NEWBIE HERE TOO!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Gryhnd
Posts: 42
Joined: Sun Jul 14, 2019 8:40 am

NEWBIE HERE TOO!

Post by Gryhnd » Tue Jul 16, 2019 4:03 pm

Am glad I found this forum! It has answered many of my questions. My biggest concern now is the leak rate. I have read all the info here & called my DME today to try a different mask. Currently using P10 nasal pillows. Last night my leak rate was 40L. Sunday night it was14L. Both nights events were both 0.2/hr. I am a side sleeper and not a mouth breather. My pressures are 8/10.2 (I changed it from 11.0 to 10.0, and slowly working my way back up. I have installed OSCAR, but DEM gave me the protective sleeve, but no SD! They are sending one, should get it Thursday. I'm going to try using the Resmed AirFit P0i nasal pillow as am clausterphobic (sp?) and a full face mask would drive me crazy.

User avatar
Grace~~~
Posts: 662
Joined: Tue May 10, 2016 3:27 pm
Location: Florida

Re: NEWBIE HERE TOO!

Post by Grace~~~ » Tue Jul 16, 2019 6:42 pm

<waves at gryhnd>

Hi Gryhnd ~~~

Zonker should be along soon to open the cage door and tell you how to add your machine and mask to the bottom of your posts. (Folks here like to just see that and not have to ask you all the time.)

...and no doubt one of the smart people will be able to ask you a few diagnostic questions.


Have you figured out from reading what OSCAR is yet?

I found this "Welcome cpap grayhound" but just realized you *could?* be grayHAND? :shock:

I'm hoping for hound. Not sure what a gray hand might mean? Low oxygen?



Image
Began XPAP May 2016. Autoset Pressure min. 8 / max 15. Ramp off. ERP set at 2. No humidity. Sleepyhead software installed and being looked at daily, though only beginning to understand the data.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: NEWBIE HERE TOO!

Post by palerider » Tue Jul 16, 2019 6:55 pm

Gryhnd wrote:
Tue Jul 16, 2019 4:03 pm
Am glad I found this forum! It has answered many of my questions. My biggest concern now is the leak rate. I have read all the info here & called my DME today to try a different mask. Currently using P10 nasal pillows. Last night my leak rate was 40L. Sunday night it was14L. Both nights events were both 0.2/hr. I am a side sleeper and not a mouth breather. My pressures are 8/10.2 (I changed it from 11.0 to 10.0, and slowly working my way back up. I have installed OSCAR, but DEM gave me the protective sleeve, but no SD! They are sending one, should get it Thursday. I'm going to try using the Resmed AirFit P0i nasal pillow as am clausterphobic (sp?) and a full face mask would drive me crazy.
Without knowing what kind of machine you have... (READ the "READ THIS FIRST" page and fill in your equipment profile), we can't even guess whether your leak rate of 40lpm is excessive, or ok. depends on the machine.

Ok, I could *guess* based on the 10.2 pressure that you're on a Resmed machine... in which case your mask is leaking too much at 40lpm, and fine at 14.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Tue Jul 16, 2019 7:19 pm

Grace~~~ wrote:
Tue Jul 16, 2019 6:42 pm


Zonker should be along soon to open the cage door and tell you how to add your machine and mask to the bottom of your posts. (Folks here like to just see that and not have to ask you all the time.)

...and no doubt one of the smart people will be able to ask you a few diagnostic questions.
meerkat.gif

are you saying i'm NOT one of the smart ones?
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Tue Jul 16, 2019 7:22 pm

palerider wrote:
Tue Jul 16, 2019 6:55 pm

Ok, I could *guess* based on the 10.2 pressure that you're on a Resmed machine... in which case your mask is leaking too much at 40lpm, and fine at 14.
oh, well spotted, sir!

i didn't even notice the ".2".
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Tue Jul 16, 2019 7:27 pm

Gryhnd wrote:
Tue Jul 16, 2019 4:03 pm
Am glad I found this forum!
good to have you here. and now that grace has outed me, welcome to the zoo.

please see my signature for some helpful links. we are each of us battling this apnea thing with different equipment. so if you can follow the directions in the newbie link below, it will help us help you.

it really makes a difference which machine you have, as palerider has pointed out.

btw, he's one of the SMART PEOPLE here. :lol:

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Gryhnd
Posts: 42
Joined: Sun Jul 14, 2019 8:40 am

Re: NEWBIE HERE TOO!

