Where did all the help GO??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
racprops
Posts: 522
Joined: Thu Mar 21, 2013 6:12 am

Re: Where did all the help GO??

Post by racprops » Wed May 15, 2013 6:07 am

Sorry to bother you again BUT:

I am reaching a near braking point.

Currently I seem to only sleep an hour of two at a time.

Four days ago I thought I had a break though when I had a good night, during that night I had a long sleep of 4+ hours, with only want seems to be only 2 Hypopneas and one snore events.

I figured I was making head way.

BUT The three nights sense have been horrible. As above, only an hour at a time and only when I am so tired I fall asleep at a drop of a head.

And the reports are also horrible: the 12th only 6.21 hours of sleep with a AHI of 14.78, the 13th 4.53 hours of sleep on the machine and a AHI of 25.52.

Tonight with only 2.52 of sleep with the machine and a AHI of 27.90 (so far)

I even gave up and tried to sleep in my bed without the machine, sleeping on my side.

No good: one hour later I woke up sore and upset..I checked the O2 recorder and sure enough I had a lot more trouble with sats during that hour, so far on desating below 88 % for a few seconds…but I feel the many more ups and down during that hours shows I cannot sleep safely.

It looks like I have to figure this out, I have no other choose.

BUT It is getting harder and harder to sleep on the machine.

I am running into two main problems, ONE I have had to tighten the mask to the point of pain, because even when I lower the max pressure the machine will blow so hard it make the mask fart..a leak like a fart and this wakes me up.

OR I wake up with so little air flow I feel like I am suffocating from lack of air. For the first time I felt closeaphobtic , trapped in a too tight a place.

Yesterday morning around 5PM this happened for the third or fourth time and I could NOT put the mask back on, so I got up.

Another too tired day with me trying to fall asleep on and off all day.

AS I have said before IF I cannot stand the mask and machine even if the treatment is not as good as I need, then it will fail completely. It nearly has.

Then there is the machine over reacting as I try to fall asleep, and taking over my breathing..or jumping to max pressure and blowing the mask off…waking me as I am just about to drop off…

AND I find that the higher the EPAP Pressure the more the machine wants to be in control.

At 9CM it follows my breathing and if I hold my breath takes a few seconds to react. And will return to my controlling breathing.

BUT at 11cm it does NOT follow my breathing but instead does its own breathing and if I fail to keep up reacts with higher and higher power pressers blows. It takes over. It seemly will not allow me to be in control.

IS this normal?

So I am going nuts: Is the settings wrong (for me) of is the machine going bad??


Is it me OR the machine OR the settings and can they be set better??


Rich

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: I USE a Responics 960 ASV machine and a New Resmed Form lined fullface Mask plus Sheepyhead

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Where did all the help GO??

Post by JohnBFisher » Wed May 15, 2013 11:34 am

racprops, I'm on the upswing from an ugly chest cold. They always knock me down (mentally and physically). I will try to look at your information and post a note later today .. but it might be tomorrow.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

racprops
Posts: 522
Joined: Thu Mar 21, 2013 6:12 am

Re: Where did all the help GO??

Post by racprops » Wed May 15, 2013 11:37 am

Thanks get well.

Rich

JohnBFisher wrote:racprops, I'm on the upswing from an ugly chest cold. They always knock me down (mentally and physically). I will try to look at your information and post a note later today .. but it might be tomorrow.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: I USE a Responics 960 ASV machine and a New Resmed Form lined fullface Mask plus Sheepyhead

racprops
Posts: 522
Joined: Thu Mar 21, 2013 6:12 am

Re: Where did all the help GO??

Post by racprops » Sat May 18, 2013 5:45 am

BREAK THOUGH!!! NO EVENTS!!!!

I felt it was time for a desperate act, so far I have not slept over 4 hours at one time on the machine..and that seemed to only happen after many nights of one hour at a time..

So last night I took a Zantac.

I woke up a little stiff and need a bathroom break.

Then I relisted I have slept for seven hours..nearly a full night’s sleep.

I had to see how I did.

So I plugged in the SD card and downloaded it.

First I felt: ON NO it malfunctioned, it failed to scan my last night how did I mess it up..

AND then I relised it had not malfunctioned that many of the readouts were reporting breathing.

Then I had a OMG moment.

I HAD A PERFICE NIGHT ZERO AHI, ZERO APNEAS, ZERO EVERYTHING.

SEE:

Image

Image

Image

Image

Image

This would and could not have happen with out all your support and help.

Rich

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: I USE a Responics 960 ASV machine and a New Resmed Form lined fullface Mask plus Sheepyhead

User avatar
49er
Posts: 5624
Joined: Mon Jan 16, 2012 8:18 am

Re: Where did all the help GO??

Post by 49er » Sat May 18, 2013 6:49 am

Rich,

I am so happy for you after all the struggles you have been through. This has really made my day. Much continued success.

49er
racprops wrote:BREAK THOUGH!!! NO EVENTS!!!!

I felt it was time for a desperate act, so far I have not slept over 4 hours at one time on the machine..and that seemed to only happen after many nights of one hour at a time..

So last night I took a Zantac.

I woke up a little stiff and need a bathroom break.

Then I relisted I have slept for seven hours..nearly a full night’s sleep.

I had to see how I did.

So I plugged in the SD card and downloaded it.

First I felt: ON NO it malfunctioned, it failed to scan my last night how did I mess it up..

AND then I relised it had not malfunctioned that many of the readouts were reporting breathing.

