Adjusting For Weight Loss

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ChicagoGranny
Posts: 15530
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Adjusting For Weight Loss

Post by ChicagoGranny » Mon May 17, 2021 9:48 am

owleyes wrote:
Mon May 17, 2021 8:53 am
What is Flex?
----------------------------->
What is Exhalation Relief like C-Flex, A-Flex, EPR, and SmartFlex?
Exhalation relief is the ability of a CPAP machine or BiPAP machine to reduce the inspiratory pressure during exhale, making it easier to breathe out against the pressure. Manufacturers have different names for exhalation relief, but they mostly work the same way. Pressure can be adjusted up to 3cm for comfort.

Respironics calls their pressure relief C-Flex, A-Flex, or C-Flex+. C-Flex lowers the pressure on exhale and quickly ramps back up to the prescribed pressure. A-Flex, available on Auto CPAP machines only, lowers the pressure on exhale and gradually increases the pressure during inhale.

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Mon May 17, 2021 11:25 am

ChicagoGranny wrote:
Mon May 17, 2021 9:41 am
owleyes wrote:
Mon May 17, 2021 8:53 am
Unfortunately I woke up far too early, unable to go back to sleep, but that's been happening lately, not related to cpap.
As you know, sleep is not all about CPAP ------------------->
What I meant was, I've made a drastic change to my diet, for the better. I've been waking much earlier than usual sometimes, but don't feel terrible as I would have in the past. I've often woken in the early morning (6 or 6:30) and I would remove the mask, roll over and get another hour or two in. I'm wondering if maybe I'm sleeping better, and 6:30 is the new 7:30. Meaning, I'm not fully rested, but too rested to go back to sleep.

I was taking melatonin, but ran out a few weeks ago. I don't like to be dependent, but everyone says it's safe. I just noticed on the forum, there's a time-release version. I wonder if that would help.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Mon Jun 07, 2021 8:03 am

Here's my recent report. Does this show anything useful?

Does it mean I can set my machine to 11 from 12? I'm not clear on how the graph works.

Thanks.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Attachments
June6Daily.jpg
June6Daily.jpg (902.21 KiB) Viewed 1473 times
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
ChicagoGranny
Posts: 15530
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Adjusting For Weight Loss

Post by ChicagoGranny » Mon Jun 07, 2021 11:35 am

owleyes wrote:
Mon Jun 07, 2021 8:03 am
Does it mean I can set my machine to 11 from 12? I'm not clear on how the graph works.
You are learning! Your pressure (IPAP) stayed at the minimum (12.0) for much of the night. If that were my chart, I would drop the minimum to 11.0 and check the results the next day.

Which position(s) do you sleep in?

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Mon Jun 07, 2021 2:01 pm

I sleep on my back unless/until I wake with the morning light and take off my mask. Then I roll over and sleep on my front (or side).

Thanks!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Adjusting For Weight Loss

Post by Dog Slobber » Mon Jun 07, 2021 2:22 pm

owleyes wrote:
Mon Jun 07, 2021 8:03 am
Does it mean I can set my machine to 11 from 12? I'm not clear on how the graph works.

By all means, drop to 11, but then leave it at 11 for a week or so.

Then compare the AHI for the week against the AHI while you were at 12. This way you aren't comparing a single night that may not be representative.

Also make note of how you slept and felt.

The OSCAR statistics page will make a new entry every time you change your machine therapy settings. Handy for comparing changes.
Battery Backup: EcoFlow Delta 2

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu Jun 10, 2021 2:32 pm

I’m on the road, so I don’t have my detailed results. But I’m using the Sleepmapper app.

My minimum pressure is set to 11.
The report says 90% pressure: 12.6,
Average CPAP pressure 11.3.

That seems odd to me.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
ChicagoGranny
Posts: 15530
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Adjusting For Weight Loss

Post by ChicagoGranny » Thu Jun 10, 2021 3:44 pm

That sounds normal. What seems odd to you?

What was your AHI?

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu Jun 10, 2021 3:47 pm

If it’s running at 12.6, 90% of the time, wouldn’t that be the average?


Sent from my iPad using Tapatalk

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Adjusting For Weight Loss

Post by Julie » Thu Jun 10, 2021 4:03 pm

That's 90% and BELOW... up to 90%, possibly only hitting it once in all.

User avatar
ChicagoGranny
Posts: 15530
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Adjusting For Weight Loss

Post by ChicagoGranny » Thu Jun 10, 2021 5:19 pm

owleyes wrote:
Thu Jun 10, 2021 3:47 pm
If it’s running at 12.6, 90% of the time, wouldn’t that be the average?
It means the pressure is 12.6 or less 90% of the time. The "or less" is the important part.

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Wed Jul 07, 2021 6:27 pm

So I've lowered my pressure to 9.5, based on results from previous settings. Does this look as though I'd be safe to continue going lower?

Out of curiosity, what is the result of going too low? Or how do I know when I get there?

Thanks.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Attachments
Oscar_Daily_7-6-21.jpg
Oscar_Daily_7-6-21.jpg (915.24 KiB) Viewed 1388 times
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: Adjusting For Weight Loss

Post by zonker » Wed Jul 07, 2021 6:35 pm

owleyes wrote:
Wed Jul 07, 2021 6:27 pm
So I've lowered my pressure to 9.5, based on results from previous settings. Does this look as though I'd be safe to continue going lower?

Out of curiosity, what is the result of going too low? Or how do I know when I get there?

Thanks.
when you go "too low", your ahi will increase.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
owleyes
Posts: 48
Joined: Thu Jun 18, 2015 9:30 am
Location: PA

Re: Adjusting For Weight Loss

Post by owleyes » Thu Jul 08, 2021 9:40 am

zonker wrote:
Wed Jul 07, 2021 6:35 pm
owleyes wrote:
Wed Jul 07, 2021 6:27 pm
So I've lowered my pressure to 9.5, based on results from previous settings. Does this look as though I'd be safe to continue going lower?

Out of curiosity, what is the result of going too low? Or how do I know when I get there?

Thanks.
when you go "too low", your ahi will increase.
Thank you, that's helpful.

Does the graph show anything about whether I've gone too low, or have more to go?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: System One HT Humid, DOM (humidifier)
Philips Respironics REMstar Auto (System One 60 Series)

Quattro Airfit F10

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Adjusting For Weight Loss

Post by Dog Slobber » Thu Jul 08, 2021 10:29 am

One of the most helpful ways to dial in therapy, make decisions based on trends and not single night anecdotal numbers is to take advantage of the Machine Setting information under the Statistics tab.

It tracks your AHI and machine settings and does the math for you?

Ensure your time frame at each level is for a few weeks.
Keep an eye on the Monthly columns to ensure your not substituting CAs for OAs (or reverse)


Machine_Settings.jpg
Machine_Settings.jpg (753.33 KiB) Viewed 1357 times
Battery Backup: EcoFlow Delta 2