Looking for a CMS-50F EXPERT
- Bert_Mathews
- Posts: 480
- Joined: Sun May 06, 2007 9:36 am
- Location: Heber, Utah
- Contact:
Looking for a CMS-50F EXPERT
First: I know an EX-sPERT (def. is am Ex <Unknown> <Spert> is a Drop under Pressure
Two Questions:
(1) what are the little black bars on the reports < Don't equal events??>
(2) Dr is a 40 mile drive so I would like to send him the DATA files?
Should be under c:/program files/spO2/data . . . . .But I'm unable to LOCATE - COPY - any files even HIDDEN / Locked?
Any HELP would be welcome
Also help in uses and Analiese
Two Questions:
(1) what are the little black bars on the reports < Don't equal events??>
(2) Dr is a 40 mile drive so I would like to send him the DATA files?
Should be under c:/program files/spO2/data . . . . .But I'm unable to LOCATE - COPY - any files even HIDDEN / Locked?
Any HELP would be welcome
Also help in uses and Analiese
_________________
| Humidifier | ||||
![]() | ||||
| Additional Comments: CozyHoze Boss™ -- Regenesis™ Pillow -- CPAP Desensitization aromatics..SleepyHead Software | ||||
" If you don't like the HEAT, Don't tickle the Dragons!!!"
Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com

Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com
-
HoseCrusher
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: Looking for a CMS-50F EXPERT
My data is in the folder you are suggesting it should be in. I have had some problems with compatibility mode hiding some of the data files so you may want to check for that.
If you are looking at the summary report, the black bars underneath the O2 and pulse rate graphs are events. They are scored according to the data you input into the tools menu.
I am no "spert," but I believe I have seen that listed on some back country menu's...
If you are looking at the summary report, the black bars underneath the O2 and pulse rate graphs are events. They are scored according to the data you input into the tools menu.
I am no "spert," but I believe I have seen that listed on some back country menu's...
_________________
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
-
SleepyParrot
- Posts: 4
- Joined: Tue Nov 01, 2011 12:28 pm
Re: Looking for a CMS-50F EXPERT
On my Windoze vista system, the CMS-50F files are in C:\Users\SleepyParrot\AppData\Local\VirtualStore\Program Files\SpO2\Data
Cheers!
Cheers!
Re: Looking for a CMS-50F EXPERT
As previously mentioned, the black lines are 'events' that occur during testing. If you are using SP02 Review, when you're looking at the data, on the bottom left corner there is a button with a wrench and screwdriver. If you click that button, that will show you what your parameters are for events (Drop for Event%, etc).
I'm using a Windows 7 computer, and my data is found in C:\Users\My_User_Name\AppData\Local\VirtualStore\Program Files (x86)\SP02\Data.
Having said that, does your doctor use SPO2 Review himself? If they do not, then the data file itself may be of no real value to them, as the file has the SPO2 and Pulse Levels only. It does not have a time stamp so your doctor won't really know what time a particular reading is. I checked one of my data files, and it has something like 19,777 readings for 5.5 hours of readings. That is 1 reading per second.
You may want to consider sending a copy of your Oximetry reports instead. To do this, you will want to download and install a pdf printer. Google "free pdf printer" to find several that can be used for home use. What you will do is print whatever report you and your doctor decide on. When it asks which printer you wish to use, use the pdf printer, give it a good descriptive name and print it. Remember the name & location of this report! You then send this report to your doctor. You can send any of the 4 reports that SPO2 has, but my doctor has always liked the Oximetry report.
Hope this helps!
-Bill
I'm using a Windows 7 computer, and my data is found in C:\Users\My_User_Name\AppData\Local\VirtualStore\Program Files (x86)\SP02\Data.
Having said that, does your doctor use SPO2 Review himself? If they do not, then the data file itself may be of no real value to them, as the file has the SPO2 and Pulse Levels only. It does not have a time stamp so your doctor won't really know what time a particular reading is. I checked one of my data files, and it has something like 19,777 readings for 5.5 hours of readings. That is 1 reading per second.
You may want to consider sending a copy of your Oximetry reports instead. To do this, you will want to download and install a pdf printer. Google "free pdf printer" to find several that can be used for home use. What you will do is print whatever report you and your doctor decide on. When it asks which printer you wish to use, use the pdf printer, give it a good descriptive name and print it. Remember the name & location of this report! You then send this report to your doctor. You can send any of the 4 reports that SPO2 has, but my doctor has always liked the Oximetry report.
Hope this helps!
-Bill
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead |
- Bert_Mathews
- Posts: 480
- Joined: Sun May 06, 2007 9:36 am
- Location: Heber, Utah
- Contact:
Re: Looking for a CMS-50F EXPERT
THANKS to all RESPONDERS - - - - The Dr. does use SP02 software along with some other software ??wm_hess wrote:As previously mentioned, the black lines are 'events' that occur during testing. If you are using SP02 Review, when you're looking at the data, on the bottom left corner there is a button with a wrench and screwdriver. If you click that button, that will show you what your parameters are for events (Drop for Event%, etc).
