Breathing out of "SYNC" with machine?
Hi All:
I have an odd issue with my mask. At times it seems to "hic-cup". Sometimes just once or twice but many times it does it for several minutes. It is almost as if I am not breathing on the same cycle as the CPAP. My mask makes a clicking sound (I think it may be from the valve in the elbow) and then sort of "Jumps" on my face. This happens mostly when I am falling asleep but I have been woken up by it several times as well. I have always been a mouth breather so must wear a fullface. Any thoughts or similar situations ????
I have an odd issue with my mask. At times it seems to "hic-cup". Sometimes just once or twice but many times it does it for several minutes. It is almost as if I am not breathing on the same cycle as the CPAP. My mask makes a clicking sound (I think it may be from the valve in the elbow) and then sort of "Jumps" on my face. This happens mostly when I am falling asleep but I have been woken up by it several times as well. I have always been a mouth breather so must wear a fullface. Any thoughts or similar situations ????
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Breathing out of "SYNC" with machine?
2 thoughts: 1st, have you checked tube for rainout or to see if your humidifier tank is overfilled. If you get droplets of water coming into the hose from the tank, they will hit your mask at high velocity and cause what you are describing. You will be able to see the droplets of water in your hose if this is happening. It happened to me once when I overfilled humidifier reservoir.
Other possibility: if you feel you are breathing out of sync - are you using EPR? If you are, have you tried turning it off to see if you still experience the same sensation?
I use the same mask as you do so I am skeptical that it is a mask issue, unless your anti-asphyxia valve is bad. You can test this by turning machine off and on (with mask off) and watch the flap/valve to see if it opens when machine is off and closes when you turn machine on.
Other possibility: if you feel you are breathing out of sync - are you using EPR? If you are, have you tried turning it off to see if you still experience the same sensation?
I use the same mask as you do so I am skeptical that it is a mask issue, unless your anti-asphyxia valve is bad. You can test this by turning machine off and on (with mask off) and watch the flap/valve to see if it opens when machine is off and closes when you turn machine on.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E |
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
- Snuff-L-Reddy
- Posts: 44
- Joined: Wed Oct 05, 2011 12:01 pm
- Location: Alaska
Re: Breathing out of "SYNC" with machine?
Hi RandyJ
I have the climate line so no rainout. The first night I used the machine I filled it up to the max line, it blew water in to my hose so I now I dont fill to max. I have considered the EPR, but as my machine is still in the rental phase so I do not want to mess with the clinical settings. I have checked the valve and even turned it over--seems to be in good condition. Mask is less than two weeks old. My pressure is only 8 so perhaps the EPR is what is causing this. It really feels like a valve is suddenly closing making a clicking sound and quickly filling my mask with air causing it to puff up on my face.
I have the climate line so no rainout. The first night I used the machine I filled it up to the max line, it blew water in to my hose so I now I dont fill to max. I have considered the EPR, but as my machine is still in the rental phase so I do not want to mess with the clinical settings. I have checked the valve and even turned it over--seems to be in good condition. Mask is less than two weeks old. My pressure is only 8 so perhaps the EPR is what is causing this. It really feels like a valve is suddenly closing making a clicking sound and quickly filling my mask with air causing it to puff up on my face.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Breathing out of "SYNC" with machine?
there is an amount of pressure (I don't know where it is) that you should not use EPR if you are at or below that pressure
you need to find out where that is
your mask could very well be unsealing when the pressure is dropped
contact your DME and get the EPR turned off for now
the quattro does make noise when water collects in the seal - been there - done that myself - check it during the night or first thing in the morning
you need to find out where that is
your mask could very well be unsealing when the pressure is dropped
contact your DME and get the EPR turned off for now
the quattro does make noise when water collects in the seal - been there - done that myself - check it during the night or first thing in the morning
Re: Breathing out of "SYNC" with machine?
Weird problem you're having. I'm a Respironics user and on my machine, the A-flex (same as EPR) is a patient comfort setting and can be easily changed w/o going into the clinician's menu. So perhaps it's accessible to you.Snuff-L-Reddy wrote: I have considered the EPR, but as my machine is still in the rental phase so I do not want to mess with the clinical settings
Jamis
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: Breathing out of "SYNC" with machine?
8 is a low pressure, so you probably do not need EPR. Try turning EPR off.
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
Re: Breathing out of "SYNC" with machine?
If EPR is set to "Patient" you should be able to turn it off without going into Clinician's Menu.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E |
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
Re: Breathing out of "SYNC" with machine?
And if EPR is NOT set to Patient, you should politely, but insistently ask the DME why they locked you out of this important patient comfort setting.RandyJ wrote:If EPR is set to "Patient" you should be able to turn it off without going into Clinician's Menu.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: Breathing out of "SYNC" with machine?
Robysue is 100% right... in the event that you are locked out of changing EPR setting, you should still be able to see whether EPR is set to "Off" or to "1," "2" or "3." If you want to change it, it is very easy to do. Going into the Clinician's Menu does not void your warranty or violate Federal law or any of the other crazy things you may hear. You are renting to own the machine, and you can make changes to settings if you want.
To see a video on going into the Clinician's Menu etc (it's really child's play), go to:
http://www.cpaplibrary.com/machines.html
and watch the first video for the S9 series. Clinician's Menu explanation starts at 11'13".
To see a video on going into the Clinician's Menu etc (it's really child's play), go to:
http://www.cpaplibrary.com/machines.html
and watch the first video for the S9 series. Clinician's Menu explanation starts at 11'13".
