Filters for Resmed S9

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Chikorita
Posts: 346
Joined: Mon May 28, 2012 7:15 am
Location: Blue Mountains, NSW, Australia

Filters for Resmed S9

Post by Chikorita » Wed Nov 28, 2012 7:40 pm

I've been looking online for filters. I notice some are made by ResMed and some are third party.
Is there anything I need to know about the difference between these please?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: :: Breathing at night is my new hobby :: Pressure range 7-12cm :: APAP since 3/8/2012 :: OSCAR ::
________
Chikorita in Australia :: APAP since 3/8/2012
1 God, 1 husband, 2 kids, 2 dogs, 8 chooks, 5 ducks, 3 budgies and a bee hive
Equipment - ResMed AirSense 10 Autoset :: AirFit P10 Nasal Pillow :: Pressure range 7-12cm :: OSCAR

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Filters for Resmed S9

Post by chunkyfrog » Wed Nov 28, 2012 8:02 pm

I've had both; can't tell the difference.
Still no clue which side goes out--if there's any difference at all.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

-tim
Posts: 710
Joined: Sat Dec 03, 2011 7:46 pm
Location: Melbourne Australia

Re: Filters for Resmed S9

Post by -tim » Wed Nov 28, 2012 11:27 pm

I use the cheap ones on the outside as a pre-filter and the resmed hypoallergenic one inside. The second filter will turn grey long before the outer filter. The resmed filter design isn't very good and if you aren't very careful, you create a gap that lets unfiltered air in.

I'm still trying to find an adapter that goes into the round filter hole so I can hook up a decent filter.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmeds overpriced SpO2

User avatar
Chikorita
Posts: 346
Joined: Mon May 28, 2012 7:15 am
Location: Blue Mountains, NSW, Australia

Re: Filters for Resmed S9

Post by Chikorita » Wed Nov 28, 2012 11:35 pm

I know this is a newbie question but why does the ai need to be filtered?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: :: Breathing at night is my new hobby :: Pressure range 7-12cm :: APAP since 3/8/2012 :: OSCAR ::
________
Chikorita in Australia :: APAP since 3/8/2012
1 God, 1 husband, 2 kids, 2 dogs, 8 chooks, 5 ducks, 3 budgies and a bee hive
Equipment - ResMed AirSense 10 Autoset :: AirFit P10 Nasal Pillow :: Pressure range 7-12cm :: OSCAR

User avatar
Pualani
Posts: 45
Joined: Mon Nov 26, 2012 9:56 am

Re: Filters for Resmed S9

Post by Pualani » Thu Nov 29, 2012 12:20 am

To filter out dust, dirt, and allergens. I am wondering how often you are supposed to change the filter. My husband has had his over a year and has changed the filter but it never changes color as mentioned above. Thanks.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Filters for Resmed S9

Post by archangle » Thu Nov 29, 2012 12:52 am

Apart from the basic health benefits of filtering the air, there is concern for the "health" of the CPAP machine itself. There's no way to clean dust off of the fan and other bits inside the machine. A lot of air flows through a CPAP machine in its operational life. Think of how much fuzz and stuff builds up on a room fan, or an air conditioner filter and imagine all that stuff inside your CPAP machine.

By the way, you need to be careful to not put the wrong kind of "custom" filter on the CPAP machine. If you block the airflow too much, the machine may overheat. There have been reports of machines overheating if the patient doesn't replace the filter reasonably often.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Filters for Resmed S9

Post by archangle » Thu Nov 29, 2012 1:08 am

Pualani wrote:To filter out dust, dirt, and allergens. I am wondering how often you are supposed to change the filter. My husband has had his over a year and has changed the filter but it never changes color as mentioned above. Thanks.
I believe that 6 months is the recommended maximum time, replacing it sooner if it gets dirty sooner.

[edit - 6 months for S9 filters only]

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Last edited by archangle on Thu Nov 29, 2012 2:48 pm, edited 2 times in total.
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

lazer
Posts: 1377
Joined: Fri Apr 20, 2012 1:56 pm
Location: Hermitage, PA
Contact:

Re: Filters for Resmed S9

Post by lazer » Thu Nov 29, 2012 7:18 am

Regarding the PSR1, are we talking the foam one and/or the white one in regards to replacement intervals suggested? Thanks

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead & Encore Basic Software & a Zeo
Image.....................................................ImagePress ESC if the animations BUG you!.....................................................Image

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Filters for Resmed S9

Post by archangle » Thu Nov 29, 2012 2:56 pm

lazer wrote:Regarding the PSR1, are we talking the foam one and/or the white one in regards to replacement intervals suggested? Thanks
Manufacturer's directions for PRS1 machines:

Grey filter - clean every two weeks, replace at 6 months.
White filter - replace every 30 days of use or when it looks dirty, do not clean.

