dream mapper reading charts

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: dream mapper reading charts

Post by palerider » Fri Mar 08, 2019 12:26 am

Marcusstephen wrote:
Thu Mar 07, 2019 11:09 pm
palerider wrote:
Thu Mar 07, 2019 5:25 pm
Pugsy wrote:
Thu Mar 07, 2019 3:57 pm
What did you set the red line threshold at?
SH is reporting 15% of the night over the red line but the only large leak you had was at around 2AM and it was extremely brief...I don't see it being 15% of the entire night.
You can either adjust your red line number up a little or just ignore the statistic and focus on the Events graph leak flagging.
I recommend just unchecking 'flag leaks over threshold', because there is no set threshold with respironics machines, it varies constantly with the pressure.

The new OSCAR should have that turned off for non resmed machines. (maybe on for some others, but i don't know offhand which.)
The new OSCAR?
At the risk of making you think you haven't been paying attention:
viewtopic/t174897/OSCAR--Open-Source-CP ... orter.html

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

sleepy head for respironics with leak flags set to 60

Post by Marcusstephen » Fri Mar 08, 2019 4:55 pm

Another night with "gaffered trap" and leak flags set to 60 for my respironics machine.
Hows it looking?
Attachments
Screen Shot 2019-03-09 at 9.51.26 am.png
Screen Shot 2019-03-09 at 9.51.26 am.png (322.23 KiB) Viewed 5106 times
BAD BABY

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: sleepy head for respironics with leak flags set to 60

Post by palerider » Fri Mar 08, 2019 4:58 pm

Marcusstephen wrote:
Fri Mar 08, 2019 4:55 pm
Hows it looking?
What do YOU think?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: sleepy head for respironics with leak flags set to 60

Post by zonker » Fri Mar 08, 2019 5:04 pm

Marcusstephen wrote:
Fri Mar 08, 2019 4:55 pm
Another night with "gaffered trap" and leak flags set to 60 for my respironics machine.
Hows it looking?
gaffered trap would make an excellent screen name here.

or for a character in a damon runyan story.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: sleepy head for respironics with leak flags set to 60

Post by Marcusstephen » Fri Mar 08, 2019 5:49 pm

palerider wrote:
Fri Mar 08, 2019 4:58 pm
Marcusstephen wrote:
Fri Mar 08, 2019 4:55 pm
Hows it looking?
What do YOU think?
I can only go on how I feel and the AHI graph, :? I guess its not too bad?
but Im particularly interested in your opinion Palerider and Pugsy, without sounding smug if its a good result
BAD BABY

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: sleepy head for respironics with leak flags set to 60

Post by palerider » Fri Mar 08, 2019 7:14 pm

Marcusstephen wrote:
Fri Mar 08, 2019 5:49 pm
palerider wrote:
Fri Mar 08, 2019 4:58 pm
Marcusstephen wrote:
Fri Mar 08, 2019 4:55 pm
Hows it looking?
What do YOU think?
I can only go on how I feel and the AHI graph, :? I guess its not too bad?
but Im particularly interested in your opinion Palerider and Pugsy, without sounding smug if its a good result
What I'm nudging you towards is looking at the chart, and seeing what you see that looks good, bad, iffy... etc. Encouraging you to start to interpret your own data :)

How many hypopneas, obstructive apneas, is the obstructive/hypopneas under 2, (1.5 if you want a stricter criteria).
how's the leak line look, do you have much LL time flagged. (or over24lpm if using a Resmed).
Since you're on a Respironics auto, do you have many of the spiky pressure probes (more is better).

I'm not trying to get you to stop asking what we think, but rather, to get you more involved in learning what to look for, so that you're even more involved in optimizing your own therapy.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65246
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: dream mapper reading charts

Post by Pugsy » Fri Mar 08, 2019 7:21 pm

I try to look at 3 things..
The detailed data....how well I felt I slept... and how I feel in general

I can have the prettiest zero leak and zero AHI report but if I slept like crap and feel like crap then it's pretty much meaningless.

On paper you are looking good.
But there's more to this therapy than just good looking (or bad looking for that matter) than just the report all by itself.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: dream mapper reading charts

Post by Marcusstephen » Fri Mar 08, 2019 7:54 pm

OK I have delayed getting into the glossary and figuring it out for myself and tended to
depend more on opinions and yes you are correct I need to establish for myself an understanding of
the graphs and how the results have consequences for myself and what I choose to tweak in order to improve
therapy. So I got into the glossary and couldn't find "Insp time" and "Exp time" perhaps these results aren't so critical.
I will continue to study the glossary. Thank you for the Guidance Pugsy and Palerider. Im very glad I found this forum.
BAD BABY

User avatar
Pugsy
Posts: 65246
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: dream mapper reading charts

Post by Pugsy » Fri Mar 08, 2019 8:03 pm

I stick to the bare basics....Events, AHI, Leak, Pressure, Flow rate and that's pretty much it.

