ResMed S9 Grounded Plug?

Post by TalonNYC » Thu Dec 08, 2011 4:09 pm

Does anyone know if ResMed makes a (US) 3-prong plug for the S9 series? All the ones I've seen are 2-prong.

For the electrical engineers among you, I am NOT trying to initiate a "what's really safe" discussion. A grounded plug is safER than a non-grounded one, that's all I'm looking at right now.

Thanks in advance.

User avatar
STL Mark
Posts: 950
Joined: Thu May 06, 2010 1:37 pm
Location: St. Louis, MO - USA

Re: ResMed S9 Grounded Plug?

Post by STL Mark » Thu Dec 08, 2011 4:17 pm

I've only seen 2 prong plugs leading to the brick that converts it to 12 volts.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Machine: S9 VPAP Adapt (eASV) 36037

Mode: ASVAuto, Min EPAP: 4, Max EPAP: 15, Min PS: 3, Max PS: 15, Ramp: Off - Original Titration: 18

User avatar
TalonNYC
Posts: 984
Joined: Tue Mar 22, 2011 7:26 am
Location: New York FREAKIN City
Contact:

Re: ResMed S9 Grounded Plug?

Post by TalonNYC » Thu Dec 08, 2011 4:24 pm

STL Mark wrote:I've only seen 2 prong plugs leading to the brick that converts it to 12 volts.
Same here, for just about every machine. It's odd that something with a water tank (if you have the humidifier) isn't grounded.

User avatar
Lizistired
Posts: 2835
Joined: Tue Dec 14, 2010 10:47 pm
Location: Indiana

Re: ResMed S9 Grounded Plug?

Post by Lizistired » Thu Dec 08, 2011 4:29 pm

Probably because it's plastic. Not much chance of you being electrocuted.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better.

User avatar
TalonNYC
Posts: 984
Joined: Tue Mar 22, 2011 7:26 am
Location: New York FREAKIN City
Contact:

Re: ResMed S9 Grounded Plug?

Post by TalonNYC » Thu Dec 08, 2011 4:38 pm

Lizistired wrote:Probably because it's plastic. Not much chance of you being electrocuted.
unless you have a ClimateLine tube, in which case it's a big heating coil.

User avatar
JeffL
Posts: 507
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: ResMed S9 Grounded Plug?

Post by JeffL » Thu Dec 08, 2011 4:41 pm

Also, once you get past the brick, there's on 24 volts DC going to the unit (I just took a look at the bottom of my brick, it's 24 not 12 volts).

_________________
MachineMask

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: ResMed S9 Grounded Plug?

Post by billbolton » Thu Dec 08, 2011 4:56 pm

TalonNYC wrote:A grounded plug is safER than a non-grounded one
The Resmed mains power adapters are a "double insulated" class of device, which does not require (or benefit from) grounding.

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
TalonNYC
Posts: 984
Joined: Tue Mar 22, 2011 7:26 am
Location: New York FREAKIN City
Contact:

Re: ResMed S9 Grounded Plug?

Post by TalonNYC » Thu Dec 08, 2011 5:07 pm

that's very cool to know! Thanks!

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: ResMed S9 Grounded Plug?

Post by chunkyfrog » Thu Dec 08, 2011 6:10 pm

I couldn't help noticing that most power tools in a workshop/garage are grounded;
while relatively few in the kitchen are--
and that's where the water is.
I always figured that it was a gender equity issue.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: ResMed S9 Grounded Plug?

Post by idamtnboy » Thu Dec 08, 2011 6:26 pm

chunkyfrog wrote:I couldn't help noticing that most power tools in a workshop/garage are grounded;
while relatively few in the kitchen are--
and that's where the water is.
I always figured that it was a gender equity issue.
Bear in mind that all kitchens, since I don't know when, have circuits protected with GFCI's, which don't require the use of a mechanical ground line, although they all have the terminal for them. You can install a GFCI into an old fashioned 2 wire circuit and be just as well protected as the 3 wire circuit. Garages have been required to have GFCI circuits for some time, but not as long as kitchens. Many tools get used who knows where and the likelihood of using a non-GFCI circuit is still pretty high. There are tools though that are double insulated so don't need a ground line. GFCIs can detect an almost unbelievably small current leak and trip.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
Randyp1234
Posts: 245
Joined: Sat Oct 22, 2011 8:13 am
Location: Decatur, Illinois

Re: ResMed S9 Grounded Plug?

Post by Randyp1234 » Thu Dec 08, 2011 6:32 pm

This reminds me of when I picked up all my equipment from Lincare last week. As they were educating me and going through their checklists of what they were supposed to tell me they must've mentioned about 5 times that I must always use a grounded outlet. I looked at the two-prong plug and thought to my self "Seriously?"

I know, I know...if it's midnight and the DME says.......

Randy

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: ResMed S9 Grounded Plug?

Post by idamtnboy » Thu Dec 08, 2011 6:34 pm

TalonNYC wrote:that's very cool to know! Thanks!
The only exposed metal on an S9 power brick are the power terminals. The power in ones of course are exposed only when the cord is not plugged in. The 24 volt output terminals have non-lethal voltage on them. The brick is probably totally waterproof, although I'm sure it's not rated as such. There really is no way to make contact with the power input side of the system without breaking it open. Of course there are fool hardy folks around who will plug in a busted brick, but even then I'll bet overall the chance of getting a lethal shock is near zero.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
TalonNYC
Posts: 984
Joined: Tue Mar 22, 2011 7:26 am
Location: New York FREAKIN City
Contact:

Re: ResMed S9 Grounded Plug?

Post by TalonNYC » Thu Dec 08, 2011 6:37 pm

True, and frankly it's plugged into a grounded, fuse-enabled UPS device. I'm just paranoid about appliances without GFI and/or ground wires.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: ResMed S9 Grounded Plug?

Post by archangle » Sat Dec 10, 2011 12:41 am

Don't worry about it. The double insulated type of design mentioned is very safe if done right. Adding a grounding wire to the design would only add a little safety.

In some ways, a grounded 3 wire design is more dangerous than a double insulated 2 wire design. If you do somehow get in contact with the DC voltage in a double insulated design, you won't necessarily get shocked because there is no path back to ground inside the machine to complete the circuit.

Of course, you shouldn't do anything stupid like open the covers, especially on the power brick, with the power on. Or chew on the climateline hose.

The fact that it's only 12 or 24 V helps, but since it's a wet environment, you might be able to get electrocuted from the low voltage.

Of course, that assumes that you don't get some sort of catastrophic failure in the power brick that lets 120/240 AC leak into the DC line to the machine. Such things usually get a "Hipot" test that checks for this and are carefully designed to isolate AC and DC.

Even with a two wire power supply, a GFCI circuit will make it a little safer. You can get portable GFCI adapters fairly cheaply these days. Just realize that you CAN electrocute yourself even with a GFCI.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
TalonNYC
Posts: 984
Joined: Tue Mar 22, 2011 7:26 am
Location: New York FREAKIN City
Contact:

Re: ResMed S9 Grounded Plug?

Post by TalonNYC » Sat Dec 10, 2011 10:10 am

archangle wrote: You can get portable GFCI adapters fairly cheaply these days. Just realize that you CAN electrocute yourself even with a GFCI.
I do, and I know, but thanks for all the information =)

The device is plugged into a UPS, which is grounded and equipped with a circuit breaker and GFCI of it's own. And I'm still careful, so no worries there.

I also promise not to chew the wires. As much as I am tempted...