(Problem Solved) S9 VPAP SmartStart Unavailable

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
5100HI
Posts: 6
Joined: Thu Mar 07, 2013 8:10 pm

(Problem Solved) S9 VPAP SmartStart Unavailable

Post by 5100HI » Sun Apr 21, 2013 5:05 am

I would appreciate assistance on this one.

When in the clinical menu my S9 VPAP's SmartStart option is greyed out and does not appear to be changeable.

I changed the mask type selection from full face to nasal and the option still is greyed out in the clinical menu and still can not be selected.

Is there any other setting anyone knows of that would disable the ability to turn on SmartStart options from the clinical menu?

Thanks,

Michael
Last edited by 5100HI on Sun Apr 21, 2013 6:30 pm, edited 1 time in total.

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: S9 VPAP SmartStart Unavailable

Post by SleepingUgly » Sun Apr 21, 2013 5:48 pm

You are able to change other settings? If so, I don't know. If not, maybe you've been locked out of changing things in the clinician's menu.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
Pugsy
Posts: 65439
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: S9 VPAP SmartStart Unavailable

Post by Pugsy » Sun Apr 21, 2013 6:03 pm

Go into the clinical menu and turn Leak Alert off and see if that will give you access to Smart Start choices.
I may have to RISE but I refuse to SHINE.

5100HI
Posts: 6
Joined: Thu Mar 07, 2013 8:10 pm

Re: S9 VPAP SmartStart Unavailable

Post by 5100HI » Sun Apr 21, 2013 6:29 pm

Problem solved, thank you for your assistance!

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: S9 VPAP SmartStart Unavailable

Post by NateS » Mon Apr 22, 2013 10:14 am

5100HI wrote:Problem solved, thank you for your assistance!
You don't want to tell us how you solved it?

Regards, Nate
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: (Problem Solved) S9 VPAP SmartStart Unavailable

Post by squid13 » Mon Apr 22, 2013 10:37 am

I agree with Nate, tell how you solved it so in the future if anyone else has this problem it maybe the fix.
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Pugsy
Posts: 65439
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: (Problem Solved) S9 VPAP SmartStart Unavailable

Post by Pugsy » Mon Apr 22, 2013 10:41 am

NateS wrote:You don't want to tell us how you solved it?
I am pretty sure it was turning off the Leak Alert.
On my S9 VPAP if the Leak alert is On I can't access Smart Start.
We can't have both functions available at the same time.
I think this is specific to the VPAPs...I don't think leak alert is available on the Elite/Autoset models but I have never seen one so I don't know for sure.
Also I didn't find it in the provider manual but I may have overlooked it.
I found it simply by turning Leak Alert off...when I did my Smart Start (which is right below leak alert) turned gray.
I figured something was a trigger and just played with stuff till I figured out which was the trigger.
I may have to RISE but I refuse to SHINE.

5100HI
Posts: 6
Joined: Thu Mar 07, 2013 8:10 pm

Re: (Problem Solved) S9 VPAP SmartStart Unavailable

Post by 5100HI » Mon Apr 22, 2013 12:21 pm

Pugsy on Sun Apr 21, 2013 7:03 pm suggested "Go into the clinical menu and turn Leak Alert off and see if that will give you access to Smart Start choices."

I did that and the problem was solved.

When I wrote problem solved it was due to following Pugsy's advice: "Go into the clinical menu and turn Leak Alert off and see if that will give you access to Smart Start choices."