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup & Travel Machines: PR System One Bipap Auto, S9 VPAP Auto, S9 Autoset, Oximeter CMS-50E |
Last edited by RandyJ on Tue Oct 11, 2011 11:13 am, edited 1 time in total.
Diagnosed March 2011, using APAP 14 - 16.5 cm, AFlex+ 2
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
Alt masks Swift FX pillows, Mirage FX nasal mask, Mirage Quattro full face mask
- Perrybucsdad
- Posts: 834
- Joined: Mon Sep 12, 2011 7:09 am
- Location: Northeast Ohio
Re: Breathing out of "SYNC" with machine?
robysue wrote:And if EPR is NOT set to Patient, you should politely, but insistently ask the DME why they locked you out of this important patient comfort setting.RandyJ wrote:If EPR is set to "Patient" you should be able to turn it off without going into Clinician's Menu.
AMEN!!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ResScan v5.9; Sleepyhead v1.0.0-beta |
- Snuff-L-Reddy
- Posts: 44
- Joined: Wed Oct 05, 2011 12:01 pm
- Location: Alaska
Re: Breathing out of "SYNC" with machine?
Well last night as far as I am aware this only happened 3 times so maybe I am learning to breathe in unison with my machine. But I guess I will have to lose my paranoia and access the clinician menu and change the EPR. Hopefully DME does not read this.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Breathing out of "SYNC" with machine?
Whether it's a rented machine or one you or your insurance paid for - it's YOURS!!! It's also YOUR therapy, YOUR body, so do whatever it takes to make this experience as comfortable as possible. It's of no business or concern of the DME how you manage your therapy, just that you are compliant.
IMHO, you need to learn to relax more - don't try to 'breathe with the machine' otherwise you'll hyperventilate and get anxious and panicky - I did - at first.
I'm still a relative newbie at CPAP, but what I still do after a year is this -
When I get into bed, I set up my TV to watch a dvr'd episode of Jeopardy (any non-stressful half hour program will work). Then I put on chinstrap and mask, hit the Play button and my SmartStart on the CPAP kicks in. I spend the 20 or so minutes watching Jeopardy adjusting my gear so it's snug, but not tight, comfy and by the time the program is over, I'm not even conscious of my breathing and can lay down and go to sleep immediately.
When we go camping, Jeopardy isn't an option, so I read my Nook for a few minutes, all the while checking and adjusting my gear (always with the machine running, of course) and by the time I'm starting to fall asleep - again, my breathing in not a conscious effort.
This might work for you if it's anxiety causing your issues.
Cheers,
xena
IMHO, you need to learn to relax more - don't try to 'breathe with the machine' otherwise you'll hyperventilate and get anxious and panicky - I did - at first.
I'm still a relative newbie at CPAP, but what I still do after a year is this -
When I get into bed, I set up my TV to watch a dvr'd episode of Jeopardy (any non-stressful half hour program will work). Then I put on chinstrap and mask, hit the Play button and my SmartStart on the CPAP kicks in. I spend the 20 or so minutes watching Jeopardy adjusting my gear so it's snug, but not tight, comfy and by the time the program is over, I'm not even conscious of my breathing and can lay down and go to sleep immediately.
When we go camping, Jeopardy isn't an option, so I read my Nook for a few minutes, all the while checking and adjusting my gear (always with the machine running, of course) and by the time I'm starting to fall asleep - again, my breathing in not a conscious effort.
This might work for you if it's anxiety causing your issues.
Cheers,
xena
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-) |
- Snuff-L-Reddy
- Posts: 44
- Joined: Wed Oct 05, 2011 12:01 pm
- Location: Alaska
Re: Breathing out of "SYNC" with machine?
Ok I accessed the clinicians menu ( Whew !!!! no alarms, police raids, or self destruct sequence ) I turned EPR and Climate control both to patient. I then set EPR to 2cmH2O. Still had some clicking but not quite as noticeable. I will lower it again tonight to see if it gets even better. I am even more happy about being able to adjust the humidity manually. I have tried different Temp. settings on Auto but would still wake up with an extremely dry mouth. After yesterdays manual humidity adjustment my mouth was not dry this morning, my wife even noticed this . Thanks to all for convincing me to take the "plunge" into the Clinicians menu as it is sure to make me more comfortable. Why were these "comfort settings" not available in the first place ?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Last edited by Snuff-L-Reddy on Wed Oct 12, 2011 10:20 am, edited 1 time in total.
Re: Breathing out of "SYNC" with machine?
Because the RT at your DME in his/her infinite lack of wisdom locked you out. And why did he/she lock you out? He/She is probably a card carrying member of the CPAP Police and believes that the less patients know about their own therapy the better.Snuff-L-Reddy wrote:Why were these "comfort settings" not available in the first place ?
Even though you've fixed the problem, I think you should complain (and loudly) to the DME about the fact that you were locked out of patient comfort settings. I'd play dumb when talking to them: Tell them you've been told to experiment with different humidity settings and different EPR settings by the doc's office to help solve the dry mouth problem and that your owners manual says you should be able to do these things and ..... You get the picture.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: Breathing out of "SYNC" with machine?
I love Jeopardy but it totally STRESSES me out when there's a category I don't know much about like Greek mythology (somehow, my education missed that topic, but thanks to Percy Jackson books my 10 year old ACES that category) or sports.When I get into bed, I set up my TV to watch a dvr'd episode of Jeopardy (any non-stressful half hour program will work).
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm