Water chamber won't close

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
trac1965

Water chamber won't close

Post by trac1965 » Wed Dec 12, 2012 3:49 am

1EVGoDYHELP

I cleaned out my hose and water chamber Sunday but now I can't get the lid back on the chamber!! I've had this problem on a much smaller scale but was almost able to make it fit so it would click together. Maybe the warm water warped the seal??

Thanks.

User avatar
xenablue
Posts: 1358
Joined: Tue Sep 07, 2010 12:52 pm
Location: Aussie living in balmy Wisconsin

Re: Water chamber won't close

Post by xenablue » Wed Dec 12, 2012 7:18 am

It would really help to know what machine you are using in order to help you.

Cheers,
xena

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-)

User avatar
StuUnderPressure
Posts: 1378
Joined: Mon Jun 18, 2012 10:34 am
Location: USA

Re: Water chamber won't close

Post by StuUnderPressure » Wed Dec 12, 2012 9:21 am

trac1965 wrote:1EVGoDYHELP

I cleaned out my hose and water chamber Sunday but now I can't get the lid back on the chamber!! I've had this problem on a much smaller scale but was almost able to make it fit so it would click together. Maybe the warm water warped the seal??

Thanks.
If you are using the ResMed S9 Series, the 4 latches are connected together by a band of the same material that wraps around the chamber.

That band sometimes slips down a little & therefore brings the latches down with it.

Go all around the chamber & make sure that band is as far up the chamber sides as it can be.
Make sure to do all 4 sides.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02

SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html

User avatar
RogerSC
Posts: 1914
Joined: Sat Jul 21, 2012 1:11 pm

Re: Water chamber won't close

Post by RogerSC » Wed Dec 12, 2012 7:22 pm

If you warm up the metal bottom with the plastic rim strip on it, like by running warm water on it or putting it in a pan or dishpan of warm water, it makes that strip much easier to move up far enough so that you can push the holder doo-dads on the sides closed *smile*. Then it tightens up again when it cools down. Otherwise, yes, it can be hard to move up high enough manually.
ResMed AirSense 10 Autoset
Philips Respironics Dreamwear nasal mask

Mine won't close...Please help

Re: Water chamber won't close

Post by Mine won't close...Please help » Sat Jun 06, 2015 7:59 am

My cover won't close. I even had my mechanically inclined son help me with no success. I thought that perhaps the seal had warped as well. I have no idea what to do to fix it, but I cannot sleep without it...Please help.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Water chamber won't close

Post by chunkyfrog » Sat Jun 06, 2015 12:11 pm

If it is an S9, slide the collar up until it stops, then fasten the clips.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Water chamber won't close

Post by archangle » Sun Jun 07, 2015 12:48 am


_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
RogerSC
Posts: 1914
Joined: Sat Jul 21, 2012 1:11 pm

Re: Water chamber won't close

Post by RogerSC » Sun Jun 07, 2015 2:17 am

See my post above from the original part of the thread. That really works, the ring actually gets loose enough to come off in warm water. So it's easy to move at that point. Has to do with the fact that when warmed, plastic expands more than metal...think about it *smile*.

RetPM
Posts: 11
Joined: Tue Jun 16, 2015 10:51 am
Location: Ohio

Re: Water chamber won't close

Post by RetPM » Fri Jul 17, 2015 11:25 am

This thread saved my sanity last night about 12:30 am when I was trying to put the water chamber back together before bedtime after cleaning it that morning!! No matter how I readjusted it I could not get the holes to line up with the tabs that snap into them. I actually thought I was losing my mind, as it had not been a problem when I took it apart earlier. I finally decided the water I washed it with must have been too hot and warped the seal or something. In desperation I googled the problem and found this thread, and of course I found an answer to the problem. The band around the metal reservoir had slipped down just a fraction, but enough to throw everything off. Moved it back up as far as I could get it and it was like magic!!

I gotta say that I love this forum! I've been reading since late May when I started on CPAP, or I guess really APAP, and have learned so much from it about being patient, fitting the mask properly, etc. It is a slow process of adjusting with no real immediate rewards, at least for me. (My AHI from the home study was a 7, just enough to be diagnosed with OSA by my internist who prescribed the machine--I'm assuming due to my having high blood pressure.) But this is the first time I have actually received immediate results from something I learned on this forum and just wanted to share and say thanks to all the experienced users who share their knowledge with us newbies.

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Water chamber won't close

Post by Sir NoddinOff » Fri Jul 17, 2015 12:20 pm

As a side note regarding similar problems on the PR sys one machine: Sometimes the plastic hook/lip on the inside of the tank that holds the lid down actually cracks where it makes its 90 degree turn. You won't actually notice it unless you look carefully. Eventually the crack gets bigger and that makes it hard for the lid to click in place while engaging the bottom of the tank. Eventually the catch lip breaks off and the catch fails... you'll have to tape the lid down securely to the bottom of the tank until you get a new tank (thin masking tape works). First make sure the silicone seal is seated properly before you go troubleshooting.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
Last edited by Sir NoddinOff on Fri Jul 17, 2015 12:22 pm, edited 1 time in total.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Water chamber won't close

Post by chunkyfrog » Fri Jul 17, 2015 12:21 pm

Zen and the art of sleeping with machinery.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Water chamber won't close

Post by Sir NoddinOff » Fri Jul 17, 2015 12:25 pm

chunkyfrog wrote:Zen and the art of sleeping with machinery.
Yeah, it's enough to drive you crazy when everything appears normal... then you have to dig deeper (or look more carefully). As my electronic's professor used to say: Observe, engage brain, observe again, then act.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

RetPM
Posts: 11
Joined: Tue Jun 16, 2015 10:51 am
Location: Ohio

Re: Water chamber won't close

Post by RetPM » Sat Jul 18, 2015 11:57 am

Sir NoddinOff wrote:
chunkyfrog wrote:Zen and the art of sleeping with machinery.
Yeah, it's enough to drive you crazy when everything appears normal... then you have to dig deeper (or look more carefully). As my electronic's professor used to say: Observe, engage brain, observe again, then act.
And if that doesn't work, come to this forum for CPAP answers and advice! Every problem I have run into since I started with CPAP has been addressed and discussed somewhere in this forum. A lot of experience, knowledge and excellent advice is found here.

paula4

Re: Water chamber won't close

Post by paula4 » Sun Nov 01, 2015 1:16 pm

THANK YOU FOR THE INFORMATION I WAS GOING CRAZY NOT BEING ABLE TO FIGURE IT OUT!



StuUnderPressure wrote:
trac1965 wrote:1EVGoDYHELP

I cleaned out my hose and water chamber Sunday but now I can't get the lid back on the chamber!! I've had this problem on a much smaller scale but was almost able to make it fit so it would click together. Maybe the warm water warped the seal??

Thanks.
If you are using the ResMed S9 Series, the 4 latches are connected together by a band of the same material that wraps around the chamber.

That band sometimes slips down a little & therefore brings the latches down with it.

Go all around the chamber & make sure that band is as far up the chamber sides as it can be.
Make sure to do all 4 sides.