The white filter is optional, but the machine might overheat if it is not replaced when dirty.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Kitatonic
Posts: 284
Joined: Fri Dec 16, 2011 9:11 pm
Location: Northern CA

Re: Filters for Resmed S9

Post by Kitatonic » Thu Nov 29, 2012 3:32 pm

Tim, I only change one filter on my S9,where is the other filter?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Use soft cervical collar

lazer
Posts: 1377
Joined: Fri Apr 20, 2012 1:56 pm
Location: Hermitage, PA
Contact:

Re: Filters for Resmed S9

Post by lazer » Thu Nov 29, 2012 3:51 pm

archangle wrote:
lazer wrote:Regarding the PSR1, are we talking the foam one and/or the white one in regards to replacement intervals suggested? Thanks
Manufacturer's directions for PRS1 machines:

Grey filter - clean every two weeks, replace at 6 months.
White filter - replace every 30 days of use or when it looks dirty, do not clean.

The white filter is optional, but the machine might overheat if it is not replaced when dirty.
Thanks for the information.

And to the follow-up post/user, I'm sorry I sorta Hi-jacked the thread with my question which pertains to a different machine. Your machine may very well have only 1 filter. I'm sure someone else can quickly confirm.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead & Encore Basic Software & a Zeo
Image.....................................................ImagePress ESC if the animations BUG you!.....................................................Image

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Filters for Resmed S9

Post by archangle » Thu Nov 29, 2012 6:04 pm

Kitatonic wrote:Tim, I only change one filter on my S9,where is the other filter?
There is only one filter at a time in the S9. You can use a regular or a hypoallergenic filter.

Many Respironics machines use a mandatory washable foam filter, and have an optional disposable fine filter under the foam filter.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

ozze_dollar
Posts: 611
Joined: Wed Jun 08, 2011 5:20 pm
Location: Sydney,Australia

Re: Filters for Resmed S9

Post by ozze_dollar » Thu Nov 29, 2012 8:30 pm

archangle wrote:
Kitatonic wrote:Tim, I only change one filter on my S9,where is the other filter?
There is only one filter at a time in the S9. You can use a regular or a hypoallergenic filter.

Many Respironics machines use a mandatory washable foam filter, and have an optional disposable fine filter under the foam filter.

I purchased some hypo allergic filters before I traveled to the Philippines and I have noticed they become a dark colour very quickly.Is this normal for these filters or is this just a lot of dirt?

User avatar
Chikorita
Posts: 346
Joined: Mon May 28, 2012 7:15 am
Location: Blue Mountains, NSW, Australia

Re: Filters for Resmed S9

Post by Chikorita » Thu Nov 29, 2012 10:19 pm

archangle wrote:Apart from the basic health benefits of filtering the air, there is concern for the "health" of the CPAP machine itself. There's no way to clean dust off of the fan and other bits inside the machine. A lot of air flows through a CPAP machine in its operational life. Think of how much fuzz and stuff builds up on a room fan, or an air conditioner filter and imagine all that stuff inside your CPAP machine.
Thank you. This makes perfect sense. My body can cope with a bit of dust but the machine can't filter itself or expel or trap dust.
I fully understand now.
I was thinking it was a human thing but forgot about the 'health' of the machine,

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: :: Breathing at night is my new hobby :: Pressure range 7-12cm :: APAP since 3/8/2012 :: OSCAR ::
________
Chikorita in Australia :: APAP since 3/8/2012
1 God, 1 husband, 2 kids, 2 dogs, 8 chooks, 5 ducks, 3 budgies and a bee hive
Equipment - ResMed AirSense 10 Autoset :: AirFit P10 Nasal Pillow :: Pressure range 7-12cm :: OSCAR

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Filters for Resmed S9

Post by archangle » Thu Nov 29, 2012 10:29 pm

ozze_dollar wrote:I purchased some hypo allergic filters before I traveled to the Philippines and I have noticed they become a dark colour very quickly.Is this normal for these filters or is this just a lot of dirt?
I suspect a bit of both. How quickly did you notice them becoming darker?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.