Well that would seem to be in conflict with the OP's current prescription, eh?4betterO2 wrote:First, your normal, usual respiratory rate must be determined, then a backup rate should be applied at -2 from the normal rate.
Which, for anybody keeping track (or giving a FF) is 14, whilemollete wrote:Well that would seem to be in conflict with the OP's current prescription...4betterO2 wrote:First, your normal, usual respiratory rate must be determined, then a backup rate should be applied at -2 from the normal rate.
the NPSG wrote:(that the respiratory rate) commonly declined to 10 to 12 bpm while asleep,
If that was measured using signal averaging over x seconds (or minutes), the dynamics of the area of respiratory disturbance would not be accurately respresented, i.e., there would be an event, followed by period of hyperventilation, followed by a central event, etc., unless this was another case of opiate-induced CSA, where RR would be decreased throughout (of course, it would be something like 4, and not 10).NPSG interpretation wrote:The respiratory events were more common when the respiratory rate was 10 bpm.
One thing bewilders me, the Rx recommendation does not add up.Nooblakahn wrote:Here is the results of my bipap asv test. Not entirely happy with the results...]
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Humidifier: DreamStation Heated Humidifier |
The prescription is acceptable: If EPAP ranges between 8cm and 10cm, then the entire PS range is allowed; but if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.4betterO2 wrote: One thing bewilders me, the Rx recommendation does not add up.
It says max pressure is 25 (that's all the PR ASV can deliver, don't know as for the Resmed equivalent machine?)
but it also says,
max EPAP= 15
max PS= 15
doesn't this add to a max pressure of 30, not 25?
...sloppy work (?)
| Mask: Quattro™ Air Full Face Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: ResMed S9 VPAP™ Adapt SV |
Thanks for the correction and reminder! In my case I'm really glad I got my max pressure reduced, I see it was the only way to get the needed limit on IPAP.-SWS wrote: if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Humidifier: DreamStation Heated Humidifier |
Don't worry about all those settings initially. However, PLEASE be certain the person who works at the DME understands how to properly set the prescription. It's pretty common that they get it wrong. I haven't followed which unit you are getting, but the prescription is based on the Respironics unit. The Resmed unit would not be able to follow the recommended prescription.Nooblakahn wrote:I appreciate the info.. but the last couple of posts here are way over my head. Still waiting on my machine. They got it Wednesday, but didn't submit stuff for the insurance auth on the Wednesday before as they were supposed to. I have an appointment set up for Monday but have no clue if they have my auth or not.. as the phones were conveniently down today at this place... So we will see what happens on Monday.
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
-SWS pardon me for correcting you a bit, but:-SWS wrote:The prescription is acceptable: If EPAP ranges between 8cm and 10cm, then the entire PS range is allowed; but if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.4betterO2 wrote: One thing bewilders me, the Rx recommendation does not add up.
It says max pressure is 25 (that's all the PR ASV can deliver, don't know as for the Resmed equivalent machine?)
but it also says,
max EPAP= 15
max PS= 15
doesn't this add to a max pressure of 30, not 25?
...sloppy work (?)

| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
Well, you lost me on that correction, John. Below is the OP's Rx followed by a table of allowable EPAP ranges for the OP's Rx. You can also see the allowable IPAP range at each EPAP value.JohnBFisher wrote:-SWS pardon me for correcting you a bit, but:-SWS wrote:The prescription is acceptable: If EPAP ranges between 8cm and 10cm, then the entire PS range is allowed; but if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.
Ranges at OP's Rx:Op's Rx wrote: Minimum EPAP 8cm, Maximum EPAP 15cm, Minimum Pressure Support 10, Maximum Pressure Support 15, Maximum Pressure 25cm, Backup rate 14 bpm, BiFlex of 3."
Thanks.-SWS wrote: If EPAP ranges between 8cm and 10cm, then the entire PS range is allowed; but if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.
| Mask: Quattro™ Air Full Face Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: ResMed S9 VPAP™ Adapt SV |
You guys should work together more often. You both said the same thing. Or close enough..JohnBFisher wrote:-SWS pardon me for correcting you a bit, but:-SWS wrote: The prescription is acceptable: If EPAP ranges between 8cm and 10cm, then the entire PS range is allowed; but if EPAP ranges above 10cm, then the resultant IPAP should be capped at max pressure.
However, since the ASV unit uses IPAP to overcome central apneas, the current maximum IPAP can range in this case upto 15cm H2O over the current EPAP. Again, if the current EPAP is 8cm and the current pressure support is 15cm H2O, then the current maximum IPAP would be 8cm + 15cm H2O .. or 23cm H2O. If the current EPAP was 15cm then the current maximum IPAP would be 15cm + 15cm H2O .. or 25cm H2O (since the maximum pressure is 25cm H2O).
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: S9 Adapt for Home and Travel, On-Board Firmware, Std. Tubing. EEP 9.8, Min PS 4.6, Max Pressure 21 |
Hey, don't blame me, Sir Banned... I said it, he expanded it with paraphrasing, and I'm still trying to understand what exactly was corrected. I suspect it might have to do with John's interpretation of how I used the term "capped".StillAnotherGuess/Banned wrote: You guys should work together more often. You both said the same thing. Or close enough..
That AHI is excellent.Nooblakahn wrote:Well... I already have the machine. Last night was one week of using it. I have a resperonics vpap adapt. The script as you said was wrong and they had to redo it. I haven't checked it to see if it is set right... but that explains why it had to be rewritten. It seems to be working.. I have been below one AHI for three or 4 out of these nights. I have been meaning to upload a few charts. Ill have to get that.
Glad they got the settings straightened out. It's a really common problem. They just don't deal with anywhere near as many ASV units as they do CPAP or BiPAP units.Nooblakahn wrote:Well... I already have the machine. Last night was one week of using it. I have a resperonics vpap adapt. The script as you said was wrong and they had to redo it. I haven't checked it to see if it is set right... but that explains why it had to be rewritten. It seems to be working.. I have been below one AHI for three or 4 out of these nights. I have been meaning to upload a few charts. Ill have to get that.