SleepyHead Qt v0.8 alpha

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
fiberfan
Posts: 262
Joined: Sat Feb 13, 2010 2:50 pm
Location: UT

Re: SleepyHead Qt v0.8 alpha

Post by fiberfan » Fri Aug 19, 2011 12:51 pm

SleepyHead would be great for beta software, for alpha it is amazing. I have had no problems importing my S9 VPAP Auto data. I write software for a living so I am almost surprised to have no problems in code as new as the importer for the S9 VPAP Auto.

Thanks!!!!!!!!!!!!!!

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.14 and SleepyHead software.
So many ideas, so much fiber, so little time - http://fiberfan.blogspot.com/

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Qt v0.8 alpha

Post by Pugsy » Fri Aug 19, 2011 1:11 pm

jedimark wrote:Just confirming this would not lead to good things happening... :-}
Understood. I am good. I pinned to Start Menu the appropriate exe file and tested multiple times. Always get 8.3 now.
I find this much easier than looking in folder each time. No chance for screw up again. I could put a shortcut on desktop like I have in the past.

Husband is home now. No time to do some instructions but I will have time Sunday after he leaves. He whines too much when I am on computer and it distracts me. Gotta go.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: SleepyHead Qt v0.8 alpha

Post by archangle » Fri Aug 19, 2011 2:52 pm

You see, Mark? I told you that putting the "randomly switch software versions on certain dates" gag would fool a bunch of people.

I can't wait until we spring the next gag on the users. It's a doozy.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by jedimark » Fri Aug 19, 2011 7:22 pm

archangle wrote:You see, Mark? I told you that putting the "randomly switch software versions on certain dates" gag would fool a bunch of people.

I can't wait until we spring the next gag on the users. It's a doozy.
Lol... If only I was that funny.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by Bookbear » Fri Aug 19, 2011 7:30 pm

Mark, could you tell me what files a PR System 1 BI-PAP would store on the SD card for a night's session? I know that the cpap and apap versions create three files ending in .001, .002, and .005 for each session. Are the ones for the BI-PAP the same? Or are there different files, or additional files?

I am trying to help someone who is having a problem with the card. It only has one night's data, and it's incomplete from what I can tell... only the .001 and .002 files. What should we be seeing?

Thanks!

_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by jedimark » Fri Aug 19, 2011 9:53 pm

Bookbear wrote:Mark, could you tell me what files a PR System 1 BI-PAP would store on the SD card for a night's session? I know that the cpap and apap versions create three files ending in .001, .002, and .005 for each session. Are the ones for the BI-PAP the same? Or are there different files, or additional files?

I am trying to help someone who is having a problem with the card. It only has one night's data, and it's incomplete from what I can tell... only the .001 and .002 files. What should we be seeing?

Thanks!
BIPAP is the same.. It will only record the .005 waveform files if the card is inserted before the session starts. Sometimes it will screw up if the card is pulled out to quickly after a session.

Depending on the devices family code the .005 will contain just the flow rate waveform, or also contain mask pressure (as in the case of family 5 ASV devices)

The other two file types (the .001 & .002) are all that's needed for calculations.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by robysue » Sat Aug 20, 2011 8:35 pm

Just want to give an update since I'd been reporting bugs on an earlier version of SleepyHead 0.8.

The problems on my Eee PC with the graphics getting stuck and not changing were fixed in 0.8.2 and remain fixed in 0.2.3.

But there's still one annoying bug that I don't think has been reported: If SH 0.8.3 is open when my computer goes to sleep or when I close the top, when I look at SH after waking the machine up, it's all goofy: The graphs are either blacked out or not there at all---when I click on "Overview", the splash page shows up. When I click back on "Daily" the Overview panel still shows through the part of the window where the graphs are, the Overview Panel is still visible. The sidebar with the summary data shows up, however. And changes appropriately when you click on a different day.

Haven't kept up with this thread at all. I notice the long term summary data that was in SH 0.7 is gone. I assume that it will be back (eventually) when other more pressing things are dealt with.

Finally,
fiberfan wrote:SleepyHead would be great for beta software, for alpha it is amazing. I have had no problems importing my S9 VPAP Auto data. I write software for a living so I am almost surprised to have no problems in code as new as the importer for the S9 VPAP Auto.

