OSCAR data display question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
unknown12345
Posts: 25
Joined: Tue Feb 03, 2015 6:03 pm

OSCAR data display question

Post by unknown12345 » Mon Mar 09, 2020 5:48 pm

Hi,

Pls see below graph from OSCAR for my AirCurve 10 ASV settings:
AirCurve_10_Settings_pic1.JPG
AirCurve_10_Settings_pic1.JPG (73.81 KiB) Viewed 391 times
Which means pressure ranges would be : (8.2 <= EPAP <= 15) and (8.2 <= IPAP <= 23) ... correct?
If so, then why in the "Statistics" section it shows "Max EPAP = 18.0" ?? (red underlined)

Below pic is the pressure blotter for the same night (3/2/2020), as you can see max EPAP (green line in the graph) never went above "15.0"
AirCurve_10_Settings_pic2.JPG
AirCurve_10_Settings_pic2.JPG (80.93 KiB) Viewed 391 times
Is there a bug in the OSCAR?

Thx.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine Mode: ASVAuto, (Min EPAP = 8.2), (Max EPAP = 15.0), (Min PS = 0.0), (Max PS = 8.0)

unknown12345
Posts: 25
Joined: Tue Feb 03, 2015 6:03 pm

Re: OSCAR data display question

Post by unknown12345 » Mon Mar 09, 2020 5:53 pm

Forgot to include OSCAR version, here it is:
AirCurve_10_Settings_OSCAR_Version.JPG
AirCurve_10_Settings_OSCAR_Version.JPG (16.1 KiB) Viewed 389 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine Mode: ASVAuto, (Min EPAP = 8.2), (Max EPAP = 15.0), (Min PS = 0.0), (Max PS = 8.0)

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: OSCAR data display question

Post by Pugsy » Mon Mar 09, 2020 6:01 pm

OSCAR has more than one little bug. :lol: :lol:
Even more bugs when people are using the various ASV kind of machines.
You found one of them...just ignore it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR data display question

Post by zonker » Mon Mar 09, 2020 6:08 pm

unknown12345 wrote:
Mon Mar 09, 2020 5:48 pm
Hi,

Pls see below graph from OSCAR for my AirCurve 10 ASV settings:
AirCurve_10_Settings_pic1.JPG

Which means pressure ranges would be : (8.2 <= EPAP <= 15) and (8.2 <= IPAP <= 23) ... correct?
If so, then why in the "Statistics" section it shows "Max EPAP = 18.0" ?? (red underlined)

Below pic is the pressure blotter for the same night (3/2/2020), as you can see max EPAP (green line in the graph) never went above "15.0"
AirCurve_10_Settings_pic2.JPG
Is there a bug in the OSCAR?

Thx.
you may want to report this in the first thread-

viewtopic/t175269/OSCAR-Suggestions-amp ... rting.html
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: OSCAR data display question

Post by Pugsy » Mon Mar 09, 2020 6:10 pm

It's already a known bug....no need to report it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR data display question

Post by zonker » Mon Mar 09, 2020 6:11 pm

Pugsy wrote:
Mon Mar 09, 2020 6:10 pm
It's already a known bug....no need to report it.
which is prolly why you didn't suggest it! :lol:

again, HOW do you keep all this "stuff" in your head?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

unknown12345
Posts: 25
Joined: Tue Feb 03, 2015 6:03 pm

Re: OSCAR data display question

Post by unknown12345 » Wed Mar 11, 2020 1:22 pm

Pugsy wrote:
Mon Mar 09, 2020 6:01 pm
OSCAR has more than one little bug. :lol: :lol:
Even more bugs when people are using the various ASV kind of machines.
You found one of them...just ignore it.
Curiosity question : From the clinician manual I am deducing that pressure(s) are determined as follows:

(Min EPAP <= EPAP <= Max EPAP) and
[ (Min EPAP <= IPAP <= (Max EPAP + Max PS) ]

correct ?

If so, then how is "Min PS" used by the algorithm?

Thx.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine Mode: ASVAuto, (Min EPAP = 8.2), (Max EPAP = 15.0), (Min PS = 0.0), (Max PS = 8.0)

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: OSCAR data display question

Post by chunkyfrog » Wed Mar 11, 2020 1:27 pm

Some heads have room for stuff that other heads disregard.

It's just how we are wired.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

unknown12345
Posts: 25
Joined: Tue Feb 03, 2015 6:03 pm

Re: OSCAR data display question

Post by unknown12345 » Wed Mar 11, 2020 1:34 pm

unknown12345 wrote:
Wed Mar 11, 2020 1:22 pm
Pugsy wrote:
Mon Mar 09, 2020 6:01 pm
OSCAR has more than one little bug. :lol: :lol:
Even more bugs when people are using the various ASV kind of machines.
You found one of them...just ignore it.
Curiosity question : From the clinician manual I am deducing that pressure(s) are determined as follows:

(Min EPAP <= EPAP <= Max EPAP) and
[ (Min EPAP <= IPAP <= (Max EPAP + Max PS) ]

correct ?

If so, then how is "Min PS" used by the algorithm?

Thx.

Forgot to include pic from manual, see below:
AirCurve_10_Pressure_Range.JPG
AirCurve_10_Pressure_Range.JPG (76.66 KiB) Viewed 326 times

Based on the above pic, then may be I got the IPAP formula wrong!!

If so, is IPAP range calculated as follows?? (note the addition of "Min PS" in red)

[ (Min EPAP + Min PS) <= IPAP <= (Max EPAP + Max PS) ]

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nano Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine Mode: ASVAuto, (Min EPAP = 8.2), (Max EPAP = 15.0), (Min PS = 0.0), (Max PS = 8.0)

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: OSCAR data display question

Post by Pugsy » Wed Mar 11, 2020 2:21 pm

unknown12345 wrote:
Wed Mar 11, 2020 1:22 pm
Pugsy wrote:
Mon Mar 09, 2020 6:01 pm
OSCAR has more than one little bug. :lol: :lol:
Even more bugs when people are using the various ASV kind of machines.
You found one of them...just ignore it.
Curiosity question : From the clinician manual I am deducing that pressure(s) are determined as follows:

(Min EPAP <= EPAP <= Max EPAP) and
[ (Min EPAP <= IPAP <= (Max EPAP + Max PS) ]

correct ?

If so, then how is "Min PS" used by the algorithm?

Thx.
In what context? Minimum PS is nothing more than the minimum difference between inhale and exhale.
What I have always wondered about (and couldn't ever find out anything) was what in the algorithm tells the machine to vary the PS or increase it from the minimum when a variable PS is available...and why and how much.
I know big PS is needed to deal with central apneas...that rather large burst of PS so that the difference between inhale and exhale is substantial....what about minor differences? I never could find any real documentation as to what, how or how much drives the algorithm to do its job.