What about system one filters?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Eureka
Posts: 146
Joined: Wed Feb 18, 2009 11:54 am

What about system one filters?

Post by Eureka » Sat Mar 22, 2014 3:01 pm

I am switching over from the Series M Respironics apap to the System one. I really like it, but wonder about the filters. Is the black filter the same size and interchangeable? And the DME (Apria) didn't recommend using the white filter along with the black one. Those of you who have this apap -- do you use both as with the M series, or just the black one?

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: What about system one filters?

Post by JDS74 » Sat Mar 22, 2014 3:17 pm

I'm not familiar with the sizes of the M series filters.
However, use both the white and the dark grey filters in the series One.
The dark grey one catches big stuff while the white one catches the ultrafine stuff.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

grillhead
Posts: 110
Joined: Fri Aug 03, 2012 2:39 pm
Location: West Virginia

Re: What about system one filters?

Post by grillhead » Sat Mar 22, 2014 3:36 pm

Definitely use the white and black filters together. My white filter is gray when I change it out because it catches a lot of stuff.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Mirage™ FX Nasal CPAP Mask with Headgear
Additional Comments: SleepyHead and EncoreBasic 2.0 Software; Pressure-12

djhall
Posts: 323
Joined: Fri Jun 21, 2013 10:47 am
Location: Vacaville, CA

Re: What about system one filters?

Post by djhall » Sat Mar 22, 2014 3:42 pm

Eureka wrote:I am switching over from the Series M Respironics apap to the System one. I really like it, but wonder about the filters. Is the black filter the same size and interchangeable? And the DME (Apria) didn't recommend using the white filter along with the black one. Those of you who have this apap -- do you use both as with the M series, or just the black one?
I have both and I'm pretty sure the black filters and the white filters without tabs can be used in either machine as is. The white filters with the tabs don't fit the System One unless you trim them down a bit.

User avatar
Pugsy
Posts: 65368
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What about system one filters?

Post by Pugsy » Sat Mar 22, 2014 3:44 pm

The black/gray foam filters are interchangeable between the M Series and the System One. The size is tiny bit different but not enough to be a problem.

I also use both the foam and the white fine filter...it gets pretty yucky looking after 3 or 4 weeks. I figure if the filter is catching that stuff I would rather it not make it to my lungs or into the inner workings of the machine. Just an extra layer of protection. It is considered optional by Respironics but not optional in my mind.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Eureka
Posts: 146
Joined: Wed Feb 18, 2009 11:54 am

Re: What about system one filters?

Post by Eureka » Sat Mar 22, 2014 4:21 pm

Thanks for the replies. I am just a month into using the System One, and am only using the black filter. I wasn't comfortable without using the white filter, but two techs said just the black one was adequate. I wash the filter every Saturday, but will order the white filters and use both. I use the machine eight to nine hours a night, and it just seems best to use both filters. Will order the white ones Monday.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 

pbriggs
Posts: 173
Joined: Thu Aug 15, 2013 11:56 am

Re: What about system one filters?

Post by pbriggs » Tue Mar 25, 2014 5:46 pm

nothing against other great ways to get supplies, but for filters (white ones) I had good luck with this person - http://www.ebay.com/itm/281100953074?ss ... 1497.l2649 - they were true Respironics one with their logo on it. So far so good with them - I change them in both my machine and my wife's machine every 3 or 4 weeks. I would never go without one especially after seeing one that has been used...

_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear
Additional Comments: 14/10 with REM Sleep Behavior Disorder

User avatar
Eureka
Posts: 146
Joined: Wed Feb 18, 2009 11:54 am

Re: What about system one filters?

Post by Eureka » Wed Mar 26, 2014 10:41 am

Thanks, pbriggs-- that's a good tip. I am placing an order today. BTW I am using the simplus mask now and loving it. Actually my numbers weren't that great at the first, but getting better. I guess it takes some time to get adjusted to a new mask.
This forum is great -- so many helpful people.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 

ClayL
Posts: 369
Joined: Mon Feb 12, 2007 8:45 am
Location: Palisade, CO now, was Full time in a motor home for 12 years- Anywhere USA

Re: What about system one filters?

Post by ClayL » Wed Mar 26, 2014 1:48 pm

I have a number of the old M Series black and the white filters. Like Pugsy said the tab on the white ones can trimmed. I did that for a while but now just fold the tab. With or without the tab I have a problem with the filters coming out. I use a rubber band placed so as to keep that from happening.
Respironics DreamStation
Pressure = 11.5 min 14.5 max
C-Flex = 1

Dreamwear FF

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: What about system one filters?

Post by bwexler » Wed Mar 26, 2014 5:04 pm

I use both the black and white filters.
I could sure use some detailed help in how to place the white filter so it does not just pop out of the machine.
The tab on the PR brand white filters acts like spring to eject both filters.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
Pugsy
Posts: 65368
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What about system one filters?

Post by Pugsy » Wed Mar 26, 2014 5:11 pm

bwexler wrote:I use both the black and white filters.
I could sure use some detailed help in how to place the white filter so it does not just pop out of the machine.
The tab on the PR brand white filters acts like spring to eject both filters.
Actually Respironics changed those tabbed filters not too long after the System One was released to help prevent that tab from pushing the filters out. The new ones for the PR S1 filter don't have the tabs and are a little lighter and more flexible and should stay in better.

If you are still getting filters with tabs...those are outdated. They don't really hurt anything I don't think but I remember reading somewhere about why Respironics changed and it had to do with airflow being better with the new non tabbed filters..in addition to the popping out thing.

Short of getting the latest filters instead of the ones designed for the M series machines...you might try cutting off the tab.
That's what I did until I used up my supply of M series filters when I got my PR S 1 machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: What about system one filters?

Post by bwexler » Wed Mar 26, 2014 5:26 pm

Thank you Pugsy.
I have only had the PRS1 since October. So once again Apria proves to right up to date and helpful.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
Pugsy
Posts: 65368
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What about system one filters?

Post by Pugsy » Wed Mar 26, 2014 5:33 pm

bwexler wrote:I have only had the PRS1 since October. So once again Apria proves to right up to date and helpful.

Yep...they are sending out old tabbed filters to get rid of old stock...and they are cheaper too.

I wish I could remember where I read the thing about the change in filters with the PR S1 but it had to do with air flow and effectiveness and it was kinda important...that's why they changed.

I think I have a spare filter laying around since I am using the S9 right now.
Send me a private message with your name and address and I will pop one in the mail for you to see.
It's a substantial difference and if it were me...I would want to be using what Respironics says needs to be used in the S1...and they changed it for a reason and it wasn't totally because the darn things kept popping out.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: What about system one filters?

Post by bwexler » Wed Mar 26, 2014 5:54 pm

Thanks again Pugsy.
I called Apria and I will convince them to replace the tab filters with the newer 60 series filters ASAP.
So you can hang on to yours until a more needy/less vocal person needs it.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i