Rescan Software

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
qjosea
Posts: 82
Joined: Mon Oct 05, 2009 7:42 pm
Location: Wilmington, N.C.

Rescan Software

Post by qjosea » Wed Dec 16, 2009 12:47 pm

I installed the software and bought a RS232 F to F DB9 Cable and cannot get my computer to recognize my machine. I have powered off the computer, then restarted it connected the cable to my machine, turned the xpap and my computer has not recognized it yet. I have tried everything I know, can anyone help with some insight into what may be causing this. My equipment info is listed on the bottom of this posting.
Thanks,

_________________
MaskHumidifier
Additional Comments: S9 VPAP Adapt

sronsen
Posts: 105
Joined: Thu Aug 13, 2009 4:08 pm
Location: Buffalo, NY

Re: Rescan Software

Post by sronsen » Wed Dec 16, 2009 1:08 pm

qjosea,

What kind of PC and what OS are you using? Need this info to trouble-shoot for you.

Steve

_________________
MachineMask
Additional Comments: SleepyHead v1.0.0

User avatar
qjosea
Posts: 82
Joined: Mon Oct 05, 2009 7:42 pm
Location: Wilmington, N.C.

Re: Rescan Software

Post by qjosea » Wed Dec 16, 2009 1:32 pm

HP with XP

tHANKS

_________________
MaskHumidifier
Additional Comments: S9 VPAP Adapt

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Rescan Software

Post by JohnBFisher » Wed Dec 16, 2009 2:01 pm

qjosea wrote:... HP with XP ...
Well, it is possible the HP computer does not see the serial port. You may need to shutdown your computer and enter the BIOS menu. You can see if it can be enabled from the BIOS menu. However, I need to warn you that sometimes with XP, I have seen times when XP did not want to recognize a serial port on the device, regardless of the BIOS settings. In those cases, I purchased a serial to USB adapter to allow me to connect a serial device to my computer. An example is:

http://www.newegg.com/Product/Product.a ... -_-Product

And that is exactly what I had to do to connect my ASV to my HP laptop to read the data via the serial port.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: Rescan Software

Post by dsm » Wed Dec 16, 2009 3:17 pm

Can you try any other device on the serial port just to see it works ?

I have several XP based notebooks & servers & all seem to recognise the ports ok + I can connect Vpap III, S7 Spirit & Vpap Auto SV - when I run Rescan it always sees any of these using std null modem db-9pin f-to-f cable.

Good luck.

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Rescan Software

Post by Goofproof » Wed Dec 16, 2009 3:19 pm

That is why the computer world has evolved to USB & Firewire, because of the flaky Serial connection game. Your serial ports need to be enabled in the BIOs, Then it played and played with, until it works, or you give up and buy a USB card reader. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Rescan Software

Post by JohnBFisher » Wed Dec 16, 2009 5:36 pm

Goofproof wrote:... or you give up and buy a USB card reader. Jim ...
Jim, in this case the user may need to get a serial to USB adapter. The machine does not have a card. It has a DB9 Serial port on the back of the machine. I use one of those adapters to connect my machine to my laptop to read the data off the machine.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

sronsen
Posts: 105
Joined: Thu Aug 13, 2009 4:08 pm
Location: Buffalo, NY

Re: Rescan Software

Post by sronsen » Wed Dec 16, 2009 5:51 pm

qjosea,

After checking the BIOS, as suggested by JohnBFisher, open the XP device manager and verify that the com port is enabled and that its settings are correct. The speed, data bits, parity, stop bits and flow control should match the VPAP's com requirements, which should be in your documentation. If you need more help, just ask.

Steve

_________________
MachineMask
Additional Comments: SleepyHead v1.0.0

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: Rescan Software

Post by dsm » Wed Dec 16, 2009 7:47 pm

am pretty sure the Resmed software works out & sets the speed. have never had to set any values for the serial port myself.
also have used AutoScan 3.0 thru to the ResScan 3.7 & have always just plugged the null-modem cable in (done this with over
5 different S7 based models) & it has always detected them. I have used those same serial ports for communicating at other
speeds with a Nellcor SpO2 device. As said, the software sets the port to the required settings & signals the flow processor.

