OSCAR Suggestions & Bug reporting
Re: OSCAR Suggestions & Bug reporting
Why does this not report the IPAP pressure????
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirTouch™ N20 Nasal CPAP Mask with Headgear |
- BlueDragon
- Posts: 570
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR Suggestions & Bug reporting
Certainly looks wrong, doesn't it? I can reproduce the problem and have logged this as an issue.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: OSCAR Suggestions & Bug reporting
I've been using OSCAR for a couple of weeks, backing up into Sleepyhead every week or so. I've been trying to check in pretty much daily because my AHI has not returned back to "normal"(usual?) since a recent surgery...I'm still having discomfort that disrupts my sleep. Anyway, this morning when I checked OSCAR, it said my AHI was 0.0. Lovely as that would be, I found it hard to believe.
It had been a few days, so I downloaded the past week's data into Sleepyhead. My AHI for last night was actually 0.51.
Does anyone have any ideas as to why the discrepany or should I just chalk it up to a glitch because OSCAR is in beta?
Thank you!
Thank you!
_________________
| Machine: AirSense 11 Autoset |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: AirMini for travel, with Eclipse/Halos. OSCAR software. |
Mask: Bleep Eclipse/Halos
Machine: ResMed AirSense 11 AutoSet
Machine: ResMed AirSense 11 AutoSet
- BlueDragon
- Posts: 570
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR Suggestions & Bug reporting
We need to know about these "discrepancies" so we can do something about them. So, thank you for your report with the images.booksfan wrote: ↑Thu May 23, 2019 6:20 pmI've been using OSCAR for a couple of weeks, backing up into Sleepyhead every week or so. I've been trying to check in pretty much daily because my AHI has not returned back to "normal"(usual?) since a recent surgery...I'm still having discomfort that disrupts my sleep. Anyway, this morning when I checked OSCAR, it said my AHI was 0.0. Lovely as that would be, I found it hard to believe.
It had been a few days, so I downloaded the past week's data into Sleepyhead. My AHI for last night was actually 0.51.
Does anyone have any ideas as to why the discrepancy or should I just chalk it up to a glitch because OSCAR is in beta?
Could you try this: In OSCAR, use Data/Advanced/Purge Selected Day to remove that day and then import it again. If Oscar is still showing 0.0, before exiting Oscar, select Help/Show Debug Pane. In the debug pane, right click and Select All. Paste into a text file and attach that to your reply to this message. If Oscar is still showing 0.0, we will probably need a zipped copy of your SD card to try to reproduce and debug the issue.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: OSCAR Suggestions & Bug reporting
Quick question.. In the event types What does the number in parenthesis mean?
Re: OSCAR Suggestions & Bug reporting
This is one big difference between OSCAR and SleepyHead, The OSCAR team actively desires this feedback to identify, and more importantly, correct these issues. With SleepyHead they were either fixed in the next major release or they fell into the proverbial bit bucket. FYI, most of the issues we are currently working on have been verified as existing in SleepyHead.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: VAuto 11-25cmH2O PS 3 --- OSCAR software, Many masks - Amara View FFM to P10 Pillows Several Nasal and FFM |
Re: OSCAR Suggestions & Bug reporting
How many there were.
In the event plot, the number is the ordinal number of that particular event, its position in the counting order of 1 to N.
_________________
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Additional Comments: Alg:AutoSet,Pressure:7.0-11.0,Ramp:Auto/4.0,EPR:1/FullTime,Climate:Auto |
Re: OSCAR Suggestions & Bug reporting
I had posted in thread, incorrectly, and someone suggested purging/reimporting. That did not work. (PaleRider told me to post hereBlueDragon wrote: ↑Thu May 23, 2019 6:35 pmWe need to know about these "discrepancies" so we can do something about them. So, thank you for your report with the images.booksfan wrote: ↑Thu May 23, 2019 6:20 pmI've been using OSCAR for a couple of weeks, backing up into Sleepyhead every week or so. I've been trying to check in pretty much daily because my AHI has not returned back to "normal"(usual?) since a recent surgery...I'm still having discomfort that disrupts my sleep. Anyway, this morning when I checked OSCAR, it said my AHI was 0.0. Lovely as that would be, I found it hard to believe.
