2 Year Mark - Giving up

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Random
Posts: 14
Joined: Thu May 06, 2010 9:02 pm

2 Year Mark - Giving up

Post by Random » Sat Jun 08, 2013 6:13 pm

Hello Everyone,

I'm now at the two year mark as a compliant CPAP user.. I even switched to an Autopap in the hopes of feeling better.

Unfortunately after 2 years of using my machine for 8 hours per night I am still miserable. I can sleep fine with it on and it lowers my AHI below 1-2, however I still can't go 10 minutes without yawning and I feel like my mind is in a deep fog all day.

I know I have a small issue with leaks, but otherwise my data appears to be very good. If anyone has any suggestions I would appreciate it.

I'm not too familiar with sleepyhead but I've posted a few screenshots here.

Image
Image

Thank you.

_________________
Mask

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: 2 Year Mark - Giving up

Post by BlackSpinner » Sat Jun 08, 2013 6:41 pm

Cpap therapy does a lot more for your body then make you feel awake - tiredness is just one symptom and not even the most important one. There is blood pressure, stroke, diabetes and many other organs that work better with it.

Your tiredness could be coming from something altogether different.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
STL Mark
Posts: 950
Joined: Thu May 06, 2010 1:37 pm
Location: St. Louis, MO - USA

Re: 2 Year Mark - Giving up

Post by STL Mark » Sat Jun 08, 2013 6:42 pm

Before you quit using the machine I would consider talking with your physician and letting them know how you are feeling. It is quite possible that you have other issues in addition to sleep apnea that require treatment. If you have already stopped CPAP therapy it will be difficult to get medical personnel to take you seriously when your not doing your part to solve the problems.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: S9 VPAP Adapt (eASV) 36037

Mode: ASVAuto, Min EPAP: 4, Max EPAP: 15, Min PS: 3, Max PS: 15, Ramp: Off - Original Titration: 18

Wulfman...

Re: 2 Year Mark - Giving up

Post by Wulfman... » Sat Jun 08, 2013 7:52 pm

Random wrote:Hello Everyone,

I'm now at the two year mark as a compliant CPAP user.. I even switched to an Autopap in the hopes of feeling better.

Unfortunately after 2 years of using my machine for 8 hours per night I am still miserable. I can sleep fine with it on and it lowers my AHI below 1-2, however I still can't go 10 minutes without yawning and I feel like my mind is in a deep fog all day.

I know I have a small issue with leaks, but otherwise my data appears to be very good. If anyone has any suggestions I would appreciate it.

I'm not too familiar with sleepyhead but I've posted a few screenshots here.

Thank you.
It could be that the pressure changes are wrecking your sleep. The APAPs use air flow and algorithms determine whether they need to change pressures. If you're entering a deep sleep stage or REM, your breathing may change and the machine may misinterpret your breathing and trigger pressure increases. This may bump you out of the much-needed stages of sleep and leave you with sub-optimal sleep quality.
Your comments lead me to think that might be part of your problem. Another thing is that your pressure settings are also not optimal (minimum pressure too low).


Den

.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: 2 Year Mark - Giving up

Post by Goofproof » Sat Jun 08, 2013 7:59 pm

You have had two sucessful years of xpap treatment, you never died from sleep apnea. Maybe if you quit, lack of treatment will end your suffering.

You posted leak rate is far from ldeal, no wonder you don't see any improvement. I consider 35 to 38 lpm leak rate to be the max allowed for good treatment.

I'm sure quiting will work for you, it's real easy to learn, almost no effort is needed to succeeded. The downside you may need to use your major medical a little more, till your body fails from lack of O2, it takes a while. Jim
Last edited by Goofproof on Sat Jun 08, 2013 9:04 pm, edited 1 time in total.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

hyperlexis
Posts: 876
Joined: Tue Aug 14, 2012 6:56 am
Location: Illinois

Re: 2 Year Mark - Giving up

Post by hyperlexis » Sat Jun 08, 2013 8:21 pm

Honestly, none of us (I think) are MDs.

Your doctor - or a doctor - is who you should be talking to, not a bunch of people on the internet. As well meaning as we are at the end of the day you need an MD.

You obviously have some sleep or neurological problem that is making you so tired. What else is left? Your mattress?

You have an auto machine and that should be getting your sleep disordered breathing under control as best as is technologically possible in 2013.

The rest is in your sleep doctor's purview to try and fix. However, simply ending cpap may do more harm than good.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: 2 Year Mark - Giving up

Post by BlackSpinner » Sat Jun 08, 2013 8:44 pm

Yes see a doctor

Image

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: 2 Year Mark - Giving up

Post by DoriC » Sat Jun 08, 2013 8:47 pm

Please don't give up. I think the advice to try straight pressure is a good one, maybe 8cms and if necessary 9cms. You also need to work on your leaks, have you tried any other masks? It's always best to check with your Dr but I think you could really benefit from adjusting your pressure settings and controlling your leaks, that's probably what's disturbing your sleep. Keep us posted.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: 2 Year Mark - Giving up

Post by jdm2857 » Sat Jun 08, 2013 9:08 pm

Could you post detailed data for a few typical nights. They can be more useful
that the summary data you posted.
jeff

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: 2 Year Mark - Giving up

Post by kaiasgram » Sat Jun 08, 2013 9:55 pm

Random,

Today is my one-year 'anniversary' of being on PAP and your story moved me to share my story with you. First, take a look at my Sleepyhead stats for the year:

Image

Impressive, isn't it? But the real story is that like you, I feel no better than before I started. I feel worse. I worked hard the first half of the year to get leaks under control, to find optimal pressure(s), even switched from cpap to bilevel which I find more comfortable to breathe with, but which still did not lead to an improvement in how I feel or function. Popular opinion is that if your therapy is looking optimal then your ongoing problems must be due to some other underlying issue. While that is probably true most of the time I'm no longer so sure that it's always the case. I do wonder about the potential iatrogenic effects of PAP therapy that may not be well understood yet. I do have a pre-existing condition that has some effect on my sleep, but even so I would have thought that effectively treating my severe sleep apnea for a year would have brought about some improvement in how I feel.