My 1st suspect would be the cable that was provided. 2nd suspect would be the serial port setting (these are normally active
by default, but I tend to use com1. I'll try to get some time this w/e to look at the various versions of the software & how
they set the port.

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
qjosea
Posts: 82
Joined: Mon Oct 05, 2009 7:42 pm
Location: Wilmington, N.C.

Re: Rescan Software

Post by qjosea » Thu Dec 17, 2009 12:11 pm

I went a bought serial to usb and still could not get it to work. Maybe I will try my wifes laptop, she is running windows vista. I have a mac, I don't belive the rescan software will work with a mac. I can be wrong, does anynoe know the answer to this.

thanks,

_________________
MaskHumidifier
Additional Comments: S9 VPAP Adapt

User avatar
GaryG
Posts: 537
Joined: Thu Jul 23, 2009 5:55 am
Location: CT

Re: Rescan Software

Post by GaryG » Thu Dec 17, 2009 12:20 pm

qjosea wrote:I went a bought serial to usb and still could not get it to work. Maybe I will try my wifes laptop, she is running windows vista. I have a mac, I don't belive the rescan software will work with a mac. I can be wrong, does anynoe know the answer to this.

thanks,
No, need to run the Resscan on an Windows OS. But I run it very successfully with Boot Camp partition with XP on my mac.
Still, its worth a try on your wife's machine, but I recall some folks having a challenge on Vista.

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Re: Rescan Software

Post by ozij » Thu Dec 17, 2009 12:57 pm

When you turn on the machine, does the intital screen show the last average pressure you used (or pressure:0 if it after noon)? If it does, press exit to get to the screen which say "Settling:ON" (of off, as the case may be) and "menu" on the left and "exit" on the right. The machine won't communicate with the computer when it's showing recent pressure info.

The can get rid of that by changing the "auto" or "autoappear" in the smartdata options menu to off. It's counter intuitive, but this option does not change the settings.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
qjosea
Posts: 82
Joined: Mon Oct 05, 2009 7:42 pm
Location: Wilmington, N.C.

Re: Rescan Software

Post by qjosea » Sat Dec 19, 2009 2:36 pm

Finally got Rescan software to work. For those who have problems and don't want to purchase the cable to download date from ResMed Vpap Adapt SV. I went to Radio Shack others can choose to purchase from another source, you will need to DB 9 Cable Female to Female cable (RS 232) and a null modem adapter. The cable is $19.98, and the Null Modem is $10.99 which by the way is less expensive that $63.00 plus S/H is you get it from the net. It works, I was able to download my date and if need arises I can make any adjustments on my machine. Thanks to this site, I have found that any need for advise that arises can be found just by posting the question on this site.

_________________
MaskHumidifier
Additional Comments: S9 VPAP Adapt

DreamOn
Posts: 1920
Joined: Fri Oct 23, 2009 10:13 am

Re: Rescan Software

Post by DreamOn » Sat Dec 19, 2009 2:39 pm

qjosea wrote:Finally got Rescan software to work. For those who have problems and don't want to purchase the cable to download date from ResMed Vpap Adapt SV. I went to Radio Shack others can choose to purchase from another source, you will need to DB 9 Cable Female to Female cable (RS 232) and a null modem adapter. The cable is $19.98, and the Null Modem is $10.99 which by the way is less expensive that $63.00 plus S/H is you get it from the net. It works, I was able to download my date and if need arises I can make any adjustments on my machine. Thanks to this site, I have found that any need for advise that arises can be found just by posting the question on this site.
Fantastic, qjosea! Glad to hear that you got it working, and thanks for sharing the information with others. I agree, this forum is the greatest!

User avatar
secret agent girl
Posts: 574
Joined: Tue Nov 10, 2009 2:15 pm

.

Post by secret agent girl » Sat Dec 19, 2009 7:34 pm

.
Last edited by secret agent girl on Sun Feb 06, 2011 8:41 pm, edited 1 time in total.