I'm using a Windows 7 computer, and my data is found in C:\Users\My_User_Name\AppData\Local\VirtualStore\Program Files (x86)\SP02\Data.
Having said that, does your doctor use SPO2 Review himself? If they do not, then the data file itself may be of no real value to them, as the file has the SPO2 and Pulse Levels only. It does not have a time stamp so your doctor won't really know what time a particular reading is. I checked one of my data files, and it has something like 19,777 readings for 5.5 hours of readings. That is 1 reading per second.
You may want to consider sending a copy of your Oximetry reports instead. To do this, you will want to download and install a pdf printer. Google "free pdf printer" to find several that can be used for home use. What you will do is print whatever report you and your doctor decide on. When it asks which printer you wish to use, use the pdf printer, give it a good descriptive name and print it. Remember the name & location of this report! You then send this report to your doctor. You can send any of the 4 reports that SPO2 has, but my doctor has always liked the Oximetry report.
Hope this helps!
-Bill
When I tell my software to LOAD a OLDER Data fill it goes to " C:\Users\SleepyParrot\AppData\Local\VirtualStore\Program Files\SpO2\Data " and all looks good. . . .
BUT when I go in with EXPLORE WINDOW to copy the file ZIPPO NOTHING there????
Can one of you other users try transferring or posting a DATA file . . . . And let me know how you DO IT!
I have a IMAC & Windoz 7 and don't have any problem in making a PDF -BUT- Again the Dr, uses two software programs to evaluate the data < Not sure what other program > or what he does - - I just prefer not driving the 40 miles. but due love living in the country with NOBODY AROUND...
_________________
| Humidifier | ||||
![]() | ||||
| Additional Comments: CozyHoze Boss™ -- Regenesis™ Pillow -- CPAP Desensitization aromatics..SleepyHead Software | ||||
" If you don't like the HEAT, Don't tickle the Dragons!!!"
Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com

Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com
Re: Looking for a CMS-50F EXPERT
1st thing to try is when SPO2 is not running, can you find the folder by clicking on "Computer", then Drive C, then Users, then Sleepyparrot, then AppData etc, or is AppData (or any of the directories you mentioned) not there?
If you do not see the one of the directories, go back to drive c in explorer. When you're looking at Drive C: in explorer, click on organize, then folder & search options. Then click on the view tab. Under the Advanced settings: find Hidden Files & Folders. Make sure Show Hidden files, folders & Drives has a click on it, and Don't show does not.
Once you've done that, try again.
-Bill
If you do not see the one of the directories, go back to drive c in explorer. When you're looking at Drive C: in explorer, click on organize, then folder & search options. Then click on the view tab. Under the Advanced settings: find Hidden Files & Folders. Make sure Show Hidden files, folders & Drives has a click on it, and Don't show does not.
Once you've done that, try again.
-Bill
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead |
Re: Looking for a CMS-50F EXPERT
A couple of suggestions.
a program called Filemon monitors file writes and reads, ( might not work in Windoze 7 or 64 bits).
you could do a file search on your PC and sort by date, find out where the latest ones have gone.
you could print the reports and scan them.
a program called Filemon monitors file writes and reads, ( might not work in Windoze 7 or 64 bits).
you could do a file search on your PC and sort by date, find out where the latest ones have gone.
you could print the reports and scan them.
- Bert_Mathews
- Posts: 480
- Joined: Sun May 06, 2007 9:36 am
- Location: Heber, Utah
- Contact:
Re: Looking for a CMS-50F EXPERT
THANKS :wm_hess wrote:1st thing to try is when SPO2 is not running, can you find the folder by clicking on "Computer", then Drive C, then Users, then Sleepyparrot, then AppData etc, or is AppData (or any of the directories you mentioned) not there?
If you do not see the one of the directories, go back to drive c in explorer. When you're looking at Drive C: in explorer, click on organize, then folder & search options. Then click on the view tab. Under the Advanced settings: find Hidden Files & Folders. Make sure Show Hidden files, folders & Drives has a click on it, and Don't show does not.
Once you've done that, try again.
-Bill
I have tried those tricks even went to DOS Command and did a xcopy of the DATA Directory. . . . Along wit a FULL SEARCH of the drive with EXPLORE / SEARCH / Xtree <If anybody can remember that program> opened the Windows drive in the network and searched with IMAC ???
NO HIDDEN FILES OR FOLDERS or LOCKED FILES???
_________________
| Humidifier | ||||
![]() | ||||
| Additional Comments: CozyHoze Boss™ -- Regenesis™ Pillow -- CPAP Desensitization aromatics..SleepyHead Software | ||||
" If you don't like the HEAT, Don't tickle the Dragons!!!"
Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com

Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com
Re: Looking for a CMS-50F EXPERT
Okay, I have gone to the directory in explorer, and copied the files and emailed them to myself without any issue.
Since you can see them in SPO2 Review let's try this.
Start SPO2 Review and Review an old file. When you are supposed to choose a file, make sure you're looking at ALL files, NOT All Supported Files, SPO2 Review Files or SPO2 Real Time Files.
You should now see that each days' data consists of two files. Right click on one of the files, and select copy. I use Outlook, so I would then go to a new message and paste into the new message. That file then is attached to the message. Go back to SPO2 Review, and copy the other file, and paste it into the message. Now your message should have both files, and you can email that to your doctor.
If you use a web based email, paste them into a temporary folder, and attach them to your email.
Hope this works until you can figure out where they are!
As a note, I am using Windows 7 (64 bit). My data files are in almost the same directory path as yours except where yours says ....\Program Files\...., mine says ...\Program Files (x86)\.... Could you possibly have that directory in your path as well?
As a further note, when I searched the C: drive (in explorer) it did not see the files. I had to do a custom search, and move down to the appdata directory before it would find the files.
-Bill
Since you can see them in SPO2 Review let's try this.
Start SPO2 Review and Review an old file. When you are supposed to choose a file, make sure you're looking at ALL files, NOT All Supported Files, SPO2 Review Files or SPO2 Real Time Files.
You should now see that each days' data consists of two files. Right click on one of the files, and select copy. I use Outlook, so I would then go to a new message and paste into the new message. That file then is attached to the message. Go back to SPO2 Review, and copy the other file, and paste it into the message. Now your message should have both files, and you can email that to your doctor.
If you use a web based email, paste them into a temporary folder, and attach them to your email.
Hope this works until you can figure out where they are!
As a note, I am using Windows 7 (64 bit). My data files are in almost the same directory path as yours except where yours says ....\Program Files\...., mine says ...\Program Files (x86)\.... Could you possibly have that directory in your path as well?
As a further note, when I searched the C: drive (in explorer) it did not see the files. I had to do a custom search, and move down to the appdata directory before it would find the files.
-Bill
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead |
- Bert_Mathews
- Posts: 480
- Joined: Sun May 06, 2007 9:36 am
- Location: Heber, Utah
- Contact:
Re: Looking for a CMS-50F EXPERT
I was NOT able to see the files....wm_hess wrote:Okay, I have gone to the directory in explorer, and copied the files and emailed them to myself without any issue.
-Bill
If you have had SUCCESS look at the fill you have sent and DISCOVER it's JUST a "ShortCut" not the full kb size of the original.....?wm_hess wrote: Since you can see them in SPO2 Review let's try this.
Start SPO2 Review and Review an old file. When you are supposed to choose a file, make sure you're looking at ALL files, NOT All Supported Files, SPO2 Review Files or SPO2 Real Time Files.
You should now see that each days' data consists of two files. Right click on one of the files, and select copy. I use Outlook, so I would then go to a new message and paste into the new message. That file then is attached to the message. Go back to SPO2 Review, and copy the other file, and paste it into the message. Now your message should have both files, and you can email that to your doctor.
-Bill
One computer is a small PC with win7 --- another desk PC 64 bit --- IMAC OSX 10.6wm_hess wrote:As a further note, when I searched the C: drive (in explorer) it did not see the files. I had to do a custom search, and move down to the appdata directory before it would find the files.
-Bill
As a note the data can be downloaded from the CMS-50F more than once onto different computers -BUT- the data fills are still ALLUSIVE????????
STILL open for MORE IDEAS???
_________________
| Humidifier | ||||
![]() | ||||
| Additional Comments: CozyHoze Boss™ -- Regenesis™ Pillow -- CPAP Desensitization aromatics..SleepyHead Software | ||||
" If you don't like the HEAT, Don't tickle the Dragons!!!"
Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com

Be careful about reading health books. You may die of a misprint.
http://www.sharpstones.com
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: Looking for a CMS-50F EXPERT
Have you checked the PROPERTIES/SECURITY tap on that folder to see if you have rights to it?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |
ResScan: http://www.resmed.com/int/assets/html/s ... c=patients
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
Re: Looking for a CMS-50F EXPERT
Sorry Bert but you've got me!
Every thing I've told you was a complete success on my end. I could see the files, they were full files (not shortcuts or paths) I emailed them to myself and was able to point SPO2 Review to the directory where I put the copies, and it read them fine.
Sorry I can't think of any thing further to try.
Every thing I've told you was a complete success on my end. I could see the files, they were full files (not shortcuts or paths) I emailed them to myself and was able to point SPO2 Review to the directory where I put the copies, and it read them fine.
Sorry I can't think of any thing further to try.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead |
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: Looking for a CMS-50F EXPERT
I copied them from the folder wm_ pointed you to, to my desktop. I think you don't have rights to the folder, or they are hidden.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |