New Trilogy User Needs Help!!! Anyone??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

New Trilogy User Needs Help!!! Anyone??

Post by Madalot » Tue Apr 29, 2014 8:21 am

We have a new member with some serious medical issues using a Trilogy. I am WAY out of my element when it comes to trying to help other than specifics on where to find things on the Trilogy.

Can any of you offer up some assistance to this member?

viewtopic/t97397/saturation-C02-levels- ... o-use.html

She just told me via PM that she is using the same mode as I am -- S/T with AVAPS enabled. Beyond that, I am not sure exactly what her settings are. But I am scared to do too much because her medical issues are serious.

Below is a new thread from Indi's Gmomma with test result information:

viewtopic/t97713/Diagnosis-settings-results-oh-my.html

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7
Last edited by Madalot on Fri May 09, 2014 4:59 am, edited 5 times in total.

User avatar
Pugsy
Posts: 65373
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sludge, -SWS, Pugsy - Can Anyone Help?

Post by Pugsy » Tue Apr 29, 2014 8:29 am

I have seen the posts. The problem is way beyond my comfort level to offer any ideas.
I am not even sure what the problem is ...much less fix it. It's not typical OSA or even CSA, from what I gather, and more of a respiratory issue with the imbalance between carbon dioxide and oxygen that may be caused by something other than sleep apnea...or at least sleep apnea hasn't been confirmed and the Trilogy trial was a stab at trying something.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Sludge, -SWS, Pugsy - Can Anyone Help?

Post by Madalot » Tue Apr 29, 2014 8:33 am

Pugsy wrote:I have seen the posts. The problem is way beyond my comfort level to offer any ideas.
I am not even sure what the problem is ...much less fix it. It's not typical OSA or even CSA, from what I gather, and more of a respiratory issue with the imbalance between carbon dioxide and oxygen that may be caused by something other than sleep apnea...or at least sleep apnea hasn't been confirmed and the Trilogy trial was a stab at trying something.
Thanks, Pugsy. That's kind of how I feel as well. I am happy to offer up help on the Trilogy, but when it comes to changing the settings because of the issues she's describing, it's way beyond my experience and comfort level as well.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Indi's Gmomma
Posts: 29
Joined: Fri Mar 21, 2014 7:46 pm

Re: Sludge, -SWS, Pugsy - Can Anyone Help?

Post by Indi's Gmomma » Tue Apr 29, 2014 9:15 am

This helps a great deal, just knowing its not normal is very helpfull and confirms our thoughts something is wrong here, just wanted to rule out the easy stuff, but being so new to all this we just weren't sure thanks

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Sludge, -SWS, Pugsy-Trilogy Can Anyone Help?

Post by Madalot » Tue Apr 29, 2014 4:37 pm

Bump - so it doesn't go off the page.

Robysue has been wonderful and offered some great insight and suggestions. Maybe if we all pull together and can get ideas, Indi's Gmomma can get her medical team to make this better for her!!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Sludge, -SWS, Pugsy-Trilogy Can Anyone Help?

Post by chunkyfrog » Tue Apr 29, 2014 5:04 pm

I wonder if the Trilogy people might have some tech info they could share with the doctor.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Sludge, -SWS, Pugsy-Trilogy Can Anyone Help?

Post by Madalot » Wed Apr 30, 2014 5:06 am

chunkyfrog wrote:I wonder if the Trilogy people might have some tech info they could share with the doctor.
Good idea (plus another bump from me) --

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Sludge, -SWS, Pugsy-New Trilogy User. Help?

Post by Madalot » Thu May 01, 2014 5:08 am

Another bump -- sorry to keep doing this, but I am hoping Sludge sees this and takes an interest. I'm hoping he might have some good insights.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
StuUnderPressure
Posts: 1378
Joined: Mon Jun 18, 2012 10:34 am
Location: USA

Re: Sludge, -SWS, Pugsy-New Trilogy User. Help?

Post by StuUnderPressure » Thu May 01, 2014 3:56 pm

No need to be sorry.

It is important enough for that action.

And, hey, I just saved you another bump.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02

SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Sludge, -SWS, Pugsy-New Trilogy User. Help?

Post by Sir NoddinOff » Thu May 01, 2014 4:24 pm

Kind of like asking how many people on the forum have done a valve job on a Lamborghini. Not an impossible question, but tough one. I hope that person gets the info they need.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: Sludge, -SWS, Pugsy-New Trilogy User. Help?

Post by Madalot » Fri May 02, 2014 5:01 am

Sir NoddinOff wrote:Kind of like asking how many people on the forum have done a valve job on a Lamborghini. Not an impossible question, but tough one. I hope that person gets the info they need.
This is true. But there are a handful of people that have some pretty good insights on situations like this (Sludge, -SWS, Pugsy) being the ones that came to mind when I was writing the original post.

Robysue, JohnBFisher, jnk -- thank so much for offering input on it even though I didn't mention you all specifically. I know that Indi's Gmomma appreciates all the input she can get.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: New Trilogy User Needs Help!!! Anyone??

Post by Madalot » Sat May 03, 2014 7:18 am

I am bumping again. We've got her a bit better with a lower Backup Rate, but she still is nowhere near dialed in....

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: New Trilogy User Needs Help!!! Anyone??

Post by chunkyfrog » Sat May 03, 2014 1:58 pm

Madalot, you are so sweet. Imdi's Gmomma has an angel indeed.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Indi's Gmomma
Posts: 29
Joined: Fri Mar 21, 2014 7:46 pm

Re: New Trilogy User Needs Help!!! Anyone??

Post by Indi's Gmomma » Sat May 03, 2014 3:19 pm

chunkyfrog wrote:Madalot, you are so sweet. Indi's Gmomma has an angel indeed.
I couldn't agree more, and I believe I even mentioned it in a pm

User avatar
Madalot
Posts: 4287
Joined: Sat Jan 30, 2010 7:47 am

Re: New Trilogy User Needs Help!!! Anyone??

Post by Madalot » Sun May 04, 2014 9:55 am

chunkyfrog wrote:Madalot, you are so sweet. Imdi's Gmomma has an angel indeed.
No, I'm no angel. I just know what it's like to be in Indi's Gmomma's situation and I was lucky enough to get a lot of help when I got here. I'm just trying to help her get the same help I received.
Indi's Gmomma wrote:I couldn't agree more, and I believe I even mentioned it in a pm
You have but it's not necessary. Just trying to help the way I was helped.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy EVO. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7