A lot of that other stuff that shows up in the Statistics have such a wide range of norms and we can't really do anything about it anyway so I don't bother with it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: dream mapper reading charts

Post by palerider » Fri Mar 08, 2019 11:25 pm

Marcusstephen wrote:
Fri Mar 08, 2019 7:54 pm
OK I have delayed getting into the glossary and figuring it out for myself and tended to
depend more on opinions and yes you are correct I need to establish for myself an understanding of
the graphs and how the results have consequences for myself and what I choose to tweak in order to improve
therapy. So I got into the glossary and couldn't find "Insp time" and "Exp time" perhaps these results aren't so critical.
I will continue to study the glossary. Thank you for the Guidance Pugsy and Palerider. Im very glad I found this forum.
For the most part, the rest of the charts are .... more confusion than worth. The SH replacement that's being worked on probably won't have them on the regular page, or they'll be turned off by default... Hopefully the accursed "resp rate" and "AHI" charts will go away forever.

Please do ask questions, but I'd like you to say "Ok, here's what I think about this chart"... when you post one :)

And, btw, your most recent one looks perfectly acceptable, anything under 1.0 is great, as far as I'm concerned.

Like pugsy says, though, that's not all there is to good sleep, but it looks like you're getting the breathing part of it under control.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: dream mapper reading charts

Post by Marcusstephen » Fri Mar 08, 2019 11:38 pm

palerider wrote:
Fri Mar 08, 2019 11:25 pm
Marcusstephen wrote:
Fri Mar 08, 2019 7:54 pm
OK I have delayed getting into the glossary and figuring it out for myself and tended to
depend more on opinions and yes you are correct I need to establish for myself an understanding of
the graphs and how the results have consequences for myself and what I choose to tweak in order to improve
therapy. So I got into the glossary and couldn't find "Insp time" and "Exp time" perhaps these results aren't so critical.
I will continue to study the glossary. Thank you for the Guidance Pugsy and Palerider. Im very glad I found this forum.
For the most part, the rest of the charts are .... more confusion than worth. The SH replacement that's being worked on probably won't have them on the regular page, or they'll be turned off by default... Hopefully the accursed "resp rate" and "AHI" charts will go away forever.

Please do ask questions, but I'd like you to say "Ok, here's what I think about this chart"... when you post one :)

And, btw, your most recent one looks perfectly acceptable, anything under 1.0 is great, as far as I'm concerned.

Like pugsy says, though, that's not all there is to good sleep, but it looks like you're getting the breathing part of it under control.
My first question is the Green and Red lines in the Pressure graph. Why are there 2 lines?
The second question is to do with Leak Rate having an upper and lower graph line also?
BAD BABY

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: dream mapper reading charts

Post by palerider » Fri Mar 08, 2019 11:44 pm

Marcusstephen wrote:
Fri Mar 08, 2019 11:38 pm
My first question is the Green and Red lines in the Pressure graph. Why are there 2 lines?
The second question is to do with Leak Rate having an upper and lower graph line also?
Confusing things, yes.

First, the green and red lines, red is your actual auto adjusting pressure, then when you use *flex (probably aflex), it drops the pressure a bit when you start to exhale, and that reduced pressure is shown by the green line.

There are two leak lines because your machine only reports *total* air flow, whereas ResMed report 'excess' airflow, or leak. The SH software *guesses* at what is probably the excess (leak) airflow amounts, based on average mask vent rates.

The top line is the reported total airflow from the machine. The bottom line is the guess at how much of that is excess (leak).

Make sense?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: dream mapper reading charts

Post by zonker » Sat Mar 09, 2019 11:42 am

palerider wrote:
Fri Mar 08, 2019 11:44 pm

First, the green and red lines, red is your actual auto adjusting pressure, then when you use *flex (probably aflex), it drops the pressure a bit when you start to exhale, and that reduced pressure is shown by the green line.
pardon me, mark. palerider? this does make sense. look, i know i've been here for over three years, but that doesn't mean i retain much. so i was happy to read this explanation.

carry on!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: dream mapper reading charts

Post by palerider » Sat Mar 09, 2019 1:57 pm

zonker wrote:
Sat Mar 09, 2019 11:42 am
palerider wrote:
Fri Mar 08, 2019 11:44 pm

First, the green and red lines, red is your actual auto adjusting pressure, then when you use *flex (probably aflex), it drops the pressure a bit when you start to exhale, and that reduced pressure is shown by the green line.
pardon me, mark. palerider? this does make sense. look, i know i've been here for over three years, but that doesn't mean i retain much. so i was happy to read this explanation.

carry on!
All you have to do is ask :D

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: dream mapper reading charts

Post by Marcusstephen » Sat Mar 09, 2019 4:30 pm

palerider wrote:
Sat Mar 09, 2019 1:57 pm
zonker wrote:
Sat Mar 09, 2019 11:42 am
palerider wrote:
Fri Mar 08, 2019 11:44 pm

First, the green and red lines, red is your actual auto adjusting pressure, then when you use *flex (probably aflex), it drops the pressure a bit when you start to exhale, and that reduced pressure is shown by the green line.
pardon me, mark. palerider? this does make sense. look, i know i've been here for over three years, but that doesn't mean i retain much. so i was happy to read this explanation.

carry on!
All you have to do is ask :D
thanks again Palerider, that makes a lot of sense.
BAD BABY