Me and My AirSense Now with screenshots!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65139
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Me and My AirSense Now with screenshots!

Post by Pugsy » Sun Aug 19, 2018 4:34 am

palerider wrote:
Sat Aug 18, 2018 7:47 pm
well, If the machine is spiking the pressure, then, probably, no matter what the graph looks like on sleepyhead, it's worth investigating, and if it looks horrible on SH and the machine is going "pfft, don't care".... ;)
I am not saying just immediately dismiss ugly FL graphs. All I am saying is that if the AHI is nice and low and boring...and I am sleeping good and feeling good....that I wouldn't go spending enormous time worrying and doing all sorts of pressure increases (or letting the machine go higher if going higher causes problems) just to clean up a FL graph that might not clean up anyway.

Investigate....yes and maybe even try some more pressure experiments but if the FL graph is the only thing ugly and the person is sleeping good and feeling good...and the more pressure experiments wasn't cleaning up the FL graph then maybe back up a bit and not go to war with the FL graph if more pressure doesn't clean it up and there's nothing else we want to fight.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 7:57 am

BlueDragon wrote:
Sat Aug 18, 2018 9:47 pm
fyi, I have run my machine in "For Her" mode once to see what the obvious differences were. The two major differences I noted were:
  • The way For Her mode established a new minimum pressure after it finds it necessary to increase the pressure. The pressure line never returns to the minimum pressure.
  • The pressure increases--presumably for sleeping on my back--are gentler. (Gentler than APAP "soft response" mode too.)
The Flow Limit chart looks quite different but I didn't pay much attention to it.
Hey thanks for doing this. Since you have resscan, would you be willing to update the post with a screenshot of the "for her" FL chart in from resscan? After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Me and My AirSense Now with screenshots!

Post by palerider » Sun Aug 19, 2018 8:21 am

djams wrote:
Sun Aug 19, 2018 7:57 am
After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
The real question isn't trying to get resscan to imitate SH, the question is, if Resscan is set to defaults, what does the data actually look like, and is SH broken, (as it is in so many other edge cases) (see the post I just made where it's showing a respiration rate of 50 during an apnea viewtopic.php?f=1&t=172686&p=1265275#p1265275)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 8:46 am

palerider wrote:
Sun Aug 19, 2018 8:21 am
djams wrote:
Sun Aug 19, 2018 7:57 am
After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
The real question isn't trying to get resscan to imitate SH, the question is, if Resscan is set to defaults, what does the data actually look like, and is SH broken, (as it is in so many other edge cases) (see the post I just made where it's showing a respiration rate of 50 during an apnea viewtopic.php?f=1&t=172686&p=1265275#p1265275)
Understand your point. My thinking was to get everything same as it can be, and just compare the chart shapes. My gut feeling is that SH is the problem too.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 9:17 am

And wow, that's a wonderful thing you did in the linked thread. You're a helluva guy!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Me and My AirSense Now with screenshots!

Post by palerider » Sun Aug 19, 2018 9:20 am

djams wrote:
Sun Aug 19, 2018 8:46 am
palerider wrote:
Sun Aug 19, 2018 8:21 am
djams wrote:
Sun Aug 19, 2018 7:57 am
After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
The real question isn't trying to get resscan to imitate SH, the question is, if Resscan is set to defaults, what does the data actually look like, and is SH broken, (as it is in so many other edge cases) (see the post I just made where it's showing a respiration rate of 50 during an apnea viewtopic.php?f=1&t=172686&p=1265275#p1265275)
Understand your point. My thinking was to get everything same as it can be, and just compare the chart shapes. My gut feeling is that SH is the problem too.
The chart *shapes* are fine. SH is recording the same FL events as Resscan... it's just the scale of the y axis that's different.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 9:39 am

palerider wrote:
Sun Aug 19, 2018 9:20 am
djams wrote:
Sun Aug 19, 2018 8:46 am
palerider wrote:
Sun Aug 19, 2018 8:21 am
djams wrote:
Sun Aug 19, 2018 7:57 am
After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
The real question isn't trying to get resscan to imitate SH, the question is, if Resscan is set to defaults, what does the data actually look like, and is SH broken, (as it is in so many other edge cases) (see the post I just made where it's showing a respiration rate of 50 during an apnea viewtopic.php?f=1&t=172686&p=1265275#p1265275)
Understand your point. My thinking was to get everything same as it can be, and just compare the chart shapes. My gut feeling is that SH is the problem too.
The chart *shapes* are fine. SH is recording the same FL events as Resscan... it's just the scale of the y axis that's different.
That's interesting. To me it appears that they are being overly widened. This is a gigantic assumption on my part.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
zonker
Posts: 11364
Joined: Fri Jun 19, 2015 4:36 pm

Re: Me and My AirSense Now with screenshots!

