How to Check Accumulated Hours on a ResMed S9 AutoSet

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by StevenXXXX » Mon Sep 12, 2011 2:01 pm

How does one check the accumulated hours (Therapy Hours & Hours it was on) on a ResMed S9 AutoSet ?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

User avatar
Pugsy
Posts: 65268
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by Pugsy » Mon Sep 12, 2011 2:43 pm

Uncle_Bob has the link for clinician manual in his signature line. I don't know the steps to check therapy hours and blower hours but I bet it is in the clinician manual.

memberlist.php?mode=viewprofile&u=38643

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by StevenXXXX » Mon Sep 12, 2011 2:55 pm

Pugsy wrote: I don't know the steps to check therapy hours and blower hours but I bet it is in the clinician manual.
I thought that too, but it is not in there. I have already checked that entire manual.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

User avatar
Pugsy
Posts: 65268
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by Pugsy » Mon Sep 12, 2011 3:03 pm

StevenXXXX wrote:I thought that too, but it is not in there. I have already checked that entire manual.
Well that is a surprise.. I thought for sure it would. I saw the instructions somewhere...awhile back..don't remember where. I will see if I can jar my memory.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by Uncle_Bob » Mon Sep 12, 2011 3:12 pm

Advance Info menu (hold down both rectangular buttons) - as described in the clincial menu at the bottom of page 15
This shows the sleep report - shown in the clinical info menu at the bottom of page 16

Shows
Days used
Hours used
Average hours
etc

is this not what you are looking for?

StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by StevenXXXX » Mon Sep 12, 2011 3:28 pm

Uncle_Bob wrote:Advance Info menu (hold down both rectangular buttons) - as described in the clincial menu at the bottom of page 15
This shows the sleep report - shown in the clinical info menu at the bottom of page 16

Shows
Days used
Hours used
Average hours
etc

is this not what you are looking for?
Not sure that this is the accumulated hours on the machine from Day 1?
(i.e. a brand new machine never used - not even for Demo - would show 0 hours).

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by Uncle_Bob » Mon Sep 12, 2011 4:25 pm

StevenXXXX wrote:
Uncle_Bob wrote:Advance Info menu (hold down both rectangular buttons) - as described in the clincial menu at the bottom of page 15
This shows the sleep report - shown in the clinical info menu at the bottom of page 16

Shows
Days used
Hours used
Average hours
etc

is this not what you are looking for?
Not sure that this is the accumulated hours on the machine from Day 1?
(i.e. a brand new machine never used - not even for Demo - would show 0 hours).
Thats my understanding although i know on the S8 you could reset those values back to zero via the menus so its not as reliable as a car odometer.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by archangle » Mon Sep 12, 2011 5:13 pm

PRS1 machines report both "blower hours" and "machine hours." Blower hours can be reset, machine hours can't.

I don't know if ResMed has similar numbers recorded.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by apneawho » Mon Sep 12, 2011 5:35 pm

On my home menu, using the info button on right and then the dial on the right, I can change the time period from 1 day to 1 week to 1 month to 6 months to 1 year. That is where I can see the total accumulated days and hours used. Is that what you are asking for? I can also see this info in Rescan.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

StevenXXXX
Posts: 165
Joined: Tue Jan 04, 2011 4:22 pm

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by StevenXXXX » Tue Sep 13, 2011 7:26 am

apneawho wrote:On my home menu, using the info button on right and then the dial on the right, I can change the time period from 1 day to 1 week to 1 month to 6 months to 1 year. That is where I can see the total accumulated days and hours used. Is that what you are asking for? I can also see this info in Rescan.
So, if you had the machine for more than 1 year, you could not determine what the accumulated to date hours were (say 3 years)?

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed ResScan Software Version 3.16 Do NOT use either Ramp or EPR

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by apneawho » Tue Sep 13, 2011 7:40 am

[So, if you had the machine for more than 1 year, you could not determine what the accumulated to date hours were (say 3 years)?[/quote]

I will have to check that. I think you can still see total hours logged on machine on same screen I am talking about. I think it is the sleep report screen.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by archangle » Tue Sep 13, 2011 4:46 pm

The clinician's manual says "Allows the clinician to erase all data stored in the unit and SD
card (except for machine hours). Settings, date and time are not affected." for the "erase data" function. That sounds like there's a non-resettable "odometer" machine hours counter.

I didn't see how to access "machine hours" in the menu from the manual, though.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: How to Check Accumulated Hours on a ResMed S9 AutoSet

Post by chunkyfrog » Tue Sep 13, 2011 4:58 pm

OK, first press the 'i' button; you will see the sleep report screen, then hold down both the silver rectangular buttons
until the little wrench appears on the bottom line, then turn the knob to the right of those silver buttons ONE click--you
will see a different screen; then you the turn the knob until the 'run time' line appears. This is it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her