Post by Gryhnd » Tue Jul 16, 2019 7:46 pm

I have updated the equipment as requested. It is gryhnd...greyhound...I have several! I will download my reports from OSCAR when I get my SD card from DME...hopefully by Thursday.!

Janknitz
Posts: 8517
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: NEWBIE HERE TOO!

Post by Janknitz » Tue Jul 16, 2019 8:08 pm

Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested. It is gryhnd...greyhound...I have several! I will download my reports from OSCAR when I get my SD card from DME...hopefully by Thursday.!
Welcome and totally off topic, but we have adopted a lab/greyhound mix. We've always had lab mixes but never a greyhound and we are finding him VERY challenging. I shall have to PM you with questions about greyhounds, we are clueless.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

Gryhnd
Posts: 42
Joined: Sun Jul 14, 2019 8:40 am

Re: NEWBIE HERE TOO!

Post by Gryhnd » Tue Jul 16, 2019 8:16 pm

yes, Janknitz...PM me. I have been with greyhound rescue for 24 years & was a veterinarian technician!!! :D

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: NEWBIE HERE TOO!

Post by palerider » Tue Jul 16, 2019 8:31 pm

Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested.
You do have the wrong humidifier, the one you have listed is from a wholly different manufacturer, just leave that field blank.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Tue Jul 16, 2019 9:59 pm

palerider wrote:
Tue Jul 16, 2019 8:31 pm
Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested.
You do have the wrong humidifier, the one you have listed is from a wholly different manufacturer, just leave that field blank.
dayum! nothing wrong with your eyesight today.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

SDBud
Posts: 266
Joined: Thu Jan 03, 2019 3:24 pm
Location: Sunny SoCal

Re: NEWBIE HERE TOO!

Post by SDBud » Wed Jul 17, 2019 12:07 am

zonker wrote:
Tue Jul 16, 2019 9:59 pm
palerider wrote:
Tue Jul 16, 2019 8:31 pm
Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested.
You do have the wrong humidifier, the one you have listed is from a wholly different manufacturer, just leave that field blank.
dayum! nothing wrong with your eyesight today.
EYESIGHT?? We're NOT here to discuss eyesight, but BREATHING. geeze.
Airsense 10 Autoset
AirTouch and AirFit F10 masks
:roll:

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Wed Jul 17, 2019 11:50 am

SDBud wrote:
Wed Jul 17, 2019 12:07 am
zonker wrote:
Tue Jul 16, 2019 9:59 pm
palerider wrote:
Tue Jul 16, 2019 8:31 pm
Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested.
You do have the wrong humidifier, the one you have listed is from a wholly different manufacturer, just leave that field blank.
dayum! nothing wrong with your eyesight today.
EYESIGHT?? We're NOT here to discuss eyesight, but BREATHING. geeze.
wrong forum! thought i was at eyeweartalk.com.

sorry!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

SDBud
Posts: 266
Joined: Thu Jan 03, 2019 3:24 pm
Location: Sunny SoCal

Re: NEWBIE HERE TOO!

Post by SDBud » Wed Jul 17, 2019 6:45 pm

zonker wrote:
Wed Jul 17, 2019 11:50 am
SDBud wrote:
Wed Jul 17, 2019 12:07 am
zonker wrote:
Tue Jul 16, 2019 9:59 pm
palerider wrote:
Tue Jul 16, 2019 8:31 pm
Gryhnd wrote:
Tue Jul 16, 2019 7:46 pm
I have updated the equipment as requested.
You do have the wrong humidifier, the one you have listed is from a wholly different manufacturer, just leave that field blank.
dayum! nothing wrong with your eyesight today.
EYESIGHT?? We're NOT here to discuss eyesight, but BREATHING. geeze.
wrong forum! thought i was at eyeweartalk.com.

sorry!!
Leave an adequate donation in the poor box as you leave,
and we MIGHT forgive you.
Airsense 10 Autoset
AirTouch and AirFit F10 masks
:roll:

User avatar
zonker
Posts: 11416
Joined: Fri Jun 19, 2015 4:36 pm

Re: NEWBIE HERE TOO!

Post by zonker » Wed Jul 17, 2019 7:07 pm

SDBud wrote:
Wed Jul 17, 2019 6:45 pm
We're NOT here to discuss eyesight, but BREATHING. geeze.
wrong forum! thought i was at eyeweartalk.com.

sorry!!
[/quote]

Leave an adequate donation in the poor box as you leave,
and we MIGHT forgive you.
[/quote]

squint.gif
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg