Sleepyhead data question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Sleepyhead data question

Post by Zzzzzzzzzzz... » Mon Mar 12, 2012 6:12 am

Hey all- it's my first post! (and there was much celebration)

Quick question. I've got three distinct data sections (nap/quick start stop/sleep) that obviously make up one full day and my AHI average is based on that. How do I highlight or isolate the actual section when I was out for the night so
I can analyze it?

Thanks and it's good to be here.

-Z

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: Sleepyhead data question

Post by b360155 » Mon Mar 12, 2012 7:15 am

I think you're saying you captured 3 sessions within one day and you only want to be looking at the 3rd one only? If that's what you're asking, go to the daily tab and on the lower lefthand side of the screen you should see 3 sessions showing the start and stop times of each and a toggle with all 3 being on (green). Just toggle off (red) the two that you don't want and you'll be left with the 3rd session only in your data.

AND, welcome and congratulations on your first post!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: Sleepyhead data question

Post by Zzzzzzzzzzz... » Mon Mar 12, 2012 8:04 am

b360155 wrote:I think you're saying you captured 3 sessions within one day and you only want to be looking at the 3rd one only? If that's what you're asking, go to the daily tab and on the lower lefthand side of the screen you should see 3 sessions showing the start and stop times of each and a toggle with all 3 being on (green). Just toggle off (red) the two that you don't want and you'll be left with the 3rd session only in your data.

AND, welcome and congratulations on your first post!

That's the ticket. And thanks much!

Z

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: Sleepyhead data question

Post by SleepingUgly » Mon Mar 12, 2012 8:35 am

Welcome, Z!!
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: Sleepyhead data question

Post by SleepingUgly » Mon Mar 12, 2012 8:50 am

Question about Sleepyhead... Does the leak graph subtract out intentional venting/leak so that theoretically no leaks should be 0? Or does one need to do the math oneself (e.g., Using Swift FX at 10cm, expected leak is around 33l)?
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
Roger2
Posts: 210
Joined: Sat Jul 09, 2011 10:40 am
Location: Englewood, Colorado
Contact:

Re: Sleepyhead data question

Post by Roger2 » Mon Mar 12, 2012 8:56 am

Is there any way to print the graphic screens other than a screen capture?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: My humidifier is the same listed but without the climate control. My software is version 3.11.009

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead data question

Post by Pugsy » Mon Mar 12, 2012 9:00 am

SleepingUgly wrote:Question about Sleepyhead... Does the leak graph subtract out intentional venting/leak so that theoretically no leaks should be 0? Or does one need to do the math oneself (e.g., Using Swift FX at 10cm, expected leak is around 33l)?
SleepyHead reports whatever the machine reports.
ResMed machines report only excess leak...so that is what SH reports. Just like ResScan. We can see possible leak of 0.0. because the machine does the subtracting.

Respironics machines report Total leak (including vent rate) and hidden in the machine is a way to get excess leak only reporting but Mark hasn't got that perfected yet.
So for now..with PR S1 machines we have to do our own math. We won't see on the reports any leak of 0.0...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: Sleepyhead data question

Post by Zzzzzzzzzzz... » Mon Mar 12, 2012 9:01 am

SleepingUgly wrote:Question about Sleepyhead... Does the leak graph subtract out intentional venting/leak so that theoretically no leaks should be 0? Or does one need to do the math oneself (e.g., Using Swift FX at 10cm, expected leak is around 33l)?

Thanks SH!

I've attached last nights data. Anyone care to comment on the leak chart? I'm not that adept just yet. (no rhyme intended)

-Z


Image

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead data question

Post by Pugsy » Mon Mar 12, 2012 9:02 am

Roger2 wrote:Is there any way to print the graphic screens other than a screen capture?
Unsure as to status of the print option in SH. It was buggy and I don't know if Mark fixed it yet.
If you can't get it to work from SleepyHead....then there is no other way except doing a screen shot and then printing the saved screen shot..that I know of anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead data question

Post by Pugsy » Mon Mar 12, 2012 9:08 am

Zzzzzzzzzzz... wrote:Anyone care to comment on the leak chart? I'm not that adept just yet.
You have a ResMed machine. It already subtracts the vent rate for your mask before reporting leak so what you see is all excess leak.
ResMed says that they can handle leaks to 24 L/min...so we try to stay below that amount of leak.

You are new to this therapy..you can easily see that you spent a lot of time near or above that 24 L/min mark.....so you have some significant leaks to work on. Though for a good bit of the night your leak was still under the 24 L/min line.

