Provider manual above...it explains how to get to the clinical setup menu where things can be changed.
Constant puffing
Re: Constant puffing
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Constant puffing
I get to the menu, but when I click on the mode option, I can't change it away from CPAP. It's stuck.Pugsy wrote:http://www.apneuvereniging.nl/forum/pdf ... manual.pdf
Provider manual above...it explains how to get to the clinical setup menu where things can be changed.
Re: Constant puffing
Ted,
If you have an s9 Elite, it cannot be changed from CPAP mode. You'd need an Autoset model to change that setting.
As a relatively new s9 Eilte user, I've never experienced anything remotely close to what you are describing. I'd bring it to the DME and have them plug it in and test it. Sounds like you probably need it fixed/replaced. No way would I sleep with something puffing at me every second!
I'd address it right away. After finding out about all the nasty health effects that go along with apnea, I wouldn't want to sleep without my machine.
If you have an s9 Elite, it cannot be changed from CPAP mode. You'd need an Autoset model to change that setting.
As a relatively new s9 Eilte user, I've never experienced anything remotely close to what you are describing. I'd bring it to the DME and have them plug it in and test it. Sounds like you probably need it fixed/replaced. No way would I sleep with something puffing at me every second!
I'd address it right away. After finding out about all the nasty health effects that go along with apnea, I wouldn't want to sleep without my machine.
_________________
| Mask: Eson™ Nasal CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Last edited by Iowamv on Sun Sep 29, 2013 9:29 am, edited 1 time in total.
- BlackSpinner
- Posts: 9742
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Constant puffing
The elite only does cpap. You need an autoset to do auto adjusting pressures.tedthefed wrote:
Does anyone know how to change the modes on a Resmed S9 Elite? I don't know if it'd make a difference, but it's set to CPAP mode, and I can't change it away from that mode. I think changing to Auto would be worth a shot, but it won't change away from CPAP.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Constant puffing
@tedthefed:
Have you tried a cod reset.
Just unplug the machine, count to 20 (not fast) and plug back in.
You have an Elite if it says "Elite" just over the power button; if not, please tell us what it does say there.
Have you tried a cod reset.
Just unplug the machine, count to 20 (not fast) and plug back in.
You have an Elite if it says "Elite" just over the power button; if not, please tell us what it does say there.