S9 Autoset Shutting Off

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BigLou
Posts: 63
Joined: Thu Jun 28, 2012 9:05 pm

Re: S9 Autoset Shutting Off

Post by BigLou » Sun Jul 01, 2012 9:16 pm

It's the machine not the software. I was using RS v.3.14 earlier in June (was using Ramp then) and then updated to v4.1 a few days ago. All the data from both versions is stored on my computer. During the use of v.3.14 and v4.1 the machine just seems to record that 5-10 minute gap whenever. When I was ramping before, I know the machine must have malfunctioned because it would start back up from the set ramp upon the initiation of data recording. Auto start/stop has been activated and has been working since the beginning. I don't power off my machine during the night and uncouple from the machine only to pee if needed. I always look at the clock when I do this and the times just don't correlate with the gaps. What's up with this?
:) :) :)

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 Autoset Shutting Off

Post by archangle » Mon Jul 02, 2012 6:35 am

DoriC wrote:I also don't think it's anything to do with power blips because I have one of those alarms plugged in to wake me in case of a power outage. It's very sensitive and did go off a few months ago when we had a brief 5 second outage. I couldn't even see it on the data if I remember.
A much shorter power blip might confuse the machine and not set off the alarm. Also a "brownout" vs. a "blackout" might confuse the machine and not set off the alarm. I don't necessarily think that's the cause, though.

You don't happen to have a battery backup system, do you?
DoriC wrote:@Arch, how much do you charge and when will you arrive here to do my checkup?
Free. Just send first class air tickets and a prepaid rental car reservation. I'll need a recliner to sleep in, too.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 Autoset Shutting Off

Post by DoriC » Mon Jul 02, 2012 11:53 am

[quote="archangle"][quote="DoriC"]You don't happen to have a battery backup system, do you?

http://www.officedepot.com/a/products/2 ... A-Battery/

Can't believe you're asking that! Our wonderful aide went and bought this one at Radio Shack this morning and just hooked it up, they told her it would serve our purpose. It looks a lot better than the crummy cheap surge protector I was using which the guy at Radio Shack said wasn't working so giving no protection at all.

As for your accomodations, I will supply Mike's recliner which comes complete with cozy blankie!!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: S9 Autoset Shutting Off

Post by MaxDarkside » Mon Jul 02, 2012 12:51 pm

A few months ago my S9 Autoset rebooted during sleep. It woke me. I don't know what the issue was, it must have had a fault. It hasn't done it since.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: S9 Autoset Shutting Off

Post by chunkyfrog » Mon Jul 02, 2012 1:45 pm

Aliens, Max. Be calm. Act nonchalant. Maybe they'll go away.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
MaxDarkside
Posts: 1199
Joined: Sun Dec 18, 2011 4:21 pm
Location: Minneapolis, MN

Re: S9 Autoset Shutting Off

Post by MaxDarkside » Mon Jul 02, 2012 2:45 pm

chunkyfrog wrote:Aliens, Max. Be calm. Act nonchalant. Maybe they'll go away.
LOL... Probably the CIA_FBI_NSA remotely installing new firmware through the power lines to tap my brain for Alpha-Bravo classified information and reprogram my brain as a sleeper agent bot. I don't success they were thinkful... ful.. ful.. (poof)

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50E Oximeter, Zeo Bedside, Software: Comm'l grade AI analytics server & tools & SleepyHead
Do or Die... Sleep Apnea killed me, but I came back. Click for my story
Please visit my My Apnea Analytics blog. Maybe we can help each other.
54 yrs, 6' 1", 160->172 lbs

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 Autoset Shutting Off

Post by archangle » Tue Jul 03, 2012 4:00 am

DoriC wrote: Can't believe you're asking that! Our wonderful aide went and bought this one at Radio Shack this morning and just hooked it up, they told her it would serve our purpose. It looks a lot better than the crummy cheap surge protector I was using which the guy at Radio Shack said wasn't working so giving no protection at all.
Good. You do realize it will only power the machine for maybe a half hour or so if the power goes off, don't you?

Plug the power outage alarm into the wall, not the UPS.