It had been a few days, so I downloaded the past week's data into Sleepyhead. My AHI for last night was actually 0.51.
Does anyone have any ideas as to why the discrepancy or should I just chalk it up to a glitch because OSCAR is in beta?
Could you try this: In OSCAR, use Data/Advanced/Purge Selected Day to remove that day and then import it again. If Oscar is still showing 0.0, before exiting Oscar, select Help/Show Debug Pane. In the debug pane, right click and Select All. Paste into a text file and attach that to your reply to this message. If Oscar is still showing 0.0, we will probably need a zipped copy of your SD card to try to reproduce and debug the issue.
I tried to attach the text file to this message, but got an error message that it is an invalid file extension.
I know I could copy and paste the data right here, but it's 33 pages
BTW, it's only that one day that says 0.0, the last 2 have been fine.
_________________
| Machine: AirSense 11 Autoset |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: AirMini for travel, with Eclipse/Halos. OSCAR software. |
Mask: Bleep Eclipse/Halos
Machine: ResMed AirSense 11 AutoSet
Machine: ResMed AirSense 11 AutoSet
- BlueDragon
- Posts: 570
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR Suggestions & Bug reporting
PM sent.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: OSCAR Suggestions & Bug reporting
My goodness it's been a long time since I heard that term. Makes me feel like a young man again.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Re: OSCAR Suggestions & Bug reporting
Ive been running Sleepyhead for a number of years, and decided to Install OSCAR. The Installation seemed to be going well, with OSCAR transferring my old Sleepyhead files. However, when OSCAR opened, I got a popup telling me that OSCAR needs to upgrade its database for Philips Respironics System One, and that I need to backup my database. (I've attached a screenshot, since I can't seem to post it in the message window). Whether I answer Yes or No, a window opens showing my data file, and then OSCAR shuts down.. I've made a copy of the file, but every time I try to open OSCAR the same thing happens.
What confuses me is that I do not have a Respironics CPAP; I had been using a ResMed S9 and recently upgraded to a ResMed AirSense10. I had previously updated my profile on Sleepyhead to show the change in machines.
Any ideas on what's causing this and, mopre importantly, how can I correct this so I can use OSCAR? If it matters, my computer is running Windows 7, 64 .
What confuses me is that I do not have a Respironics CPAP; I had been using a ResMed S9 and recently upgraded to a ResMed AirSense10. I had previously updated my profile on Sleepyhead to show the change in machines.
Any ideas on what's causing this and, mopre importantly, how can I correct this so I can use OSCAR? If it matters, my computer is running Windows 7, 64 .
_________________
| Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle |
| Mask: AirFit F30 Complete Mask + AirMini Mask Setup Pack Bundle |
- Attachments
-
- OSCAR.png (412.12 KiB) Viewed 8929 times
- BlueDragon
- Posts: 570
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR Suggestions & Bug reporting
The debug log you sent me seems to show that you were importing data from the SleepyHead backup files into OSCAR, not from the SD card. Try importing from the SD card instead of reimporting from SleepyHead every time. There's generally no need to import from the Sleepyhead backups more than once to get started.BlueDragon wrote: ↑Thu May 23, 2019 6:35 pmWe need to know about these "discrepancies" so we can do something about them. So, thank you for your report with the images.booksfan wrote: ↑Thu May 23, 2019 6:20 pmI've been using OSCAR for a couple of weeks, backing up into Sleepyhead every week or so. I've been trying to check in pretty much daily because my AHI has not returned back to "normal"(usual?) since a recent surgery...I'm still having discomfort that disrupts my sleep. Anyway, this morning when I checked OSCAR, it said my AHI was 0.0. Lovely as that would be, I found it hard to believe.
It had been a few days, so I downloaded the past week's data into Sleepyhead. My AHI for last night was actually 0.51.
Does anyone have any ideas as to why the discrepancy or should I just chalk it up to a glitch because OSCAR is in beta?