Then I had a OMG moment.

I HAD A PERFICE NIGHT ZERO AHI, ZERO APNEAS, ZERO EVERYTHING.

SEE:

Image

Image

Image

Image

Image

This would and could not have happen with out all your support and help.

Rich

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Where did all the help GO??

Post by NateS » Sat May 18, 2013 8:56 am

Congratulations, Rich!

Very happy to hear the good news!

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: Where did all the help GO??

Post by DoriC » Sat May 18, 2013 9:37 am

To everyone who has given Rich their support, expertise and Patience, please take a bow!! AMAZING!!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Where did all the help GO??

Post by HoseCrusher » Sat May 18, 2013 9:55 am

This is great news.

One night doesn't make a trend so now the challenge is to develop a trend of good nights sleep. It looks like you have all the pieces in place so this may be as easy as falling asleep...

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

racprops
Posts: 522
Joined: Thu Mar 21, 2013 6:12 am

Re: Where did all the help GO??

Post by racprops » Sat May 18, 2013 12:48 pm

It took 29 days to reach this.

And as I tried to go back asleep and got another 2.75 hours of sleep the machine the racked up 3.24 events 1.05 CA and 2.19 Hypopneas.

So I do know this will not be every night, but I also now know it can happen and that I am on the right track.

Rich

PS I will point out it took a Xanax and sleeping in my recliner is a near upright position. There is still work to do.
HoseCrusher wrote:This is great news.

One night doesn't make a trend so now the challenge is to develop a trend of good nights sleep. It looks like you have all the pieces in place so this may be as easy as falling asleep...

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Additional Comments: I USE a Responics 960 ASV machine and a New Resmed Form lined fullface Mask plus Sheepyhead
Last edited by racprops on Sat May 18, 2013 2:34 pm, edited 1 time in total.

Claudio
Posts: 3
Joined: Sat May 18, 2013 1:06 pm

Re: Where did all the help GO??

Post by Claudio » Sat May 18, 2013 2:02 pm

"I keep waking up and ripping off my mask as I feel suffocated..."

I don't have much technical knowledge in CPAP but could only share what happened to me when I felt suffocated and had to rib off my own mask.

This may not apply to most people and may somewhat help a few but it is worth mentioning in case it is the same happening to a few as it has happened to me.

The first thing that has happened to me was that my nose is usually congested and needs to be clean, sometimes very minor but hard to breath. Even when it is clean, when I lay down, my nose air passage gets smaller and hard to breath. Sometimes it is hard to breath from the start and other times it is later own when sleeping.

One remedy was to use nasal pillow mask and I usually need to use nasal strips as well. This worked the best for me, along with a few other things.

The other problem (because of my nose problem), was my pressure was too high; it was set that way at first because I needed the air flow. Then I had a hard time expelling air and suffocating, the ramp, inspiratory and expiratory pressure had to be adjusted. The auto CPAP machines must adjust to each individual. I tried a ResMed VPAP ADAPT machine for the first time last night.

When I felt suffocated in the past, it was a combination of nose problem, air intake, air out, air pressure and sometimes rain out. I also had it bad in certain positions of my head on my back, the air flow got worse.

Once I learned the best mask to use and fit, the best position, humidity, in/out air pressure and better nose air flow, things are way better.

For those who have no problem breathing properly, that’s great, but for a few of us it is a learning experience.

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: Where did all the help GO??

Post by NateS » Sat May 18, 2013 2:57 pm

racprops wrote:It took 29 days to reach this.
Everyone told you it would take a month!

Regards, Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

sleepinow
Posts: 462
Joined: Thu Apr 25, 2013 9:40 am

Re: Where did all the help GO??

Post by sleepinow » Sat May 18, 2013 3:14 pm

Keep up the good work racPROPS, you have helped me through my struggle too. We both have the same machines so yeah it was hard. Good to see things change for you and thanks for your help!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Trying my best to get quality rest. PR System ONE REMstar BiPAP Auto SV Advanced
System One Respironics Bipap AutoSV Advanced Quattro Fx

Min EPAP: 13.0 CmH20
Min Pressure Support: 3.5
Max EPAP: 25.0
Max Pressure Support: 5.0
Max Pressure: 20.0
Flex Setting: Bi-Flex - 3
Backup Rate:Auto
Humidification Mode:off
Humidifier Setting:c5

User avatar
RandyJ
Posts: 1673
Joined: Thu Apr 21, 2011 7:22 pm
Location: Connecticut, USA

Re: Where did all the help GO??

Post by RandyJ » Sat May 18, 2013 5:44 pm

While every night might not yield as great a result, at least you know that your current settings are good and you can relax in that regard.

Good luck!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Where did all the help GO??

Post by JohnBFisher » Sat May 18, 2013 7:36 pm

Zantac as in an antacid? Hmmm... It's possible the stomach acid was causing you to awaken during the night. That would trigger sleep state transition apneas - which appear to be a real struggle for you. Yes, I can believe that that's what helped.

But don't panic if it doesn't always work. It sounds as if getting to you fall asleep and stay asleep can be a struggle. Those nights always are lousy for me, as well. Be patient. You will find you tend to sleep better more frequently. But it requires patience on your part.

And here's hoping you will have found the secret to help you get better sleep. Best wishes for much more!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

racprops
Posts: 522
Joined: Thu Mar 21, 2013 6:12 am

Re: Where did all the help GO??

Post by racprops » Sat May 18, 2013 7:44 pm

Sorry I misquote it was xanax.

Rich