CPAP Help: Feel suffocation.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
coldsideofpillow
Posts: 29
Joined: Wed Feb 06, 2019 12:57 pm

Re: CPAP Help: Feel suffocation.

Post by coldsideofpillow » Fri Feb 08, 2019 12:51 pm

so this has been my progress so far, making the change to pressure of 8 and EPR at 3. i attached the pictures.

last night thursday was my best night as far as using it. i did the same thing i put my nose strip on to open my nostrils and used some zicam decogestant and put the cpap on and fell asleep with my mouth closed. im trying to remember if at any point i took off the mask but i cant remember but i feel like i might have...i was in such deep sleep i was actually dreaming. i did wake up twice and the first time i was sleeping on my side with the cpap still on but i was drooling. i was a bit disappointed because i dont know how long i was like that :( .

after that happened i was determined to not let that make me want to stop so i laid flat on my back and woke up for my alarm at 8 am. i was on my back with my mouth closed.

i feel good. its certainly easier to get out of bed and i dont need as much coffee to wake up. i guess i gotta give it a couple of weeks till i feel those life changing effects. thanks alot yall i would have not got here without you guys. just a few questions.

with my mouth open during cpap, is mouth taping suggested? does keeping your mouth open during cpap sleep affect how effective the cpap treatment is? ive tried taping before but it gives me that asphyxiation feeling when im breathing out. if i have to learn to breathe through the cpap with tape over my mouth then i guess i have to. thoughts?
Attachments
E37B07A1-93D0-485F-914C-066387C419D2.jpeg
E37B07A1-93D0-485F-914C-066387C419D2.jpeg (884.02 KiB) Viewed 5162 times
CE56B6B7-2F65-40AB-8323-A3E74D30E930.jpeg
CE56B6B7-2F65-40AB-8323-A3E74D30E930.jpeg (872.89 KiB) Viewed 5162 times
F5287DEF-30E2-4429-B78B-CE1D1CC515D2.jpeg
F5287DEF-30E2-4429-B78B-CE1D1CC515D2.jpeg (884.73 KiB) Viewed 5162 times

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP Help: Feel suffocation.

Post by Pugsy » Fri Feb 08, 2019 1:08 pm

Mouth opening or mouth breathing doesn't always mean that there is a massive loss of therapy pressure out the mouth.

I have a leak graph where I was awake and mouth breathing for sure....because I woke up mouth breathing (gently...no massive tornado out the mouth) and I figured it would be a good time to document just how much mouth breathing might actually impact the leak graph.
The known mouth breathing time is circled in red at the end of the night. I would have done it longer but my bladder wouldn't take much more stretching.
It's certainly well within the machine's ability to compensate for leak and actually well below some other bigger leaks earlier in the night.
I don't know what caused those...probably mask moving around with tossing and turning...and I don't really care. I slept through all of it and that's all I care about. I never worry about minor short lived excursions into large leak territory as long as the leaks don't wake me up.

If taping to get a "perfect" leak line causes more sleep disruption than any little leaks might....don't tape. Let a little leak slide.
You gotta get the sleep first....without the sleep the rest of this stuff doesn't matter much does it? A little drooling...is it a big deal...nope...I do it all the time. Annoying yes...big deal as to impacting therapy...I don't see it happening. Drooling can happen with just a tiny bit of opening in the lips...doesn't even have to cause any mouth breathing at all.



Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

coldsideofpillow
Posts: 29
Joined: Wed Feb 06, 2019 12:57 pm

Re: CPAP Help: Feel suffocation.

Post by coldsideofpillow » Fri Feb 08, 2019 2:04 pm

thanks a lot for the reply. i find it weird that i even got to the point of drooling with the mask on but im unsure if i even did any mouth breathing. but i will never know until i get one of those APAP machines.

do you think a few more weeks of this and i will see some progress in my energy and brain fog and mood?

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP Help: Feel suffocation.

Post by Pugsy » Fri Feb 08, 2019 2:16 pm

coldsideofpillow wrote:
Fri Feb 08, 2019 2:04 pm
do you think a few more weeks of this and i will see some progress in my energy and brain fog and mood?
I wish I could give you a definitive answer but there's so much more to getting good quality restorative sleep than just dealing with sleep apnea events all by themselves.

I will say this though...until you can consistently get a decent number of hours of sleep with minimal awakenings for whatever reason...you don't stand much chance of feeling much improvement.
And there's so much more than just sleep apnea stuff that can impact our sleep quality.
It's not only hours of sleep but the quality of those hours and then there are the other things that can impact sleep quality and how we feel....like medication side effects or other health issues or whatever.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Help: Feel suffocation.

Post by zonker » Fri Feb 08, 2019 4:48 pm

coldsideofpillow wrote:
Fri Feb 08, 2019 12:51 pm
so this has been my progress so far, making the change to pressure of 8 and EPR at 3. i attached the pictures.
<snip>

with my mouth open during cpap, is mouth taping suggested? does keeping your mouth open during cpap sleep affect how effective the cpap treatment is? ive tried taping before but it gives me that asphyxiation feeling when im breathing out. if i have to learn to breathe through the cpap with tape over my mouth then i guess i have to. thoughts?
keep on keepin' on, as we used to say in the 70's. :lol:

instead of taping, which i hate because of my beard, you may want to try this-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

it works GREAT for me. maybe it can help you?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

coldsideofpillow
Posts: 29
Joined: Wed Feb 06, 2019 12:57 pm

Re: CPAP Help: Feel suffocation.

