General Discussion on any topic relating to CPAP and/or Sleep Apnea.
-
booksfan
- Posts: 460
- Joined: Fri Feb 09, 2018 7:20 pm
- Location: Triangle area, NC
Post
by booksfan » Tue May 28, 2019 3:16 pm
BlueDragon wrote: ↑Tue May 28, 2019 8:50 am
booksfan wrote: ↑Mon May 27, 2019 2:23 pm
Sorry, I haven't been on for a few days so I've just seen your reply.
I use a Flash SD card and FlashPap to import my data. As far as I know, the setup has always been the same,
1. Scan FlashAir SD card
2. Copy xPap data for selected date(s)
3. Exit and Run Tracking Software (I did change that from Sleepyhead to OSCAR)
Once OSCAR opens, I click on the CPAP importer as I have always done. Maybe I have directed it to the wrong location? But it doesn't look like it is labeled as a backup location.
I use FlashPap too; it is certainly useful. So C:/Users/...../Desktop/SLEEPyheadDATA is the name of your FlashPap directory?
I do see unusual messages about the period from Feb 2, 2018 through April 3,2018. But that's in the past and shouldn't affect today.
Is OSCAR consistently reporting AHI of 0.0? If you move back a day at a time, do you find a time when AHI was being reported correctly?
It is the name of the directory where the data is stored. OSCAR only reported AHI of 0.0 only the one day, May 21. Before and since, it has been reporting normally. There was a gap btw 2/21-3/13 (ish) when I was traveling and using my travel CPAP so no data and a few days when I was recovering from surgery and throwing up so I didn't use it around March 20-23. My AHI was all over the map for a few weeks then as well. Things are getting back to normal now, but not that good

, which is why it kind of raised a flag for me when I saw the 0.0.
I apologize if I gave the impression it was an on-going issue.
Mask: Bleep Eclipse/Halos
Machine: ResMed AirSense 11 AutoSet
-
ClayL
- Posts: 369
- Joined: Mon Feb 12, 2007 8:45 am
- Location: Palisade, CO now, was Full time in a motor home for 12 years- Anywhere USA
Post
by ClayL » Tue Jun 04, 2019 9:38 am
What do the two items in the bottom left of the left panel represent? The ones that say "3h 28m and 3h 56m" in the picture -

