Resmed Water Resevoir Will Not Close Please help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
StaceyA

Resmed Water Resevoir Will Not Close Please help

Post by StaceyA » Sat Jun 06, 2015 8:03 am

I took my Resmed water resevoir apart for cleaning and it won't go back together. I had my son who is very mechanically inclined try to help me put it back together, with no success. I have done it many times before. All the parts align except for where it snaps together. I thought that perhaps the seal swelled when I cleaned it with hot water...I believe that it is the S9 series. Please help!

User avatar
Pugsy
Posts: 65159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Water Resevoir Will Not Close Please help

Post by Pugsy » Sat Jun 06, 2015 8:13 am

The yellow gasket thing has slipped. I had this happen fairly often. Just move the gasket (I think up) all the way and then try to close the snap things. The yellow gasket does slide up or down just a little and it is this sliding that causes the malalignment.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

StaceyA

Re: Resmed Water Resevoir Will Not Close Please help

Post by StaceyA » Sat Jun 06, 2015 8:19 am

I do not have a yellow gasket. The rubber seal is firmly attached, and doesn't slide anywhere. I wish that I could post a pic.

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Resmed Water Resevoir Will Not Close Please help

Post by avi123 » Sat Jun 06, 2015 8:26 am

This is one of the Resmed H5i humidifier weak design. Also, to raise the cover you need to use both hands instead of it being springy and open by itself after you press the latch. It was designed by lousy engineers. Another problem with the Flip Lid is that it needs to be centered on the lower tank to work properly.


Replacement or Backup Humidifer Flip Lid Seal

The ResMed H5i Humidifier Flip Lid Seal for the H5i Heated Humidifier is a sillicone gasket that effectively locks the flip lid and water chamber, which prevents water from leaking into the machine. Easy to maintain, the gasket will need to be replaced as soon as it becomes damaged or torn.




Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Sat Jun 06, 2015 10:35 am, edited 4 times in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Water Resevoir Will Not Close Please help

Post by Pugsy » Sat Jun 06, 2015 8:26 am

What color is the rubber seal or gasket on the chamber that you do have?
And exactly which chamber do you have?
The standard chamber has a yellow ring thing
https://www.cpap.com/productpage/resmed ... amber.html
and the dishwasher chamber has a black ring thing
https://www.cpap.com/productpage/resmed ... amber.html
It's the ring thing that is causing the problem and it will move slightly and cause the snaps not to close.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Thanks...that is exactly what was wrong!

Re: Resmed Water Resevoir Will Not Close Please help

Post by Thanks...that is exactly what was wrong! » Sat Jun 06, 2015 8:34 am

Thanks! You were right! It wasn't yellow, but it was the bottom part of the clip part. It had slipped. Thanks so much!

StaceyA

Re: Resmed Water Resevoir Will Not Close Please help

Post by StaceyA » Sat Jun 06, 2015 8:39 am

Thanks so much! I am very grateful for this site! It has saved me now on two occasions! It took me 120 days to get used to my CPAP, but now I am addicted to the point that I panic when I can't get it working correctly, and since the place where I bought it is closed on Saturdays, I was in a bind. The last time it happened, it was in the middle of the night and I found some kind person to help then, too! Thanks soooooo much!

User avatar
Pugsy
Posts: 65159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Water Resevoir Will Not Close Please help

Post by Pugsy » Sat Jun 06, 2015 9:00 am

When this first happened to me I just kept fiddling with things and then it would close but I didn't know what I did to fix it. Then one day it happened and instead of going into frustration mode I slowly started fiddling with things until I discovered that little movement of that little ring thing. Once you know exactly where the problem is you should be able to easily fix it the next time it happens.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Resmed Water Resevoir Will Not Close Please help

Post by chunkyfrog » Sat Jun 06, 2015 11:24 am

That silly collar has slipped down on my tank, too.
Luckily the bullfrog figured it out in a snap.
Resmed's omission of some kind of label fails to consider our lack of mechanical prowess.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Water Resevoir Will Not Close Please help

Post by palerider » Sat Jun 06, 2015 5:17 pm

avi123 wrote:This is one of the Resmed H5i humidifier weak design. Also, to raise the cover you need to use both hands instead of it being springy and open by itself after you press the latch. It was designed by lousy engineers.
wrong again, avi, as usual, you've failed to understand the issue before climbing on your soapbox.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Resmed Water Resevoir Will Not Close Please help

Post by archangle » Sun Jun 07, 2015 12:45 am

Sounds like the problem was the translucent square plastic ring/collar that goes on the outside of the metal base and has the flaps that snap the top on. i.e. the "ears" on 4 sides of this picture.

Image

If this slides down a little, it makes it hard or impossible to close the top.

One trick is to run some warm water over the ring and get some of the water between the ring and the metal base. It will slide up and down more easily.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
RogerSC
Posts: 1911
Joined: Sat Jul 21, 2012 1:11 pm

Re: Resmed Water Resevoir Will Not Close Please help

Post by RogerSC » Sun Jun 07, 2015 2:14 am

archangle wrote:Sounds like the problem was the translucent square plastic ring/collar that goes on the outside of the metal base and has the flaps that snap the top on. i.e. the "ears" on 4 sides of this picture.

If this slides down a little, it makes it hard or impossible to close the top.

One trick is to run some warm water over the ring and get some of the water between the ring and the metal base. It will slide up and down more easily.
Yes, if you run warm water on the bottom of the container with the ring on it, the plastic ring expands more than the metal due to the heat, so the ring loosens up quite a bit, and you can move it around so that it allows the tank to close again. Found that out when I was washing the tank and the plastic ring came off easily in warm water *smile*.