Thanks!!!!!!!!!!!!!!
I agree 100% with this. SleepyHead is amazing for alpha and as it stands would be a respectable piece of beta software. Keep up the good work Jedimark. We're really lucky you've been willing to take on this project and we're even luckier that you've been determined to do it right. Thanks!

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by GumbyCT » Sat Aug 20, 2011 9:00 pm

jedimark wrote:Sometimes it will screw up if the card is pulled out to quickly after a session.
Bookbear fwiw- I have seen new users, eager to see the data, remove the card too soon after turning the machine off and before the data is written to the card. The machine is not sure if the current session is complete just yet and may need up to an hour to finalize the session.

The good thing is - these new fangled machines are sposed to write the data when the card is re-inserted.

The bad thing is - I think removing the card when the machine is writing to it results in damaged and unreadable cards and many are not figuring out why yet.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
Breathe Jimbo
Posts: 954
Joined: Sun Sep 12, 2010 3:02 pm
Location: Pasadena, CA

Re: SleepyHead Qt v0.8 alpha

Post by Breathe Jimbo » Sat Aug 20, 2011 9:24 pm

GumbyCT wrote:I have seen new users, eager to see the data, remove the card too soon after turning the machine off and before the data is written to the card. The machine is not sure if the current session is complete just yet and may need up to an hour to finalize the session.
Ah, yes, a bathroom break or letting the dogs out in the middle of the night.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap

User avatar
Otter
Posts: 1063
Joined: Sat Mar 19, 2011 5:02 pm

Re: SleepyHead Qt v0.8 alpha

Post by Otter » Sat Aug 20, 2011 10:06 pm

GumbyCT wrote:Bookbear fwiw- I have seen new users, eager to see the data, remove the card too soon after turning the machine off and before the data is written to the card. The machine is not sure if the current session is complete just yet and may need up to an hour to finalize the session.
Pushing the stop button doesn't give it a clue?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Software: SleepyHead 0.9 beta

User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Re: SleepyHead Qt v0.8 alpha

Post by Bookbear » Sun Aug 21, 2011 10:10 am

Thanks... I'll pass that tip along.

_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

User avatar
Otter
Posts: 1063
Joined: Sat Mar 19, 2011 5:02 pm

Re: SleepyHead Qt v0.8 alpha

Post by Otter » Sun Aug 21, 2011 2:44 pm

I enabled "display crashy snapshots" today to see what it was. Never did figure that out, but I guess it works. SH crashed.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Software: SleepyHead 0.9 beta

User avatar
came;ady
Posts: 3
Joined: Thu May 19, 2011 2:27 pm
Location: NE Wisconsin

Re: SleepyHead Qt v0.8 alpha

Post by came;ady » Sun Aug 21, 2011 6:46 pm

Greetings all,
Still can't get SH to import, and I'm wondering if it's because my SD card doesn't show .005 files, it shows .004/b] along with the .001 and .002

Don't know what .004 is supposed to show but could this be the reason SH doesn't find anything? I'll try dealing with the DME tomorrow but based on the past six months, I don't hold out any hope of finding anything out.

Thanks to all who have helped me so far.
Linda

pharm30
Posts: 59
Joined: Wed Aug 17, 2011 12:36 am

Re: SleepyHead Qt v0.8 alpha

Post by pharm30 » Mon Aug 22, 2011 11:22 am

anyone want to help someone who has no clue what I just read on any of the previous pages?
I tried to download the zip files and the program would just stop working. I wasn't sure
how to locate a folder for my card reader? I can find the device, but don't know how to extract the data.

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepyHead Qt v0.8 alpha

Post by Pugsy » Mon Aug 22, 2011 12:45 pm

pharm30 wrote:anyone want to help someone who has no clue what I just read on any of the previous pages?
I tried to download the zip files and the program would just stop working. I wasn't sure
how to locate a folder for my card reader? I can find the device, but don't know how to extract the data.
Could you please confirm which machine you are using??
The one showing in your profile is not supported by SleepyHead.
If you use the DT3500 card reader, out of luck.
Now if you have one of the new PR System One machines...maybe I can help. I have screen shots in a couple of my posts (scroll back in the posts to see) and the last one shows the "folder" that SH .0.8.3 needs.
Also you might try 8.1 as it is a bit easier to manage for a newbie.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.