Adjusting For Weight Loss
Re: Adjusting For Weight Loss
Respironics machines don't do a flow limitation graph like the ResMed machines do.
Instead Respironics flags flow limitations as individual events and shows them up on the Events graph...see FL line? That's flow limitations....and only in auto adjusting mode will Respironics machines flag FLs. So fixed mode (if anyone is using fixed mode and doesn't see any FLs flagged) 0 flagged FLs doesn't necessarily mean they didn't happen. .....this is fixed mode only.
You are in auto mode so if you have any FLs they will be flagged up on the Events graph.
The only graphs we really need to see from a Respironics machine report
The normal stuff on the left side
and on the right side graphs...
Events
Flow rate
Pressure
Leak
Instead Respironics flags flow limitations as individual events and shows them up on the Events graph...see FL line? That's flow limitations....and only in auto adjusting mode will Respironics machines flag FLs. So fixed mode (if anyone is using fixed mode and doesn't see any FLs flagged) 0 flagged FLs doesn't necessarily mean they didn't happen. .....this is fixed mode only.
You are in auto mode so if you have any FLs they will be flagged up on the Events graph.
The only graphs we really need to see from a Respironics machine report
The normal stuff on the left side
and on the right side graphs...
Events
Flow rate
Pressure
Leak
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Adjusting For Weight Loss
On Respironics, Flow Limitations are flagged as events an not presented as a trace graph.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Adjusting For Weight Loss
I'm trying to figure out what to show people.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: System One HT Humid, DOM (humidifier) |
Philips Respironics REMstar Auto (System One 60 Series)
Quattro Airfit F10
Quattro Airfit F10
Re: Adjusting For Weight Loss
Isn't that what I showed before?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: System One HT Humid, DOM (humidifier) |
Philips Respironics REMstar Auto (System One 60 Series)
Quattro Airfit F10
Quattro Airfit F10
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Adjusting For Weight Loss
You showed more, a lot unnecessary. Because of the extra information, the important information was compressed and difficult to read.
Put only what was asked, so the important stuff isn't squished.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Adjusting For Weight Loss
Is this better?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: System One HT Humid, DOM (humidifier) |
- Attachments
-
- Daily_Report.jpg (750.84 KiB) Viewed 1167 times
-
- P2.jpg (96.17 KiB) Viewed 1167 times
Philips Respironics REMstar Auto (System One 60 Series)
Quattro Airfit F10
Quattro Airfit F10
Re: Adjusting For Weight Loss
perfect!
is this night typical of the nights you've been having?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Adjusting For Weight Loss
Don't zoom in.
We need to see the pressure graph line for the entire night.
Otherwise the most recent screenshot shows what we like to see.
We need to see the pressure graph line for the entire night.
Otherwise the most recent screenshot shows what we like to see.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Adjusting For Weight Loss
Someone else might correct me but I do not believe Philips machines have flow limitations.
It might help if you could also put your machine & mask in your signature. Then those helping won't have to ask and it would always be available for viewing. (You can see my example and others as well)
_________________
| Machine: AirSense 11 Autoset |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: AirMini for travel, with Eclipse/Halos. OSCAR software. |
Mask: Bleep Eclipse/Halos
Machine: ResMed AirSense 11 AutoSet
Machine: ResMed AirSense 11 AutoSet
Re: Adjusting For Weight Loss
See my reply above. Respironics machines will do flow limitation flagging but it's just not recorded in the same way as ResMed records FLs.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Adjusting For Weight Loss
OH!
wondered what was up with the pressure; didn't notice it was for just a few hours.
(one of these days, i'll learn how to read oscar graphs.)
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Adjusting For Weight Loss
Until I need to know, I will remain blissfully ignorant.
Such as I can manage . . .
Such as I can manage . . .
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Adjusting For Weight Loss
my mother told me that ignorance isn't bliss, it's a blister.chunkyfrog wrote: ↑Fri May 14, 2021 10:10 pmUntil I need to know, I will remain blissfully ignorant.
Such as I can manage . . .
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Adjusting For Weight Loss
Okay, hopefully this is right.
I set my lower pressure to 12 from 15. Does this show the lower number was helpful?
Anything else?
I set my lower pressure to 12 from 15. Does this show the lower number was helpful?
Anything else?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: System One HT Humid, DOM (humidifier) |
- Attachments
-
- SleepBlurb.jpg (62.19 KiB) Viewed 1119 times
-
- NewDaily.jpg (300.23 KiB) Viewed 1119 times
Philips Respironics REMstar Auto (System One 60 Series)
Quattro Airfit F10
Quattro Airfit F10
Re: Adjusting For Weight Loss
It does appear that setting the minimum to 12 still doesn't create a situation where the machine wants to go higher.
So at this point you may not need 12 cm minimum.
Next step...try 10 cm minimum and see if the pressure graph moves upwards trying to prevent apneas.
The little spiky saw tooth increases are pressure probes and not the machine responding to airway restrictions....so ignore those.
So at this point you may not need 12 cm minimum.
Next step...try 10 cm minimum and see if the pressure graph moves upwards trying to prevent apneas.
The little spiky saw tooth increases are pressure probes and not the machine responding to airway restrictions....so ignore those.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.