Respironics DreamStation
Pressure = 11.5 min 14.5 max
C-Flex = 1
Dreamwear FF
-
NoOnesPerfect
- Posts: 163
- Joined: Thu May 09, 2019 7:29 am
- Location: Washington, USA
Post
by NoOnesPerfect » Tue Jun 04, 2019 9:47 am
It means you turned the machine off and back on - those two times are the lengths of each session.
-
ClayL
- Posts: 369
- Joined: Mon Feb 12, 2007 8:45 am
- Location: Palisade, CO now, was Full time in a motor home for 12 years- Anywhere USA
Post
by ClayL » Tue Jun 04, 2019 10:52 am
Thanks. I forgot that I turned off to use the restroom.
Respironics DreamStation
Pressure = 11.5 min 14.5 max
C-Flex = 1
Dreamwear FF
-
palerider
- Posts: 32299
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Wed Jun 05, 2019 11:28 am
Bonjour found out why you can't load new data for a day without deleting that day...
Because Mr Watkins just didn't give a f*.
Code: Select all
auto rit = resdayList.find(date);
if (rit != resdayList.end()) {
// Already seen this record.. should check if the data is the same, but meh. <---------=
continue;
}
At least he was good enough to document that he didn't care.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
Dog Slobber
- Posts: 4269
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Post
by Dog Slobber » Fri Jun 07, 2019 11:15 am
Jas_williams wrote: ↑Fri Jun 07, 2019 11:06 am
Dave Wise wrote: ↑Fri Jun 07, 2019 9:29 am
Where does OSCAR save data? I had a mishap this morning and I want to save a copy before doing anything else.
I imported one oximeter file, commanded Rebuild, and OSCAR crashed. When I started it again, it acted like there was no data to display. Debug Pane mentioned it failed to read a summary file. I'll get details tonight.
If your using windows it’s in your Documents folder by default called OSCAR_Data
Same for Mac
Battery Backup: EcoFlow Delta 2
-
CaptPat
- Posts: 35
- Joined: Tue May 20, 2014 4:04 pm
Post
by CaptPat » Sun Jun 09, 2019 6:07 am
Bug I think
I'm using a new Aircurve 10 VAUTO, on the leftmost pane of the OSCAR screen it indicates that Climate Control in is "manual," however on the machine itself, climate control is in AUTO.
-
BlueDragon
- Posts: 547
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Post
by BlueDragon » Sun Jun 09, 2019 6:41 am
CaptPat wrote: ↑Sun Jun 09, 2019 6:07 am
Bug I think
I'm using a new Aircurve 10 VAUTO, on the leftmost pane of the OSCAR screen it indicates that Climate Control in is "manual," however on the machine itself, climate control is in AUTO.
I would need to see one of the files on your SD card to investigate this problem. I'll send you a PM about which file would be most useful.
Thanks for the report.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
-
BlueDragon
- Posts: 547
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Post
by BlueDragon » Tue Jun 11, 2019 8:40 am
CaptPat wrote: ↑Sun Jun 09, 2019 6:07 am
Bug I think
I'm using a new Aircurve 10 VAUTO, on the leftmost pane of the OSCAR screen it indicates that Climate Control in is "manual," however on the machine itself, climate control is in AUTO.
It looks like the labels are reversed. My AUTOSET is set to "manual" but is reported as "auto." I'll log this as an issue to be addressed.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
-
emcee13
- Posts: 33
- Joined: Sat May 25, 2019 10:19 am
Post
by emcee13 » Wed Jun 12, 2019 9:00 pm
I’d like to suggest a side to side scroll feature. I have to look at the charts carefully because many of my type of events just won’t get flagged. It’s very difficult to do without a way to “scrub” through the charts. That’d be a very nice addition, unless there is already a way to do this(?). Keep up the great work!
-
palerider
- Posts: 32299
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Wed Jun 12, 2019 9:45 pm
emcee13 wrote: ↑Wed Jun 12, 2019 9:00 pm
I’d like to suggest a side to side scroll feature. I have to look at the charts carefully because many of my type of events just won’t get flagged. It’s very difficult to do without a way to “scrub” through the charts. That’d be a very nice addition, unless there is already a way to do this(?). Keep up the great work!
You mean like with the arrow keys?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
emcee13
- Posts: 33
- Joined: Sat May 25, 2019 10:19 am
Post
by emcee13 » Thu Jun 13, 2019 10:09 am
palerider wrote: ↑Wed Jun 12, 2019 9:45 pm
emcee13 wrote: ↑Wed Jun 12, 2019 9:00 pm
I’d like to suggest a side to side scroll feature. I have to look at the charts carefully because many of my type of events just won’t get flagged. It’s very difficult to do without a way to “scrub” through the charts. That’d be a very nice addition, unless there is already a way to do this(?). Keep up the great work!
You mean like with the arrow keys?
I read through the instructions throughly and didn’t see this as an option. Thankfully your “question” prompted me to just try the arrow keys. It worked after some fiddling with the graph size. Thanks!
-
palerider
- Posts: 32299
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Thu Jun 13, 2019 12:10 pm
emcee13 wrote: ↑Thu Jun 13, 2019 10:09 am
palerider wrote: ↑Wed Jun 12, 2019 9:45 pm
emcee13 wrote: ↑Wed Jun 12, 2019 9:00 pm
I’d like to suggest a side to side scroll feature. I have to look at the charts carefully because many of my type of events just won’t get flagged. It’s very difficult to do without a way to “scrub” through the charts. That’d be a very nice addition, unless there is already a way to do this(?). Keep up the great work!
You mean like with the arrow keys?
I read through the instructions throughly and didn’t see this as an option. Thankfully your “question” prompted me to just try the arrow keys. It worked after some fiddling with the graph size. Thanks!
Sounds like something for the doc people
ps, up and down arrows zoom in and out on the charts.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
cacososa
- Posts: 5
- Joined: Thu May 16, 2019 4:00 pm
Post
by cacososa » Thu Jun 20, 2019 4:25 pm
Hello
I would like to report a Bug
On the Welcome tab, the text with the information about the Last Usage statistic does not read well because the background is a light yellow and the text color is white. In order to see the text I have to highlight it, otherwise you can not see it. I'm attaching a screen shot
Regards.
CES

- Screen Shot 2019-06-20 at 6.19.29 PM.png (543.9 KiB) Viewed 10752 times
-
BlueDragon
- Posts: 547
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Post
by BlueDragon » Thu Jun 20, 2019 4:34 pm
OSCAR does not support "dark mode" on the Mac. That's because the underlying development framework we use, QT, does not support dark mode. In the next release of OSCAR, we plan to include a workaround. Meanwhile, you should be able to read the messages if you set your Mac to use Light mode.
cacososa wrote: ↑Thu Jun 20, 2019 4:25 pm
I would like to report a Bug
On the Welcome tab, the text with the information about the Last Usage statistic does not read well because the background is a light yellow and the text color is white. In order to see the text I have to highlight it, otherwise you can not see it. I'm attaching a screen shot
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.