Does a reducing 90% avg pressure indicate an improving apnea

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
VVV
Posts: 531
Joined: Mon Mar 14, 2011 11:28 am

Re: Does a reducing 90% avg pressure indicate an improving apnea

Post by VVV » Mon Oct 31, 2011 7:23 pm

DoriC wrote:Usually the therapy works better if your minimum pressure is closer to your 90% pressure, in your case 9-10cms. So a minimum pressure of 8cm might be better, the max is probably OK at 14 should you need the extra pressure. A narrower range also helps with any leak issues.
That is an excellent point about the leaks. If Gary starts out at 4 his mask will likely not leak. If he really needs, for example, 10-14, after he falls asleep, has some apneas, machine ramps to somewhere between 10 and 14, then the mask may start leaking badly.

When I first started out the DME loaned me a machine set 4 - 20. I did not know any better at the time and had a disastrous three weeks with that machine.
.....................................V

LonghornGary
Posts: 15
Joined: Tue Oct 04, 2011 3:10 pm

Re: Does a reducing 90% avg pressure indicate an improving apnea

Post by LonghornGary » Mon Oct 31, 2011 9:14 pm

My ahi has been running between 2,8 to 3.0, how would I go about narrowing the range of my pr sustem one apap?

User avatar
Pugsy
Posts: 65030
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Does a reducing 90% avg pressure indicate an improving apnea

Post by Pugsy » Mon Oct 31, 2011 10:18 pm

Do you have the clinician/Provider manual? It tells you how to get into the clinical setup..there is where you can alter settings.
For instructions on how to get to the clinical menu with pictures...here..http://www.apneaboard.com/PR-System-One ... -Setup.htm

If you do not have the manual you can request it here...Just follow the directions.
Be sure to call it by proper name
PR System One Remstar Auto CPAP with A Flex.
http://www.apneaboard.com/CPAP%20Adjustment.htm

Your reported AHI is fine though. If you wish to alter pressure making a more narrow range...increase the minimum a bit. You may not gain much in improvement in AHI and again you might reduce them even further. No need to raise maximum unless you spend a lot of time at the maximum.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

LonghornGary
Posts: 15
Joined: Tue Oct 04, 2011 3:10 pm

Re: Does a reducing 90% avg pressure indicate an improving apnea

Post by LonghornGary » Tue Nov 01, 2011 4:31 pm

Alright, I've changed my minimum pressure to 8. We'll see if it makes a difference in AHI. I never had a leak problem before. Thanks for the help and suggestions.