setting time on PR System One

Post by msat » Sat Nov 23, 2013 12:48 am

Hi all,

When I download to Sleepyhead, the times are wrong -- it used to register about 25 minutes later than I actually slept, and now it's closer to 35 minutes. I believe it's my machine, not Sleephyhead that's the problem -- I tried to compare with what the technician downloaded on his professional software the last time I was in, but we were a bit confused going back and forth between the two formats and didn't definitively decide what was wrong.

In any case, I kind of use the data to keep track of when I actually went to bed (as opposed to when I intended to go to bed ) and when I woke up, and this really bugs me. The technician didn't have a clue how to adjust the time, either. Surely there must be a way of doing so -- I mean, what if I moved to a different time zone?

Thanks for any help you can offer.

DeepBreathing
Posts: 18
Joined: Mon Oct 21, 2013 12:23 am
Location: Perth

Re: setting time on PR System One

Post by DeepBreathing » Sat Nov 23, 2013 6:19 am

My understanding is that the PR machines have an internal clock set to UTC (or GMT, depending on where you went to school). The clock can only be set by a service technician (no doubt using arcane incantations). Seems like a silly system to me.

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead software

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: setting time on PR System One

Post by robysue » Sat Nov 23, 2013 7:26 am

DeepBreathing,

You are correct: The System One's internal clock cannot be set via the clinical menu or the patient's menu and it is a stupid system.

Msat,

SleepyHead does have the ability to make a "time correction" in the user preferences. It's under the CPAP tab---look for "CPAP clock drift." Keep in mind that the drift is measured in seconds, so you'll want to multiply that 35 minute drift by 60 before you change the drift setting in SH

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: setting time on PR System One

Post by Pugsy » Sat Nov 23, 2013 7:54 am

As already mentioned by Robysue and Deep Breathing...it's a known issue with Respironics machines of all ages and models.
The internal battery clock is often off from real time and most of the time it is late and there's not a thing we can do about it at the machine level short of the machine going back to the factory and getting a factory reset fix for the clock and a new battery.

You can do what Robysue says with SleepyHead and the software will make the correction. That's about all you can do if knowing the exact time the machine gets turned on is important to you.

We don't know why the clock gets so off with some Respironics. I have had 6 or 7 different machines over the years and my very first machine was a 2 year old M series machine with zero hours of use and it was only about 5 minutes off. Go figure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: setting time on PR System One

Post by JDS74 » Sat Nov 23, 2013 9:22 am

DeepBreathing wrote:My understanding is that the PR machines have an internal clock set to UTC (or GMT, depending on where you went to school). The clock can only be set by a service technician (no doubt using arcane incantations). Seems like a silly system to me.
Accurate Biomedical Services can perform this service for you.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: setting time on PR System One

Post by Goofproof » Sat Nov 23, 2013 11:02 am

JDS74 wrote:
DeepBreathing wrote:My understanding is that the PR machines have an internal clock set to UTC (or GMT, depending on where you went to school). The clock can only be set by a service technician (no doubt using arcane incantations). Seems like a silly system to me.
Accurate Biomedical Services can perform this service for you.
And in 3 months you can send it in again, and again, because your xpap isn't a Timex, and it's clock isn't meant to keep accurate time. If you need the clock to match clock time subtract or add the error factor, but time is relative, what matters is the events and getting your Apnea controlled. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

msat
Posts: 34
Joined: Fri Dec 09, 2011 4:21 pm

Re: setting time on PR System One

Post by msat » Sat Nov 23, 2013 7:21 pm

Thanks for your replies; I'll stop trying to figure out how to make this change on my machine.

Robysue, I can't find that setting on Sleepyhead. The tabs I've got are Welcome, Daily, and Overview, and the menus I have are File, View, Data, and Help. The File menu has preferences, but I can't find anything to adjust for time correction or drift there. Am I missing something?

And Goofproof, if my issue were only sleep apnea, then I might agree with you about not needing to know the exact time, but I'm also trying to manage my sleep-wake cycle in general. Also, sometimes it's useful to correlate with something that happened in the environment, like a phone call or a power outage or whatever. Sure, it's possible to subtract the time difference, but it's a PITA and sometimes confusing. I keep daily notes, and I'd really like the times I write to correspond with the times on the chart. Last but not least, psychologically I'd feel a lot better with my sleep-wake times not looking even later than they are. None of this is a big deal in terms of my therapy, but if there had been a simple fix, it sure would have been nice.

M.