Good point about the surge protector. Most of them have a light that will go out if it's not surge protecting any more. Everyone needs to figure out where this light is and how to check it. Lots of surge protectors have a little gadget inside that absorbs surges and often sacrifices itself in the process.
DoriC wrote:As for your accomodations, I will supply Mike's recliner which comes complete with cozy blankie!!
Good. Can I bring taco puppy along, too?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

-tim
Posts: 710
Joined: Sat Dec 03, 2011 7:46 pm
Location: Melbourne Australia

Re: S9 Autoset Shutting Off

Post by -tim » Tue Jul 03, 2012 4:52 am

If you unplug the S9 to reset it, press the power button. The thing uses a computer chip that can run off nearly no power and the 24V->3.3V supply could keep the computer happy in sleep mode for weeks. Pressing a button will wake it up and it will drain all the stored power.

If you have a surge problem, Is it a window A/C or whole house one? Window units tend to be very noisy and in that case find the circuit breaker that your A/C and O2 generator are on. Unplug most things, plug a light in where the A/C or O2 gen is and start flipping circut breakers until it goes off and write down the position and then try the other devices include the cpap. They may be on different sides or on the same side. If they are all on one side, find an outlet that isn't on that side as well. Post a picture of your breaker box and someone here might be able to tell you which breakers on different phases which could help.

Another solution is plug in a long extension cord into an outlet on the other side of the house. It is not recommended by any CPAP manufacture or your local fire brigade but it can reduce some types of power surges and if it works, tell us and we can let you know your options such as a UPS or house surge protector. Your S9 power brick doesn't care about high voltage spikes. Mine sees 377 Volts all day every day. It will care about power dropouts but those tend to show up with dimming light bulbs (if you have any of the old type left)

Your best option is get a bit of card board and cut it out about the size of the machine and stick it on top. If its not there in the morning, you don't have surges.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmeds overpriced SpO2

jean126
Posts: 2
Joined: Fri Sep 30, 2011 8:46 pm

Re: S9 Autoset Shutting Off

Post by jean126 » Tue Jul 03, 2012 8:19 am

My S9 autoset was doing the same thing for months. I first I thought i was shutting it off in my sleep, but after moving it to where that wasn't possible it was still happening every night. I took it in and they reviewed the data, and called resmed. I ended up getting a new machine. Since February I have not had that issue. My humidifier decided to quit few weeks ago and they replaced that as well. I worry when i am not under the warranty.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 Autoset Shutting Off

Post by DoriC » Tue Jul 03, 2012 10:21 am

archangle wrote:
DoriC wrote: Good point about the surge protector. Most of them have a light that will go out if it's not surge protecting any more. Everyone needs to figure out where this light is and how to check it. Lots of surge protectors have a little gadget inside that absorbs surges and often sacrifices itself in the process.
DoriC wrote:As for your accomodations, I will supply Mike's recliner which comes complete with cozy blankie!!
Good. Can I bring taco puppy along, too?
I knew the light on the surge protector wasn't working anymore but I didn't think about it! Dont know if it's coincidence but the data for last night looks perfect, the machine sounded even more quiet, best night in a long time!

I've got a better idea, send me the puppy and you can stay home!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: S9 Autoset Shutting Off

Post by archangle » Tue Jul 03, 2012 7:04 pm

DoriC wrote:I've got a better idea, send me the puppy and you can stay home!
She isn't qualified to diagnose electrical problems, but I'm sure she'd be glad to add some extra holes to your mask and hose.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 Autoset Shutting Off

Post by DoriC » Sat Jul 07, 2012 3:16 pm

Update! After another night of intermittent shut offs, I called for an electrician (we live in a condo assoc so that service is provided). Guess what, we had a faulty outlet with a half broken wire which was causing intermittent "electrical vibrations"( I think that's what he said). Anyway, he replaced the outlet, checked out everything else and last night was uneventful! Arch, why didn't you think of that!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: S9 Autoset Shutting Off

Post by HoseCrusher » Sat Jul 07, 2012 4:49 pm

Actually...

I was just going to mention that...

Glad you have this problem resolved.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: S9 Autoset Shutting Off

Post by DoriC » Sat Jul 07, 2012 5:28 pm

Hose Crusher, Remember what happens to hoses when you lie!!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: S9 Autoset Shutting Off

Post by RandyJ » Sat Jul 07, 2012 5:39 pm

Wow DoriC, I'm glad you figured it out... the last thing I would have thought of was faulty wiring in the house...