no humid air flow

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rick pen

no humid air flow

Post by rick pen » Sat Aug 04, 2012 11:10 am

i have a problem with my icon cpap machine. the humidifier plate is warm, the air is blowing, but after a few seconds of use the air drys out. any idea what the problem could be?

User avatar
LSAT
Posts: 13323
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: no humid air flow

Post by LSAT » Sat Aug 04, 2012 12:24 pm

Is the humidifier using water??? If it is, you are getting moist air.

User avatar
eaglett1111
Posts: 39
Joined: Fri Aug 03, 2012 1:36 pm
Location: Northern California

Re: no humid air flow

Post by eaglett1111 » Sat Aug 04, 2012 12:30 pm

rick pen -

I was wondering the same thing last night. For some reason I just didn't feel the humidity. Kept turning the humidifier up. I think it has something to do with the fit, but not sure. The machine is definitely using water.

_________________
Mask: Mirage™ FX Nasal CPAP Mask with Headgear
Additional Comments: Not sure if what I picked re: software is right. I use Encore Basic

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: no humid air flow

Post by archangle » Sat Aug 04, 2012 3:27 pm

Please sign up for an ID and fill in the equipment on your profile. There's a link below on how to do this.

The first few seconds of air will probably be more humid than the rest even if everything is right. Airflow speeds up and there's probably humid air collected above the water and in the hose.

If the water is warm and there are no leaks, there's basically no way for the humidification to fail. It may not be as humid as you want, but the physics sort of can't fail.

Disconnect the hose from your mask, plug up the end of the air hose with your thumb and listen for air leaks.

If you have mouth leaks or leaks around the edge of your mask, that can leak out enough air to end up dehydrated.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.