Sleep study inaccurate
- ChicagoGranny
- Posts: 15508
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Sleep study inaccurate
- Grouchy Bear
- Posts: 28
- Joined: Wed Jan 20, 2021 6:48 am
- Location: Deep in a gold mine somewhere in Alaska
Re: Sleep study inaccurate
So, could comparing belt movement with a previous breath and finding about the same movement, but no/very little flow through the nasal cannula could indicate either mouth breathing or an apnea? I think it gets complicated in that case, but one might be able to discern which is actually happening, with the detailed test results and graphs available to study.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Oscar on Linux; SCC; O2 Ring |
I'm RETIRED
Quit asking me to do stuff
ResMed AirSense 10 Autoset - F&P Vitera
ResMed AirSense 10 Autoset - F&P Vitera
- ChicagoGranny
- Posts: 15508
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Sleep study inaccurate
It certainly seems to be over your head and a few others commenting here.
Re: Sleep study inaccurate
While commonly referred to as “respiratory effort”, in actuality it measures the changes in chest circumference. It does not measure effort. The amount of effort to expand the chest is independent of change in dimension.
But that is not really the topic of this thread.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: Sleep study inaccurate
Well Dulltack0, someone disagrees.
https://www.google.com/search?client=fi ... dy+measure
Sleep studies screening for obstructive sleep apnea (OSA) and other sleep-related breathing disorders generally measure both respiratory flow and respiratory movements [1]. Respiratory inductance plethysmography (RIP) belts are typically used to determine the presence of respiratory effort [1].Jan 7, 2021
JPB
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Sleep study inaccurate
It appears that the key here is MATH;
whereby that which can be easily measured
can be used to estimate that which cannot be measured as easily.
Smart people do it all the time--others, not so much.
whereby that which can be easily measured
can be used to estimate that which cannot be measured as easily.
Smart people do it all the time--others, not so much.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- Grouchy Bear
- Posts: 28
- Joined: Wed Jan 20, 2021 6:48 am
- Location: Deep in a gold mine somewhere in Alaska
Re: Sleep study inaccurate
Since you imply that it is not over your head, please enlighten all who don't fully understand how the belts and breathing are related.ChicagoGranny wrote: ↑Wed Jun 15, 2022 5:44 pmIt certainly seems to be over your head and a few others commenting here.![]()
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Oscar on Linux; SCC; O2 Ring |
I'm RETIRED
Quit asking me to do stuff
ResMed AirSense 10 Autoset - F&P Vitera
ResMed AirSense 10 Autoset - F&P Vitera
Re: Sleep study inaccurate
The key there is “determine the presence of” which is not the same as “measures respiratory effort”.jimbud wrote: ↑Wed Jun 15, 2022 6:59 pm
Well Dulltack0, someone disagrees.
https://www.google.com/search?client=fi ... dy+measure
Sleep studies screening for obstructive sleep apnea (OSA) and other sleep-related breathing disorders generally measure both respiratory flow and respiratory movements [1]. Respiratory inductance plethysmography (RIP) belts are typically used to determine the presence of respiratory effort [1].Jan 7, 2021
JPB
RIP belts are based on inductance sensors, those sensors measure changes in volume (of the chest and abdomen) , not effort.
Even the dimensions (units) that RIPs display are in milliliters.
Yes it takes effort to increase the volume of the chest, but the belts don’t measure the amount of effort required to expand the chest.
Can we get back on topic? I believe something like : Are HSATs accurate if the apparatus ignores oral breathing.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- Dog Slobber
- Posts: 4330
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Sleep study inaccurate
Another example of you being tedious, adding nothing of value, and then dragging the topic off-topic. You then start posting about getting back on-topic.dataq1 wrote: ↑Wed Jun 15, 2022 5:54 pmWhile commonly referred to as “respiratory effort”, in actuality it measures the changes in chest circumference. It does not measure effort. The amount of effort to expand the chest is independent of change in dimension.
But that is not really the topic of this thread.
Posting about how RIP belts don't measure effort, but chest circumference. Is like making the claim that thermometers don't measure temperature but expansion of liquids or coils.
Here's a clue, if you want to keep topics on-topics, stop dragging them off topic.
...
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Sleep study inaccurate
Exactly. Just another example of it's P/A technique of derailing threads.Dog Slobber wrote: ↑Thu Jun 16, 2022 7:24 amAnother example of you being tedious, adding nothing of value, and then dragging the topic off-topic. You then start posting about getting back on-topic.
Laying it's down crap, thus forcing someone to correct said crap. Then trying to get everyone to move on and accept it's crap. Attempting to chalk up another win in Dulltack0's column. Tiresome.
Ignorance can be overcome...stupidity just feeds on itself.
JPB
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |
Re: Sleep study inaccurate
and @ Dogslobber,jimbud wrote: ↑Thu Jun 16, 2022 7:58 amExactly. Just another example of it's P/A technique of derailing threads.Dog Slobber wrote: ↑Thu Jun 16, 2022 7:24 amAnother example of you being tedious, adding nothing of value, and then dragging the topic off-topic. You then start posting about getting back on-topic.
Laying it's down crap, thus forcing someone to correct said crap. Then trying to get everyone to move on and accept it's crap. Attempting to chalk up another win in Dulltack0's column. Tiresome.
Ignorance can be overcome...stupidity just feeds on itself.
JPB
Are you contending that 'RIP belts measure volume, not respiratory effort' is crap? That said "crap" was intentionally injected into this discussion in an attempt to derail the topic?
Grouchy Bear introduced the "body belts" into the discussion. Granny offered a explanation as to what those belts did. Subsequently I clarified Granny's explanation. That should have been the end of that.
I don't really understand your animus towards me, but your animus/ criticism is really off topic and has nothing whatsoever to do with CPAP functions.
Do you expect me to ignore your attacks on my motives?
If you don't appreciate my comments, I invite you to either ignore them or foe list me or begin a new topic. It's all the same to me.
Now, can we get back on Sanjay's topic.......
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Sleep study inaccurate
Thanks to Dulltack0's obnoxious behavior, returning to the original topic is
NEXT TO IMPOSSIBLE.
Please go away--this is a group help forum, meant so we can help each other-
NOT SO SOME DUMBASS CAN FEED HIS EGO.
PATHETIC!
NEXT TO IMPOSSIBLE.
Please go away--this is a group help forum, meant so we can help each other-
NOT SO SOME DUMBASS CAN FEED HIS EGO.
PATHETIC!
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- Dog Slobber
- Posts: 4330
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Sleep study inaccurate
If you don't appreciate my comments, I invite you to either ignore them or foe list me or begin a new topic. It's all the same to me.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Sleep study inaccurate
I'd like to go back to the part where OP claims there are 2 belts, cause there should only be one:
https://document.resmed.com/documents/p ... lo_eng.pdf
I don't think it's RIP technology either.
And unless a RIP belt is calibrated (cRIP) it doesn't measure volume, it measures motion.
https://document.resmed.com/documents/p ... lo_eng.pdf
I don't think it's RIP technology either.
And unless a RIP belt is calibrated (cRIP) it doesn't measure volume, it measures motion.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: Sleep study inaccurate
And even if you did calibrate it, it would only be accurate until the guy decided to roll over.
That said, uncalibrated RIP belts are far better than piezo because the signal is superior.