As sad, frustrated, and disappointed as I feel about my year on PAP, I intend to stick with it and I hope you will too. Because if nothing else, we are not having oxygen deprivation all night long, and that counts for something -- actually a lot -- even if we don't really feel better. Any time I feel like quitting (which is pretty much every day ) I look at the desaturation graph from my first night with an oximeter and no cpap and I know I can't let that continue. I dutifully dragged the damn machine through the airport for the first time last week so that I wouldn't have four nights without treatment while I was out of town.

You are young, and it really matters that you keep breathing at night. Your leaks could be messing significantly with your sleep quality, causing arousals or mini arousals. As a few others suggested, perhaps you are sensitive to pressure changes and would fare better on straight cpap or with a smaller apap range. And of course, there is the question of whether other health issues are interfering with your sleep. So do follow up with your doctor. I have a neurology appointment next week to see if I can get more help with whatever is keeping me out of deep sleep and pushing me dangerously close to disability.

So, let's make a deal -- we both keep trying. OK?

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Always tired
Posts: 252
Joined: Sat Sep 15, 2012 4:03 pm
Location: Northern Nevada

Re: 2 Year Mark - Giving up

Post by Always tired » Sat Jun 08, 2013 10:48 pm

kaiasgram wrote:So, let's make a deal -- we both keep trying. OK?
Random,
Excellent advice, don't give up.
A T

_________________
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: Also PR TrueBlue nasal, Mirage Quattro, Aloha, pressure 6.5--10.0,
Om Mani Padme Hum

User avatar
caffeinatedcfo
Posts: 690
Joined: Sun Mar 24, 2013 9:19 am
Location: Upstate NY

Re: 2 Year Mark - Giving up

Post by caffeinatedcfo » Sun Jun 09, 2013 2:48 am

You need professional advice, but I wonder what else exists in your environment.

For example, I'm up at 4:30 posting here because my wife's snoring woke me up and was keeping me up. I will be a tired & cranky today because of this, not ineffective treatment.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead software; using APAP mode 10-12cm & EPR 3

User avatar
Denial Dave
Posts: 1334
Joined: Wed Sep 19, 2012 6:45 am
Location: Connecticut

Re: 2 Year Mark - Giving up

Post by Denial Dave » Sun Jun 09, 2013 5:07 am

I'm up before 5AM every day... M - F because of work schedule.... weekends, I can't sleep more than 7 hours anymore.

my mother in law used to participate in the "don't ask, don't tell" program with her doctor.....since he didn't ask, she didn't tell him her issues.

make a list of your concerns & issues and go over each of them with your doctor.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Started at VPAP pressure setting of 20/14.4. I've survived Stage 1 cancer and lost 80+ lbs. Pressure is now 14.5 / 11
if you don't know where you are going... any road will take you there.... George Harrison

DocWeezy
Posts: 468
Joined: Thu Nov 11, 2010 12:04 pm

Re: 2 Year Mark - Giving up

Post by DocWeezy » Sun Jun 09, 2013 11:52 am

I can really empathize with your frustration! I'm a little over 2 years on CPAP and am just now beginning to really solve my sleep problems. HOWEVER, treating my OSA was only the FIRST step to sleeping and feeling better....not the last step. Without treating the OSA, you can never really figure out what's happening because the OSA will cover up all other symptoms or potential improvement. At any rate, your body functions better with appropriate oxygen at night! I really wanted to be one of those people who had a miraculous improvement once on CPAP, but that was not to be. I've spent over two years trying everything under the sun to stay asleep and wake up rested. Turns out I really had a deep level of chronic insomnia...who knows, maybe it was caused by a lifetime of suffocating in my sleep and in turn it seriously impacted my metabolism and hormonal system.

It is very possible that you have other physical issues going on that are ruining your sleep. But you HAVE to treat the OSA or else nothing else will ever get better because the chronic sleep deprivation keeps your body from healing or working well at all.

You may want to find an endocrinologist or other physician who is really attuned to the hormonal/physical subtleties that can impact sleep. Our bodies aren't as simple as we'd like them to be and it can take a lot of work and trial and error to find the right solution. You may need different nutrition, thyroid or other hormonal medications, or even supplements (amino acids have really become my friend and are making an impact!).

CPAP is not a magic bullet for many of us--darn it! But it is the first step towards solving sleep issues. You can't fix anything if you're still suffocating at night!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hybrid is alternate mask

Zorki1c

Re: 2 Year Mark - Giving up

Post by Zorki1c » Sun Jun 09, 2013 7:23 pm

I'm very new at this but after a few nights I discovered the best results by turning off ramping and all the auto pressure changes and just going from start to finish of the night with the doctor prescribed 8.0 pressure. Also, quite a few people would kill for the stats you posted. I'm think there are issues outside of CPAP at work here.