Could you try this: In OSCAR, use Data/Advanced/Purge Selected Day to remove that day and then import it again. If Oscar is still showing 0.0, before exiting Oscar, select Help/Show Debug Pane. In the debug pane, right click and Select All. Paste into a text file and attach that to your reply to this message. If Oscar is still showing 0.0, we will probably need a zipped copy of your SD card to try to reproduce and debug the issue.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: OSCAR Suggestions & Bug reporting
A trendline would be great in the overview. This can be achieved in various ways through either filtering, averaging or fitting. In effect, it would give an idea of whether things are getting better, worse or are stable by de-emphasizing daily variations. I currently export my data and do this in a different package, but I think most people here would appreciate this functionality. Happy to talk more
Felix
Felix
_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: OSCAR Suggestions & Bug reporting
Sorry, I haven't been on for a few days so I've just seen your reply.BlueDragon wrote: ↑Sun May 26, 2019 2:49 pmThe debug log you sent me seems to show that you were importing data from the SleepyHead backup files into OSCAR, not from the SD card. Try importing from the SD card instead of reimporting from SleepyHead every time. There's generally no need to import from the Sleepyhead backups more than once to get started.BlueDragon wrote: ↑Thu May 23, 2019 6:35 pmWe need to know about these "discrepancies" so we can do something about them. So, thank you for your report with the images.booksfan wrote: ↑Thu May 23, 2019 6:20 pmI've been using OSCAR for a couple of weeks, backing up into Sleepyhead every week or so. I've been trying to check in pretty much daily because my AHI has not returned back to "normal"(usual?) since a recent surgery...I'm still having discomfort that disrupts my sleep. Anyway, this morning when I checked OSCAR, it said my AHI was 0.0. Lovely as that would be, I found it hard to believe.
It had been a few days, so I downloaded the past week's data into Sleepyhead. My AHI for last night was actually 0.51.
Does anyone have any ideas as to why the discrepancy or should I just chalk it up to a glitch because OSCAR is in beta?
Could you try this: In OSCAR, use Data/Advanced/Purge Selected Day to remove that day and then import it again. If Oscar is still showing 0.0, before exiting Oscar, select Help/Show Debug Pane. In the debug pane, right click and Select All. Paste into a text file and attach that to your reply to this message. If Oscar is still showing 0.0, we will probably need a zipped copy of your SD card to try to reproduce and debug the issue.
I use a Flash SD card and FlashPap to import my data. As far as I know, the setup has always been the same,
1. Scan FlashAir SD card
2. Copy xPap data for selected date(s)
3. Exit and Run Tracking Software (I did change that from Sleepyhead to OSCAR)
Once OSCAR opens, I click on the CPAP importer as I have always done. Maybe I have directed it to the wrong location? But it doesn't look like it is labeled as a backup location.
_________________
| Machine: AirSense 11 Autoset |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: AirMini for travel, with Eclipse/Halos. OSCAR software. |
Mask: Bleep Eclipse/Halos
Machine: ResMed AirSense 11 AutoSet
Machine: ResMed AirSense 11 AutoSet
- BlueDragon
- Posts: 570
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: OSCAR Suggestions & Bug reporting
I use FlashPap too; it is certainly useful. So C:/Users/...../Desktop/SLEEPyheadDATA is the name of your FlashPap directory?booksfan wrote: ↑Mon May 27, 2019 2:23 pmSorry, I haven't been on for a few days so I've just seen your reply.
I use a Flash SD card and FlashPap to import my data. As far as I know, the setup has always been the same,
1. Scan FlashAir SD card
2. Copy xPap data for selected date(s)
3. Exit and Run Tracking Software (I did change that from Sleepyhead to OSCAR)
Once OSCAR opens, I click on the CPAP importer as I have always done. Maybe I have directed it to the wrong location? But it doesn't look like it is labeled as a backup location.
I do see unusual messages about the period from Feb 2, 2018 through April 3,2018. But that's in the past and shouldn't affect today.
Is OSCAR consistently reporting AHI of 0.0? If you move back a day at a time, do you find a time when AHI was being reported correctly?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.