Adjusting For Weight Loss

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Adjusting For Weight Loss

Post by owleyes » Thu May 13, 2021 12:09 pm

I've been on cpap for almost 6 years. My weight was in the mid 170s, which is about 25 more that normal. I'm now 25 pounds lighter (not a huge difference, I know). But I'm wondering if my cpap needs may have changed. My machine is set to 15lbs, which I think is the max. I'm hoping a lower setting might be called for now. Can the Sleepyhead report show anything? I looked and didn't see anything, but frankly, I really don't know what to look for.

Thanks!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Adjusting For Weight Loss

Post by Okie bipap » Thu May 13, 2021 12:19 pm

Most of us have quit using Sleepyhead and have switched over to OSCAR. The Sleepyhead developer has quit making any updates to it, but the OSCAR team continues to update OSCAR. The only way you can tell if you need less pressure is to lower your pressure and see how many events you have. Other than that you can try to request another titration study to determine your pressure needs.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu May 13, 2021 12:25 pm

Thanks for the info. I'll try Oscar.

I was going to try that last night, but I forgot there were two numbers (7 and 15). Does it make sense to change both at once, or just the higher or lower? I was thinking of trying 5 and 12. Does that seem reasonable?

And will the stats show if there's improvement, or do I have to determine that by my eyelids :| ?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Adjusting For Weight Loss

Post by Miss Emerita » Thu May 13, 2021 12:31 pm

Congratulations on the weight loss! I would suggest that once you're up and running with Oscar, you spend at least one night at your current settings and post the daily chart so we can take a look. From there it'll be trial and error, as Okie bipap has said.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu May 13, 2021 12:35 pm

Cool. Thank you.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu May 13, 2021 1:02 pm

Is this right page to look at?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Attachments
SleepData.jpg
SleepData.jpg (365.74 KiB) Viewed 1968 times
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
zonker
Posts: 11455
Joined: Fri Jun 19, 2015 4:36 pm

Re: Adjusting For Weight Loss

Post by zonker » Thu May 13, 2021 1:41 pm

owleyes wrote:
Thu May 13, 2021 1:02 pm
Is this right page to look at?
no. we want the daily view.

sorta like this-

Image

we'll want to see how your pressure is doing currently with it's battle against your events.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Adjusting For Weight Loss

Post by Julie » Thu May 13, 2021 2:00 pm

Oh, and FYI - it's cm (of water), not lbs.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Adjusting For Weight Loss

Post by chunkyfrog » Thu May 13, 2021 3:01 pm

Julie wrote:
Thu May 13, 2021 2:00 pm
Oh, and FYI - it's cm (of water), not lbs.
15 centimeters of water cannot blow up a balloon.
15 psi (pounds per square inch) will inflate a car tire.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu May 13, 2021 6:54 pm

Okay, thanks for the corrections.

Here's my daily report.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Attachments
SleepMapperDaily.jpg
SleepMapperDaily.jpg (927.67 KiB) Viewed 1938 times
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Adjusting For Weight Loss

Post by Julie » Thu May 13, 2021 8:07 pm

Can you get rid of the snore and resp. rate graphs which aren't needed and shrink the others?

User avatar
zonker
Posts: 11455
Joined: Fri Jun 19, 2015 4:36 pm

Re: Adjusting For Weight Loss

Post by zonker » Thu May 13, 2021 8:59 pm

owleyes wrote:
Thu May 13, 2021 6:54 pm
Okay, thanks for the corrections.

Here's my daily report.
that's no where near my example. please try again.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65336
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Adjusting For Weight Loss

Post by Pugsy » Fri May 14, 2021 7:23 am

viewtopic/t158560/How-to-post-images-for-review.html

See above for examples of which graphs to include.

If you want to see if you could get by with less pressure by evaluating what the cpap machine does..
Change your minimum pressure to 10 cm and leave the max and 20 and see what happens. Normally we would suggest an even lower minimum but your current minimum is 15 and if you go a lot lower you may have trouble in terms of comfort with breathing at a lower setting. You may feel a bit stifled or suffocated. If you do at 10 cm...try 12 cm or whatever is less than 15 that is comfortable for you.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Fri May 14, 2021 7:31 am

Thank you! That's very helpful. I'll re-post my daily results shortly.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Fri May 14, 2021 12:23 pm

I'd like to re-post my chart, but I can't see what's missing.

There is no Flow Limit in my list.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10