Post by coldsideofpillow » Fri Feb 08, 2019 5:55 pm

Pugsy wrote:
Fri Feb 08, 2019 2:16 pm
coldsideofpillow wrote:
Fri Feb 08, 2019 2:04 pm
do you think a few more weeks of this and i will see some progress in my energy and brain fog and mood?
I wish I could give you a definitive answer but there's so much more to getting good quality restorative sleep than just dealing with sleep apnea events all by themselves.

I will say this though...until you can consistently get a decent number of hours of sleep with minimal awakenings for whatever reason...you don't stand much chance of feeling much improvement.
And there's so much more than just sleep apnea stuff that can impact our sleep quality.
It's not only hours of sleep but the quality of those hours and then there are the other things that can impact sleep quality and how we feel....like medication side effects or other health issues or whatever.
i understand. it was silly of me to ask this with barely my first night of “decent” sleep on cpap. thanks!

coldsideofpillow
Posts: 29
Joined: Wed Feb 06, 2019 12:57 pm

Re: CPAP Help: Feel suffocation.

Post by coldsideofpillow » Fri Feb 08, 2019 7:47 pm

zonker wrote:
Fri Feb 08, 2019 4:48 pm
coldsideofpillow wrote:
Fri Feb 08, 2019 12:51 pm
so this has been my progress so far, making the change to pressure of 8 and EPR at 3. i attached the pictures.
<snip>

with my mouth open during cpap, is mouth taping suggested? does keeping your mouth open during cpap sleep affect how effective the cpap treatment is? ive tried taping before but it gives me that asphyxiation feeling when im breathing out. if i have to learn to breathe through the cpap with tape over my mouth then i guess i have to. thoughts?
keep on keepin' on, as we used to say in the 70's. :lol:

instead of taping, which i hate because of my beard, you may want to try this-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

it works GREAT for me. maybe it can help you?
is that the right link? not sure im following here...

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP Help: Feel suffocation.

Post by Pugsy » Fri Feb 08, 2019 8:01 pm

That's probably not the link Zonker was thinking of...he'll be along in a bit and correct it.
I don't have the one he is thinking about handy or I would give it to you.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: CPAP Help: Feel suffocation.

Post by Okie bipap » Fri Feb 08, 2019 8:29 pm

I think this is the post he was looking for viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Help: Feel suffocation.

Post by zonker » Fri Feb 08, 2019 10:23 pm

Okie bipap wrote:
Fri Feb 08, 2019 8:29 pm
I think this is the post he was looking for viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296
THAT'S it. thanks for saving my bacon, okie!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Help: Feel suffocation.

Post by zonker » Fri Feb 08, 2019 10:24 pm

coldsideofpillow wrote:
Fri Feb 08, 2019 7:47 pm

is that the right link? not sure im following here...
that was merely a test to see if you were paying attention! :lol:

okie bipap has the correct link.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

coldsideofpillow
Posts: 29
Joined: Wed Feb 06, 2019 12:57 pm

Re: CPAP Help: Feel suffocation.

Post by coldsideofpillow » Sat Feb 16, 2019 10:06 am

hello guys ive been using my machine for over a week now and i have been using it fully for 6-8 hours these past 5 days at pressure of 8 EPR of 3. Thanks to you guys Ive been sleeping much better and im greatful for this message board.

Im wondering now where do i go from here now? Do i increase pressure? As far as my sleep quality its improved but i wonder if it gets better?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CPAP Help: Feel suffocation.

Post by palerider » Sat Feb 16, 2019 12:07 pm

coldsideofpillow wrote:
Sat Feb 16, 2019 10:06 am
Im wondering now where do i go from here now? Do i increase pressure? As far as my sleep quality its improved but i wonder if it gets better?
Ya know how we talk about posting charts? well, this is one of those times.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

klv329
Posts: 500
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: CPAP Help: Feel suffocation.

Post by klv329 » Wed Feb 20, 2019 9:38 pm

after years if taping, I switched to a 3 inch soft cervical collar that is working fine. a stiffer 4 inch collar was too uncomfortable. just got tired of taping, I guess. lots of collars on amazon.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: CPAP Help: Feel suffocation.

Post by Chris8243 » Wed Feb 20, 2019 11:13 pm

coldsideofpillow wrote:
Wed Feb 06, 2019 8:10 pm
Do you think I would benefit from using a mouth and nose mask even though im a mouth breather? Like if i was do get a mask with mouth and nose would CPAP still be effective even if im breathing through my mouth on cpap?
Yes! I use Amara View (click on it on my post so you can see it). It covers the mouth and under the nose. I don't worry about or even think about trying to keep my mouth closed. I don't keep my mouth closed naturally when I'm awake so I know I'm a mouth breather.