I have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
Is IT Carbon Dioxide?
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Is IT Carbon Dioxide?
sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:13 pmI have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
It would be really good to have graphs and breakdowns on what the type and number of events your having, turning epr off could have reduced CA’s or reduced OA’s as it increases epap by one we just don’t kno as we are guessing that’s why detailed information of the screen or OSCAR is so useful
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
been told my machine wont produce that info.. working on getting new machine but making use of wht I have nowJas_williams wrote: ↑Fri Oct 25, 2019 2:21 pmsidesleeper7 wrote: ↑Fri Oct 25, 2019 2:13 pmI have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
It would be really good to have graphs and breakdowns on what the type and number of events your having, turning epr off could have reduced CA’s or reduced OA’s as it increases epap by one we just don’t kno as we are guessing that’s why detailed information of the screen or OSCAR is so useful
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
Re: Is IT Carbon Dioxide?
Those .2 adjustments are typically inconsequential.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:13 pmI have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
Turning EPR off, however, is effectively raising your pressure by 1cm.
So, a pressure of 7.8 EPR 1 gives you an *effective* pressure of 6.8.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
gotcha and it seemed that effective pressure of 6.8 was a little too low as AHI were higher then .. even though it seemed a little more comfortable exhaling...palerider wrote: ↑Fri Oct 25, 2019 2:48 pmThose .2 adjustments are typically inconsequential.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:13 pmI have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
Turning EPR off, however, is effectively raising your pressure by 1cm.
So, a pressure of 7.8 EPR 1 gives you an *effective* pressure of 6.8.
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
Re: Is IT Carbon Dioxide?
Six pages and no charts... let's see some charts.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:52 pmgotcha and it seemed that effective pressure of 6.8 was a little too low as AHI were higher then .. even though it seemed a little more comfortable exhaling...![]()
wiki/index.php/Sleepyhead:Organize
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
been told my machine wont produce that info.. working on getting new machine but making use of wht I have nowpalerider wrote: ↑Fri Oct 25, 2019 2:58 pmSix pages and no charts... let's see some charts.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:52 pmgotcha and it seemed that effective pressure of 6.8 was a little too low as AHI were higher then .. even though it seemed a little more comfortable exhaling...![]()
wiki/index.php/Sleepyhead:Organize
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
Re: Is IT Carbon Dioxide?
I think (per profile) that he has the half assed brick CPAP that only shows the summary AHI, pressure and leak numbers.palerider wrote: ↑Fri Oct 25, 2019 2:58 pmSix pages and no charts... let's see some charts.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:52 pmgotcha and it seemed that effective pressure of 6.8 was a little too low as AHI were higher then .. even though it seemed a little more comfortable exhaling...![]()
wiki/index.php/Sleepyhead:Organize
No graphs and no event category breakdown. Nothing to really see with the software except for Bob to tell us he's got a brick.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Is IT Carbon Dioxide?
it's my humble opinion that you stay with your settings as is and let it settle down. and maybe raise the min pressure another .2 in a few nights.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
one night's data isn't much. so let's see how that goes.
i had to do similarly, raise and wait .2 at a time. for me, it was the added battle of aerophagia
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Is IT Carbon Dioxide?
Oh, right, sorry, my bad... raise the pressure another CM and see what happens.sidesleeper7 wrote: ↑Fri Oct 25, 2019 3:30 pmbeen told my machine wont produce that info.. working on getting new machine but making use of wht I have nowpalerider wrote: ↑Fri Oct 25, 2019 2:58 pmSix pages and no charts... let's see some charts.sidesleeper7 wrote: ↑Fri Oct 25, 2019 2:52 pmgotcha and it seemed that effective pressure of 6.8 was a little too low as AHI were higher then .. even though it seemed a little more comfortable exhaling...![]()
wiki/index.php/Sleepyhead:Organize
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
thanks going to give this a try .. btw what do you mean by "settle down" ? I usually only hear this from my Mom when Ive had too much too drinkzonker wrote: ↑Fri Oct 25, 2019 5:53 pmit's my humble opinion that you stay with your settings as is and let it settle down. and maybe raise the min pressure another .2 in a few nights.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
one night's data isn't much. so let's see how that goes.
i had to do similarly, raise and wait .2 at a time. for me, it was the added battle of aerophagiaas well as getting my ahi down. but slow (and i mean SLOW) and steady won the race for me.
good luck!
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
Re: Is IT Carbon Dioxide?
where do you think she got it from?sidesleeper7 wrote: ↑Fri Oct 25, 2019 6:31 pmthanks going to give this a try .. btw what do you mean by "settle down" ? I usually only hear this from my Mom when Ive had too much too drinkzonker wrote: ↑Fri Oct 25, 2019 5:53 pmit's my humble opinion that you stay with your settings as is and let it settle down. and maybe raise the min pressure another .2 in a few nights.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
one night's data isn't much. so let's see how that goes.
i had to do similarly, raise and wait .2 at a time. for me, it was the added battle of aerophagiaas well as getting my ahi down. but slow (and i mean SLOW) and steady won the race for me.
good luck!![]()
![]()
![]()
i mean give your body and brain time to get comfortable with what you are asking it to do. some here will press you to go faster. some say go slower.
i'm just giving you my take on it from my perspective.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Is IT Carbon Dioxide?
Sorry I’d missed you have the CPAP version of the machine bummer my first machine was an S9 escape which doesn’t even have AHI it too me three days to buy my own better machinesidesleeper7 wrote: ↑Fri Oct 25, 2019 2:42 pmbeen told my machine wont produce that info.. working on getting new machine but making use of wht I have nowJas_williams wrote: ↑Fri Oct 25, 2019 2:21 pmsidesleeper7 wrote: ↑Fri Oct 25, 2019 2:13 pmI have just adjusted the pressure from 7.8 with EPR @ 1 and now I am at 7.4-7.6 with EPR Turned OFF - my AHI was higher when I had the EPR ON.palerider wrote: ↑Fri Oct 25, 2019 12:44 pmWhich settings did you mess with most recently, and when? if your AHI is slowly settling down, then let it keep doing that.sidesleeper7 wrote: ↑Fri Oct 25, 2019 12:10 pmGot the AHI down to 3 AHI lastnight, feeling better today... not sure if I should mess with any settings further or just stick to this path for now..![]()
It would be really good to have graphs and breakdowns on what the type and number of events your having, turning epr off could have reduced CA’s or reduced OA’s as it increases epap by one we just don’t kno as we are guessing that’s why detailed information of the screen or OSCAR is so useful![]()
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
I am looking for a Resmed N10 Large Cushion on Amazon, not having any luck would a Nano cushion work with this nasal mask ? Thx!Pugsy wrote: ↑Tue Oct 22, 2019 8:16 pmhttps://5faacd0c97cf0d23c6dc-064f3cd61b ... -guide.pdfsidesleeper7 wrote: ↑Tue Oct 22, 2019 8:07 pmI have the AirFit N10 Nasal Mask (regular size) and it seems it may not be wide enough and blocks some airflow, is there a fitting guide for which sizes to use ?
Same guide as for the Nano. I suppose because the actual cushion itself is similar.
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask
-
sidesleeper7
- Posts: 129
- Joined: Fri Jan 25, 2013 12:36 pm
Re: Is IT Carbon Dioxide?
I am looking for a Resmed N10 Large Cushion on Amazon, not having any luck would a Nano cushion work with this nasal mask ? Thx!
ResMed AirSense 10 Auto - Resmed Airfit N30 Mask