ASV SleepyHead red dots on flow rate graph?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

ASV SleepyHead red dots on flow rate graph?

Post by photonic » Mon Aug 04, 2014 7:56 am

I noticed these red dots on the flow rate graph. What do they mean? Cheers.

Image
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV SleepyHead red dots on flow rate graph?

Post by Pugsy » Mon Aug 04, 2014 8:16 am

See if this tutorial helps
http://www.healthcare.philips.com/pwc_h ... _Guide.pdf

Red dots are usually pressure pulses
Little puffs of air to help the machine decide if what it is sensing is related to something with the airway obstructed with collapsing airway tissues or cessation of breathing due to a clear airway event (central).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: ASV SleepyHead red dots on flow rate graph?

Post by JDS74 » Mon Aug 04, 2014 9:35 am

On my graphs for my 960 machine, the red dots correspond to machine triggered breaths.
Here is a sequence from my data.
Image

I think that for non-ASV machines, the red dots do correspond to the pressure pulses that the machine is creating to differentiate between obstructive and central apneas.

Edited:
I took a quick look at a similar circumstance from my 750 data and got this:
Image
The red dots there certainly correspond to pressure pulses as the 750 doesn't have machine triggered breaths.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV SleepyHead red dots on flow rate graph?

Post by Pugsy » Mon Aug 04, 2014 5:06 pm

I think Photonic has a model 960 ASV...so it may be PTB...with the equipment being missing I have to stretch my brain....anyway it should be explained in the tutorial link above for Encore as that is for the ASV machine.

It looks more like PTB and not the typical pressure pulses during a flow reduction now that I look a little closer.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

Re: ASV SleepyHead red dots on flow rate graph?

Post by photonic » Mon Aug 04, 2014 8:51 pm

Hi gang. There seems to be some debate I see about these dots. Darn the equipment is still not showing up on the bottom of the post. But comments are so I am going to put my equip in there and hope it shows up.

I AM using the PR 960 ASV.

Couldn't quite understand the abbreviation PTB so still unsure and the referred to tutorial unfortunately refers to the Encore software and its display, but doesn't seem to have these specific colored square "dots" that SleepyHead is displaying.

Hmmm, the mystery continues. Should I try to contact our fearless app builder on this question?
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software

sawinglogz
Posts: 400
Joined: Mon Aug 27, 2012 12:53 pm

Re: ASV SleepyHead red dots on flow rate graph?

Post by sawinglogz » Mon Aug 04, 2014 9:12 pm

Patient-triggered breaths.

I haven't really poked at SH's 960 support yet, so I can't tell you what the dots mean there.

User avatar
Pugsy
Posts: 65255
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV SleepyHead red dots on flow rate graph?

Post by Pugsy » Mon Aug 04, 2014 9:32 pm

photonic wrote: Should I try to contact our fearless app builder on this question?
I would.

Encore always did/does the pressure pulses in red and that's why I originally said pressure pulses because Mark tried to keep the colors the same.
I have Encore but I don't have a PR S1 ASV to use to see what it might show and compare it to Sleepyhead.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ASV SleepyHead red dots on flow rate graph?

Post by palerider » Mon Aug 04, 2014 10:18 pm

photonic wrote: Hmmm, the mystery continues. Should I try to contact our fearless app builder on this question?
I'll ask him the next time I see him on chat.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

Re: ASV SleepyHead red dots on flow rate graph?

Post by photonic » Mon Aug 04, 2014 10:27 pm

Thanks and I just PMed him....
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software

User avatar
Sludge
Posts: 953
Joined: Wed Dec 04, 2013 4:36 am

Re: ASV SleepyHead red dots on flow rate graph?

Post by Sludge » Tue Aug 05, 2014 4:37 am

Those "red dots" look more like "mauve squares" which would make them TBs.
You Kids Have Fun!!

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: ASV SleepyHead red dots on flow rate graph?

Post by igdoc » Tue Aug 05, 2014 4:43 am

Photonic
Just In case you have not heard from Jedimark, who must be incredibly busy! The red dots are 'timed breaths' and correspond to the purple dots on Encore Pro. These are where your next breath has been slower to start than the machine expected and it puts in a breath for you. So these are 'machine' triggered breaths and are related to the BPM setting (breaths per minute) which is usually set to 'Auto'. The more of these the machine has to put in then obviously the percentage of patient triggered breaths reported will be lower. I usually hover around 96% of my own breaths and 4% machine breaths. It is a crude indication of how irregular your breathing is and how often the machines has to take over for you.
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

User avatar
photonic
Posts: 261
Joined: Fri Jun 27, 2014 7:42 pm

Re: ASV SleepyHead red dots on flow rate graph?

Post by photonic » Tue Aug 05, 2014 1:56 pm

Got a great and swift reply back from jedimark. He notes that the ASV machine is very different than all the other types of lower level machines and that the "dots" in question used to refer to pressure pulses in the lower levels of machines. He purposefully just made an assumption that the dots on the AVS machine "might" be the same but notes that the data for those dots is different than what was used before and therefore he admits that he really does not know what they are really all about! He suggested that someone might want to compare those dots and data in Encore if someone is savvy enough to do that. Um that would not be me - LOL. He said if someone could figure it out, he would adjust the program.

Well there you have it. A big question mark. Nice guy to answer so swiftly. Sure appreciate his efforts.
ResMed AirCurve 10 ASV
8 years with F&P Oracle mouth mask now changing to
F&P Evora Full with S-M size
Hozer hose management device (a miracle!)
OSCAR software

igdoc
Posts: 54
Joined: Sun Dec 22, 2013 1:07 pm
Location: England

Re: ASV SleepyHead red dots on flow rate graph?

Post by igdoc » Tue Aug 05, 2014 3:42 pm

Jedimark is amazing. Sleepyhead/Encore comparison done and I can confirm the results as per my previous post.
Ian

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead v0.9.6. Encore Pro 2.12. Complex sleep apnea.
IG

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: ASV SleepyHead red dots on flow rate graph?

Post by JDS74 » Tue Aug 05, 2014 4:43 pm

just so everyone can see the Encore Pro color keys:
Image
SleepyHead doesn't have a similar display but incorporates color definitions in the various graphic displays.

Note that Encore Pro does have a color key for pressure pulses for the 960 ASV but I have not seen them in the waveform displays.
Doesn't mean they are not sometimes there, just I didn't see them and I suppose it contributes to the confusion.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
jedimark
Posts: 1300
Joined: Fri Feb 18, 2011 12:45 am
Location: Bundaberg, Australia
Contact:

Re: ASV SleepyHead red dots on flow rate graph?

Post by jedimark » Tue Aug 05, 2014 7:10 pm

JDS74 wrote:just so everyone can see the Encore Pro color keys:
Image
SleepyHead doesn't have a similar display but incorporates color definitions in the various graphic displays.

Note that Encore Pro does have a color key for pressure pulses for the 960 ASV but I have not seen them in the waveform displays.
Doesn't mean they are not sometimes there, just I didn't see them and I suppose it contributes to the confusion.
Interesting.. The 960 has PressurePulse as well as a Timed Breath code.. I'll have a hunt for that in the data and create a new channel accordingly...
I also don't think I've currently got any LargeLeak flags detected for 960 data either.. (I just added calculation capabilities for this based off the new redline threshold, but I'd rather use the machine detected large leaks on PRS1)

It would very much help having a matching encore waveform report, and the data set to go with it that demonstrates both these fields.

I'll have a bit of a think about what can be done to add a legend. But then again, I've just added hover overs in 0.9.7 so it might not be necessary..

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :)
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code