Post by zonker » Sun Aug 19, 2018 11:08 am

i'm going to go ahead and post that "odd" night, just for the record-
screenshot-20180819-094256.png
then last night-
screenshot-20180819-094256.png
obviously, i'm pretty happy with last night! :) now, i know one data point doesn't make a trend. so we'll see how it goes, moving forward.

i'll have many questions if my AHI stays nice and low. of course, i'll have some question if it doesn't, like WHY? :) but we'll see how she goes.

comments welcome if you see anything you wish to comment upon.

interesting....i thought i'd formatted the post so that i had image then comment then image. that's not how the preview turned out. and for some reason, the high AHI night got in 2x.

:roll: more "stuff" for me to learn!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
BlueDragon
Posts: 547
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Me and My AirSense Now with screenshots!

Post by BlueDragon » Sun Aug 19, 2018 12:21 pm

djams wrote:
Sun Aug 19, 2018 7:57 am
Hey thanks for doing this. Since you have resscan, would you be willing to update the post with a screenshot of the "for her" FL chart in from resscan? After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
OK, here's the ResScan FL graph with the range set to 0 to 1.
.
20180814 ResScan FL.png

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Me and My AirSense Now with screenshots!

Post by palerider » Sun Aug 19, 2018 2:15 pm

BlueDragon wrote:
Sun Aug 19, 2018 12:21 pm
OK, here's the ResScan FL graph with the range set to 0 to 1.
Compared to what?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 3:35 pm

BlueDragon wrote:
Sun Aug 19, 2018 12:21 pm
djams wrote:
Sun Aug 19, 2018 7:57 am
Hey thanks for doing this. Since you have resscan, would you be willing to update the post with a screenshot of the "for her" FL chart in from resscan? After setting resscan FL chart scale to match SH, as mentioned earlier in this thread.
OK, here's the ResScan FL graph with the range set to 0 to 1.
Thank you again! I've tried as best I can to overlay the SH and resscan FL charts here. In order to get the times lined up, I had to stretch the SH chart considerably, so everything's a bit distorted on it.

What this shows me is that I'm all wet. Not just wet, soaking wet. :lol: Looks like pretty much a duplicate to me. I was expecting some significant differences, and there are none. Doesn't look like SH has any problem at all here.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins
Attachments
FL_chart_RS_SH.PNG
FL_chart_RS_SH.PNG (202.71 KiB) Viewed 13086 times

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Me and My AirSense Now with screenshots!

Post by djams » Sun Aug 19, 2018 3:42 pm

zonker wrote:
Sun Aug 19, 2018 11:08 am
i'm going to go ahead and post that "odd" night, just for the record-
You still hanging around this thread? :lol:

Nice night last night. Feeling pretty good today?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Me and My AirSense Now with screenshots!

Post by palerider » Sun Aug 19, 2018 3:51 pm

zonker wrote:
Sun Aug 19, 2018 11:08 am
comments welcome if you see anything you wish to comment upon.
like.... how you posted the same pic twice?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11364
Joined: Fri Jun 19, 2015 4:36 pm

Re: Me and My AirSense Now with screenshots!

Post by zonker » Sun Aug 19, 2018 4:54 pm

djams wrote:
Sun Aug 19, 2018 3:42 pm
zonker wrote:
Sun Aug 19, 2018 11:08 am
i'm going to go ahead and post that "odd" night, just for the record-
You still hanging around this thread? :lol:

Nice night last night. Feeling pretty good today?
yeah, i have NO idea why! :lol: :lol: :lol:

did feel good, though i woke up lots last night.

we'll see how it goes.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11364
Joined: Fri Jun 19, 2015 4:36 pm

Re: Me and My AirSense Now with screenshots!

Post by zonker » Sun Aug 19, 2018 4:57 pm

palerider wrote:
Sun Aug 19, 2018 3:51 pm
zonker wrote:
Sun Aug 19, 2018 11:08 am
comments welcome if you see anything you wish to comment upon.
like.... how you posted the same pic twice?
yes, how DID i do that?

<sigh> i guess i should've said constructive comments welcome.

but seriously, how DID i do that? as far as i could tell, i posted one, then the other. but two of the first one showed up. then the other one.

i think maybe there should be a sticky post: ZONKER-This is how you do stuff!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg