What is a PS setting on a ResMed9 BiPap?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CoffeePgh
Posts: 11
Joined: Wed Jul 02, 2014 1:40 am

What is a PS setting on a ResMed9 BiPap?

Post by CoffeePgh » Wed Jul 16, 2014 4:12 am

My pressure is set at min 6 and max 25. My ramp is 5. My AHIs all average 8.5-10.

For the last week, my Median/Max EPAPs are : 8.56/13.96; 8.54/13.98; 7.06/13.32; and 6.65/15.16. (These are the first readings with the increase to 6.)

What is a PS setting? Mine is set at 4. Is that a constant, meaning that my IPAP will always be 4 over my EPAP? From my readings, I think that it is.

Is there any way to make the PS fluctuate, allowing me to have a higher/lower IPAP score other than 4 over the EPAP? And, do I want the PS to fluctuate?

Or, should I be focusing more on the EPAP numbers and adjusting my minimum pressure so that it is more in line with my median score?

Thank you for the guidance.

_________________
MaskHumidifier
Additional Comments: Start date: 4/23/14

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What is a PS setting on a ResMed9 BiPap?

Post by Pugsy » Wed Jul 16, 2014 4:29 am

CoffeePgh wrote:What is a PS setting? Mine is set at 4. Is that a constant, meaning that my IPAP will always be 4 over my EPAP? From my readings, I think that it is.
With the S9 VPAP Auto bilevel machine the PS is a constant fixed setting even in Auto mode. In auto mode EPAP and IPAP may vary but the difference will always be whatever the PS is set at.
The Respironics bilevel pressure machines in auto mode do offer an auto adjusting PS though.
I am unsure as to exactly why it might be of benefit to have an auto adjusting PS. I have some theories but haven't found any concrete definitive reasons. I have used both brands of bilevel auto machines. In terms of therapy effectiveness I couldn't tell that one way of doing PS was any better or worse than the other. Each has their pros and cons.
CoffeePgh wrote:My AHIs all average 8.5-10.
What is the event category breakdown for the AHI numbers? Is it primarily Obstructive in nature (OAs and hyponeas) or Centrals?
CoffeePgh wrote:should I be focusing more on the EPAP numbers and adjusting my minimum pressure so that it is more in line with my median score?
I don't know that I would fixate on the median score until I determined what the machine was wanting to do.
If the AHI is primarily composed of Obstructive in nature events then (and since your machine seems to want to increase the pressure and it won't do that for centrals) I would target EPAP if the AHI is primarily OAs.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Stevoreno_55
Posts: 369
Joined: Tue Jan 14, 2014 10:28 pm
Location: MS Gulf Coast

Re: What is a PS setting on a ResMed9 BiPap?

Post by Stevoreno_55 » Wed Jul 16, 2014 9:19 am

CoffeePgh wrote:My pressure is set at min 6 and max 25. My ramp is 5. My AHIs all average 8.5-10.

For the last week, my Median/Max EPAPs are : 8.56/13.96; 8.54/13.98; 7.06/13.32; and 6.65/15.16. (These are the first readings with the increase to 6.)

What is a PS setting? Mine is set at 4. Is that a constant, meaning that my IPAP will always be 4 over my EPAP? From my readings, I think that it is.

Is there any way to make the PS fluctuate, allowing me to have a higher/lower IPAP score other than 4 over the EPAP? And, do I want the PS to fluctuate?

Or, should I be focusing more on the EPAP numbers and adjusting my minimum pressure so that it is more in line with my median score?

Thank you for the guidance.
That's a wide pressure range; a minimum of 6cm and a max of 25cm; whoa!!! I'll soon start on a fixed Bi-Level with an IPAP of 22 and an EPAP of 18. I couldn't even feel 6cm; lucky you.

_________________
Mask
Additional Comments: Newly diagnosed pressures as of June 05, 2014 are 22cmIPAP and 18cmEPAP; first diagnosed with OSA in 1999.

CoffeePgh
Posts: 11
Joined: Wed Jul 02, 2014 1:40 am

Re: What is a PS setting on a ResMed9 BiPap?

Post by CoffeePgh » Wed Jul 16, 2014 4:24 pm

Thanks for the input. I really do appreciate it.