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: setting time on PR System One

Post by Pugsy » Sat Nov 23, 2013 8:32 pm

msat wrote:The File menu has preferences, but I can't find anything to adjust for time correction or drift there. Am I missing something?
What version and build of SleepyHead are you currently using? Go to Help and "about SleepyHead" to see version and build date.
It's there on the Windows version 0.9.3 build May 2013
Preference..then "CPAP" tab...upper right at the very top.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

msat
Posts: 34
Joined: Fri Dec 09, 2011 4:21 pm

Re: setting time on PR System One

Post by msat » Sat Nov 23, 2013 10:13 pm

Ah, I've got an older version of Sleepyhead. It automatically gets updated regularly, so I thought I was on top of things. Do I have to download a whole new version of it?
Thanks.
M.

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: setting time on PR System One

Post by Pugsy » Sat Nov 23, 2013 10:30 pm

Are you sure it is getting "updated regularly" because there was only one update ever?..most often what would happen was it would check for updates...tell you there was one...and then fail to install update because Sleepyhead didn't have Administrator privileges.
If you have older than 0.9.1 version...you need a total new install.
If you are seeing 0.9.1 and the update is failing...try opening SH with a right click of the mouse and select "run as administrator" and see if it updates to 0.9.2-something. I don't remember when the clock drift was added...it may have been 0.9.3 and in that case, yes, you need to install the new version.
If you uninstall the prior version your data folder in MyDocuments should still remain intact and I don't think you will lose anything assuming you have 9.1 version
If you have 8.8 version...just leave it on your computer and install 9.3 separately...that's what I did.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

msat
Posts: 34
Joined: Fri Dec 09, 2011 4:21 pm

Re: setting time on PR System One

Post by msat » Sat Nov 23, 2013 11:30 pm

Ah, spiffy new Sleepyhead! I had version 8.8. I guess that every time I imported data, it informed me that it was "checking for updates," and I just assumed it was being updated! So I'm having fun setting all my preferences, and hopefully I can get the time correction right (will have to make a careful note of when I actually put on the mask tonight or take it off tomorrow morning).
Thanks, Pugsy!
M.

sleepinglass
Posts: 135
Joined: Thu Mar 22, 2012 11:09 pm

Re: setting time on PR System One

Post by sleepinglass » Sun Nov 24, 2013 4:31 am

Wow Pugsy!

Learning to set the time is so useful! I have set mine to minus 1800 seconds and it works!

I now can work out from my Sleepy Head charts what time I went to bed, which for me is a useful excercise.

Pity we cannot do the same with Resscan.

Thank you for that superbly useful tip. I am over the moon!!!

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: PR System One BiPAP autoSV Advanced:Epap Min:9 Max:12. PS Min:6 PS Max:12 Max press:25 Back up:Auto.Humidif: set@ 4 DreamStation BiPap AutoSV
Legacy Remstar Auto with humidi, S9 AutoSet with H51 humidi & H.hose. Many unsuccessful attempts at bipap. NOW on PR System One BiPAP autoSV Advanced, most comfortable, best machine ever!

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: setting time on PR System One

Post by Pugsy » Sun Nov 24, 2013 6:30 am

sleepinglass wrote:I now can work out from my Sleepy Head charts what time I went to bed, which for me is a useful excercise.

Pity we cannot do the same with Resscan.
Don't need to do it with ResScan....The S9 has a clock that you can set so the time is exact.
It's available in the clinical set up menu in the Options menu.
http://www.apneaboard.com/resmed-s9-cpap-setup

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

sleepinglass
Posts: 135
Joined: Thu Mar 22, 2012 11:09 pm

Re: setting time on PR System One

Post by sleepinglass » Sun Nov 24, 2013 9:26 am

Sorry Pugsy getting myself confused!

I meant that I cannot set the time on my encore basic and my PR system one, like the other people have found.

At the moment when I go to bed, it believes I am going to bed at one o'clock in the afternoon, which is rather irritating,but not the end of the world thanks to SH.

I remember correcting the time on my S9 and that is really cool.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: PR System One BiPAP autoSV Advanced:Epap Min:9 Max:12. PS Min:6 PS Max:12 Max press:25 Back up:Auto.Humidif: set@ 4 DreamStation BiPap AutoSV
Legacy Remstar Auto with humidi, S9 AutoSet with H51 humidi & H.hose. Many unsuccessful attempts at bipap. NOW on PR System One BiPAP autoSV Advanced, most comfortable, best machine ever!

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: setting time on PR System One

Post by Pugsy » Sun Nov 24, 2013 10:56 am

sleepinglass wrote:I meant that I cannot set the time on my encore basic and my PR system one, like the other people have found.

At the moment when I go to bed, it believes I am going to bed at one o'clock in the afternoon,
EncoreBasic doesn't have a clock setting (time zone) available...instead it uses your computer clock and the internal clock in the machine to calculate time...but it normally isn't a full day off or hours and hours off.

The new day always starts on the machine at 12 Noon GMT time. Where do you live and just how much is the time off by?
When it says you turned the machine on at 1 PM..what time is it really?

Now EncorePro does allow a time zone adjustment...if you want to try to handle that PITA beast.