Sleep study inaccurate

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Tec5
Posts: 173
Joined: Mon Mar 28, 2022 10:32 am

Re: Sleep study inaccurate

Post by Tec5 » Thu Jun 16, 2022 10:50 am

Grouchy Bear wrote: ↑
Wed Jun 15, 2022 4:31 pm
So, could comparing belt movement with a previous breath and finding about the same movement, but no/very little flow through the nasal cannula could indicate either mouth breathing or an apnea? I think it gets complicated in that case, but one might be able to discern which is actually happening, with the detailed test results and graphs available to study.
Yes, that's seems reasonable to infer, particularly if the first belt expansion had a simultaneous recorded flow but the on the second expansion there was no recorded flow at the nostrils. But the algorithm to do that "look back" would be complicated, and probably not cost-effective.

I'd be really be surprised that the HSAT employed RIP belts (way too expensive and overly complicated data processing), more likely a strain-gage sensor built into the belt that measure changes in the length of the belt, or a piezoelectric transducer.

Lastly I agree with @Rubicon, the OP is likely misremembering, there is only one belt on the Resmed HSAT.

One thing that I would expect that the HSAT belt should be very good at is observing Central Apneas (because the chest is not being commanded to expand at all)
I am neither a physician nor a lawyer, so DO NOT rely on me for professional medical or legal advice.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Thu Jun 16, 2022 3:54 pm

The concluding episode of this thread can be found at viewtopic.php?f=1&t=184526&p=1413923#p1413917.
Sanjay's doctor has ordered another sleep study be cause his oral breathing was not measured in the original study.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Sleep study inaccurate

Post by Dog Slobber » Thu Jun 16, 2022 4:36 pm

dataq1 wrote: ↑
Thu Jun 16, 2022 3:54 pm
The concluding episode of this thread can be found at viewtopic.php?f=1&t=184526&p=1413923#p1413917.
Sanjay's doctor has ordered another sleep study be cause his oral breathing was not measured in the original study.
Could you quote that from the post, because thats not what he posted.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Sleep study inaccurate

Post by zonker » Thu Jun 16, 2022 5:21 pm

Dog Slobber wrote: ↑
Thu Jun 16, 2022 4:36 pm
dataq1 wrote: ↑
Thu Jun 16, 2022 3:54 pm
The concluding episode of this thread can be found at viewtopic.php?f=1&t=184526&p=1413923#p1413917.
Sanjay's doctor has ordered another sleep study be cause his oral breathing was not measured in the original study.
Could you quote that from the post, because thats not what he posted.
but it's what dataq1 read.
:roll:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Fri Jun 17, 2022 10:49 am

zonker wrote: ↑
Thu Jun 16, 2022 5:21 pm
Dog Slobber wrote: ↑
Thu Jun 16, 2022 4:36 pm
dataq1 wrote: ↑
Thu Jun 16, 2022 3:54 pm
The concluding episode of this thread can be found at viewtopic.php?f=1&t=184526&p=1413923#p1413917.
Sanjay's doctor has ordered another sleep study be cause his oral breathing was not measured in the original study.
Could you quote that from the post, because thats not what he posted.
but it's what dataq1 read.
Is there an alternate way of describing in a few words what Sanjay posted at viewtopic.php?f=1&t=184526&p=1413923#p1413917?
In any event, it would appear that he got resolution to his issue.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Sleep study inaccurate

Post by zonker » Fri Jun 17, 2022 10:55 am

dataq1 wrote: ↑
Fri Jun 17, 2022 10:49 am

Is there an alternate way of describing in a few words what Sanjay posted at viewtopic.php?f=1&t=184526&p=1413923#p1413917?
In any event, it would appear that he got resolution to his issue.
i should think that if someone said something in another thread, one would want to quote the thread rather than paraphrasing.

also, not directed to you but to members new to the forum, this is yet another reason why a person new to the forum should stick to one thread. it cuts down on confusion and helps us old timers keep track of what is what.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Sleep study inaccurate

Post by Dog Slobber » Fri Jun 17, 2022 10:58 am

dataq1 wrote: ↑
Fri Jun 17, 2022 10:49 am
zonker wrote: ↑
Thu Jun 16, 2022 5:21 pm
Dog Slobber wrote: ↑
Thu Jun 16, 2022 4:36 pm
dataq1 wrote: ↑
Thu Jun 16, 2022 3:54 pm
The concluding episode of this thread can be found at viewtopic.php?f=1&t=184526&p=1413923#p1413917.
Sanjay's doctor has ordered another sleep study be cause his oral breathing was not measured in the original study.
Could you quote that from the post, because thats not what he posted.
but it's what dataq1 read.
Is there an alternate way of describing in a few words what Sanjay posted at viewtopic.php?f=1&t=184526&p=1413923#p1413917?
In any event, it would appear that he got resolution to his issue.
Yes, and there also is an honest way.

Your paraphrase was dishonest.
Battery Backup: EcoFlow Delta 2

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Sleep study inaccurate

Post by jimbud » Fri Jun 17, 2022 11:54 am

dataq1 wrote: ↑
Fri Jun 17, 2022 10:49 am
Is there an alternate way of describing in a few words what Sanjay posted?
by Sanjay1976 » Thu Jun 16, 2022 2:55 pm

Hello all,
I share an office visit. I was breathing through my mouth. So I spoke. I could have a thing (strap?)

He was very kind and said that they would charge insurance for the second sleep study.

I am happy.

help

Sanjay
How is that for alternate way of describing in a few words what Sanjay posted.

Is it honest? Well, just about as honest as your "paraphrase". :wink:

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Fri Jun 17, 2022 12:16 pm

First I don't understand why y'all are making such a fuss about honesty.

Second: Jimbud, you failed to include:

"they would NOT charge insurance ....because it was THEIR fault"

Rather important to include that.

I interpret the "their fault" as an acknowledgment that "they" did not account for his oral breathing.

Anyone who wants to can read his actual post and draw their own interpretation.

Now can we get off the "Dataq1-bashing soap box"
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Sleep study inaccurate

Post by Rubicon » Fri Jun 17, 2022 1:45 pm

dataq1 wrote: ↑
Fri Jun 17, 2022 12:16 pm
Second: Jimbud, you failed to include:

"they would NOT charge insurance ....because it was THEIR fault"

Rather important to include that.

I interpret the "their fault" as an acknowledgment that "they" did not account for his oral breathing.

Anyone who wants to can read his actual post and draw their own interpretation.
My interpretation is that without seeing the data, it has not been decided that if, in fact, there is any fault that needs to be assigned.

Hence, both studies should be examed.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Fri Jun 17, 2022 3:48 pm

Rubicon wrote: ↑
Fri Jun 17, 2022 1:45 pm
... it has not been decided that if, in fact, there is any fault that needs to be assigned.
And yet, the provider has already accepted responsibility for a "fault", even without comparative data. (if Sanjay's narrative is accurate).

But you are entitled to your interpretation.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
VVV
Posts: 542
Joined: Mon Mar 14, 2011 11:28 am

Re: Sleep study inaccurate

Post by VVV » Fri Jun 17, 2022 4:20 pm

dataq1 wrote: ↑
Fri Jun 17, 2022 3:48 pm
But you are entitled to your interpretation.
I've read both threads. You are pushing an agenda and making assumptions that support it.
.....................................V

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Fri Jun 17, 2022 5:13 pm

Very easy to make an accusation without any specificity. Please enlighten . What agenda and what assumptions?

Better yet, send me a PM, Because your post has nothing to do with the topic
Last edited by dataq1 on Fri Jun 17, 2022 6:00 pm, edited 1 time in total.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Sleep study inaccurate

Post by jimbud » Fri Jun 17, 2022 5:16 pm

dataq1 wrote: ↑
Fri Jun 17, 2022 12:16 pm
First I don't understand why y'all are making such a fuss about honesty.
That is the attitude pointing out why you are what you are.

And why the responsible people on here have to watch what you say, so honesty/truth will prevail.

You are like a snotty, out of control, self absorbed, know it all fifteen year old.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Sleep study inaccurate

Post by dataq1 » Fri Jun 17, 2022 5:45 pm

jimbud wrote: ↑
Fri Jun 17, 2022 5:16 pm
And why the responsible people on here have to watch what you say, so honesty/truth will prevail.
JPB
specificity,
If you make accusations without any, it makes YOU sound like the….. well …..what you are suggesting .
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."