With time and experience your leak line should stabilize and be reduced. This isn't bad for a first night. I sometimes have almost as bad after nearly 3 years.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
b360155
Posts: 163
Joined: Sat Oct 22, 2011 7:46 pm
Location: North Richland Hills, TX

Re: Sleepyhead data question

Post by b360155 » Mon Mar 12, 2012 9:09 am

Pugsy wrote:
Roger2 wrote:Is there any way to print the graphic screens other than a screen capture?
Unsure as to status of the print option in SH. It was buggy and I don't know if Mark fixed it yet.
If you can't get it to work from SleepyHead....then there is no other way except doing a screen shot and then printing the saved screen shot..that I know of anyway.
Just hit the print button in the lower righthand corner of the screen and it should print off all your graphs whether in the overview mode or the daily mode. Works fine for me.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Using ResScan Version 3.16 Software and Sleepyhead 0.9.2

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: Sleepyhead data question

Post by Hawthorne » Mon Mar 12, 2012 9:10 am

Thanks b360155! I have been wanting to do that myself and never looked into. Now I don't have too!

Thanks again and thanks Z for asking the question!

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: Sleepyhead data question

Post by Zzzzzzzzzzz... » Mon Mar 12, 2012 9:13 am

Pugsy wrote:
Zzzzzzzzzzz... wrote:Anyone care to comment on the leak chart? I'm not that adept just yet.
You have a ResMed machine. It already subtracts the vent rate for your mask before reporting leak so what you see is all excess leak.
ResMed says that they can handle leaks to 24 L/min...so we try to stay below that amount of leak.

You are new to this therapy..you can easily see that you spent a lot of time near or above that 24 L/min mark.....so you have some significant leaks to work on. Though for a good bit of the night your leak was still under the 24 L/min line.

With time and experience your leak line should stabilize and be reduced. This isn't bad for a first night. I sometimes have almost as bad after nearly 3 years.

Thanks P!

Though this is after a week or so... still playing with pressures, EPR's and Ramps... I might keep this set up for awhile and see what happens. Incidentally, would you guess my leaks are from the mask or my mouth? I was a bit congested last night and used some Olba's oil around the intake- seemed to help but maybe not enough. I'm typically a nose breather at night. And didn't wake up with a really dry mouth this time. Finally, I'm on 3mg of Lunesta during my "break in" period. Hope to reduce that to 2mg soon... and ultimately off it completely. But man, it sure helps.

Thanks for the advice!

-Z

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead data question

Post by Pugsy » Mon Mar 12, 2012 9:24 am

Zzzzzzzzzzz... wrote:Incidentally, would you guess my leaks are from the mask or my mouth?
It is really impossible to tell if leaks are from mouth breathing or just nasal pillow movement.
I know that sometimes I do mouth breathe (wake up in the middle of the night with mouth open and dry mouth) but I don't see it on the leak line...no huge increase in leak...so I assumed very short lived mouth breathing.

It is possible that some of those spikes are mouth breathing...after things settle down...especially any congestion issues resolve then re-evaluate the leak line. If you still seem some evidence of prolonged time spent above that 24 L/min mark then you might add a chin strap just to see if it makes any difference. Doesn't have to be store bought...you can do a homemade chin strap to get an idea....something like a wide ace wrap would work.

I don't mouth breathe vary often and I know that sometimes the nasal pillows slip out (I am active sleeper) so sometimes my leak line looks pretty darn ugly and sometimes it looks near perfect. I don't sweat short spikes to large leak territory.
Now if any leak (large or small) is waking you up and disturbing your sleep then it needs to be dealt with because we don't want anything waking us up. Fragmented sleep for any reason will mess with how we feel the next day and thus unwanted.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Zzzzzzzzzzz...
Posts: 331
Joined: Sun Mar 04, 2012 9:19 am

Re: Sleepyhead data question

Post by Zzzzzzzzzzz... » Mon Mar 12, 2012 9:28 am

Pugsy wrote:
Zzzzzzzzzzz... wrote:Incidentally, would you guess my leaks are from the mask or my mouth?
It is really impossible to tell if leaks are from mouth breathing or just nasal pillow movement.
I know that sometimes I do mouth breathe (wake up in the middle of the night with mouth open and dry mouth) but I don't see it on the leak line...no huge increase in leak...so I assumed very short lived mouth breathing.

It is possible that some of those spikes are mouth breathing...after things settle down...especially any congestion issues resolve then re-evaluate the leak line. If you still seem some evidence of prolonged time spent above that 24 L/min mark then you might add a chin strap just to see if it makes any difference. Doesn't have to be store bought...you can do a homemade chin strap to get an idea....something like a wide ace wrap would work.

I don't mouth breathe vary often and I know that sometimes the nasal pillows slip out (I am active sleeper) so sometimes my leak line looks pretty darn ugly and sometimes it looks near perfect. I don't sweat short spikes to large leak territory.
Now if any leak (large or small) is waking you up and disturbing your sleep then it needs to be dealt with because we don't want anything waking us up. Fragmented sleep for any reason will mess with how we feel the next day and thus unwanted.

I hear ya. Thanks again!

-Z