water reservoir

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Pattipayroll

water reservoir

Post by Pattipayroll » Sun Mar 30, 2014 12:15 pm

For the past week, the water reservoir on my machine has not needed filling on every other night. One morning the reservoir is almost dry and the next morning it is almost full. Con some one explain why this happens? The room temperature has been consistent.

Thanks

xyz
Posts: 407
Joined: Fri Dec 07, 2007 1:38 pm

Re: water reservoir

Post by xyz » Sun Mar 30, 2014 12:21 pm

> One morning the reservoir is almost dry and the next morning it is almost full.
> Can some one explain why this happens?

Your numbers.

If you have apap you might have averaged 15 one night and 10 the next.

Higher numbers use more water (for me).

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: water reservoir

Post by bwexler » Sun Mar 30, 2014 12:21 pm

If you fill in your equipment profile that would help us help you. Also let us know where you live. The answer in New orleans might be different than Tucson.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: water reservoir

Post by Pugsy » Sun Mar 30, 2014 12:26 pm

What machine and humidifier are you using?
The newer machines will see water consumption vary a lot sometimes because sensors are used to help regulate humidity.

Things that affect water consumption...
Ambient room humidity...has it been raining a lot where you live? If your humidifier setting is rather low and the air in your house already has some moisture in it then maybe the machine simply doesn't need to add much more moisture to the air.
Leaks...big leaks will use up more water
Pressure...higher pressure needs will use more water

I prefer humidifier set to maximum humidity levels...when it rains outside and I sleep with the windows open...it barely uses any water at all but if the windows are shut and/or the humidity is on the low side of where it normally is...I will use almost all the water in the tank in one night and not change any setting.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/