No Answers Yet For Problems With ASV

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: An answer for low self initinated breathing on ASV??

Post by avi123 » Sun Jul 03, 2011 3:44 pm

HoseCrusher wrote:Avi, it is interesting that your O2 desaturation events work out to 15.3 per hour and your AHI is 13.2 per hour.

An average of 94.7% is good, but you do have some desaturations below 90%. Fortunately, you don't spend much time below 90%.

It would be interesting to check if your major leaks occurred during the various desaturations below 90%.

Thanks Crusher, I plan to synchronize the CMS-50E graph with the ResScan's graphs and find out. Last night, I managed to sleep non-stop for five hours. Usually, I sleep for three hours and then get up to urinate, and repeat this sequence. While going to the bathroom I stop the CPAP and take the mask off. Getting up and falling back to sleep louse- up the ResScans' graphs of leaks, events, etc., and I bet the SpO2 also.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: An answer for low self initinated breathing on ASV??

Post by HoseCrusher » Sun Jul 03, 2011 4:13 pm

Avi, when you get up in the middle of the night it will be obvious on the oximeter data. This may serve as an aid in synchronizing your data.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: An answer for low self initinated breathing on ASV??

Post by dsm » Sun Jul 03, 2011 4:48 pm

Mr Bill wrote:
Paper_Nanny wrote: MSD talked to Bob, the Respironics Rep for much of the state of Oregon. Bob did in fact say I should be on a straight BiPAP, with no ASV. Why does this make sense? According to Bob, the the ASV is causing me to hyperventilate. That causes the CO2 level in my blood to go down, supressing the urge to breathe. When the CO2 level goes up, then I breathe. But, the ASV thinks I should be breathing sooner than that and triggers a breath. Thus, the PTB rate goes down after the first few hours of sleep.

Why does the ASV do that? Doesn't the algorithm figured out how fast I was breathing and time the breaths accordingly? According to Bob, that isn't quite right. The algorithm was deisgned for people with Cheyne-Stokes Respiration and CSA secondary to the use of opiods. It works great for people with that sort of SDB, but not for people like me.
OK, I am going to take a stab at explaining this from the existential participant point of view. This is the view where I try to extrapolate my existential relationship with my own ASV into the patterns I see you having.

First, I think ASV's like the patient to have a regular well spaced breathing pattern because they are designed to break up patterns of periodic breathing. So, if you have an erratic pattern of breaths due to your MS then the ASV will keep intervening to regularize the pattern.

Second, Although the patient initiated breath graph looks horrible under full ASV control, I think probably, its not as bad as it appears. The ASV decides what your "breath timing slots" should be after measuring your average breath rate and tidal volume and probably some other timing and flow rate factors and I think it probably does some sort of predictive filtering sort of like a moving average to guess when the next breath should be. Just because it nudges you to start a breath does not mean it pushes in the whole breath. I've noticed with mine that it starts the breath but it does not follow through come hell or high water unless, unless, I persist in not breathing, then the pressure ramps up. Now pressure can ramp up both to open the airway and to inflate lung. I'm not at all informed how much more pressure differential is needed to inflate the lungs if we just lay there like a dead fish. But my RT says an ASV will actually do it if set to have a backup rate. So the patient initiated breath graph is probably a moving average. as the percentage drops with each "missed time slot". You may actually be doing 90% of the work even if you are doing 0% of the breath initiation. However, estimating that percentage of effort is probably not very easy or accurate outside of a lab. So, you see the Respironics rep is saying 'its not desirable but probably not harmful'. I suppose the only way to know for sure would be if the polysonmography could actually measure your exertion during sleep.

Third, it may be that you are having alpha wave intrusion (mild arousals while unconscious during sleep) because of these breath initiations which are making you feel less rested even though your AHI is looking pretty good.

Fourth, possibly you feel better with the recently tried BIPAP mode because you got to set your own breathing rhythm but the ASV logs every deviation from a steady train of pulses as hyponeas or clear airway apneas.

Fifth, its probably a good thing in the long run that the ASV tries to enforce a steady pattern of breaths no matter what causes breathing to be erratic. It took me months to get used to breathing 18 bpm when my natural rate is closer to 8. But I have gotten there and I keep sleeping better and better.

Now to the answer phase. I have to confess I wish I could sleep deeper most of the time. It has taken me 4 months to get over 4.5 hours a night and now my average after 7 months (just for the last 3 months) is 6.5 hours. But I can tell that I need to sleep deeper and longer. Its too easy to get woke up. I'll hazard that the ASV can treat your symptoms better than a BIPAP if the ASV could be adjusted to the widest possible leeway in breath variability. I believe somebody already mentioned that the inspiration time had to be less than 3 seconds to make the ASV happy. I wonder if setting your Inspiration time to the max would help? Also, since you have an average BPM of around 10 maybe setting a backup rate at 8 (ask Dr and RT if this is OK). On auto, maybe the unit is trying to speed you up to some default value. Finally, the rise time can be set higher, if I understand correctly, rise time is how quickly the ASV tries to bump up the pressure for an individual breath. Maybe if it is constrained to rise slowly, it will be less disturbing to your sleep as it tries to correct you to an even pattern. Lastly, I think you may start feeling more refreshed over time. I know my own experience after the first rush of getting sleep was to then realize for several months how much I needed it and how little I was getting. Maybe with time your sleeping mind will start accepting the ASV's nudges and not be aroused and then you will start really sleeping.

So, I'm not a medical doctor but I think you need the ASV in the long haul.

Edited for truth in advertising...
Mr Bill,

Enjoyed your evaluation - it is a good thing to think through what might be happening as it connects you to your therapy very closely. In the latter large para you suggest making changes to rate & other settings (rise time etc; ), the challenge in doing such changes is in knowing if they improved anything. One would hope the nightly data is granular enough to spell out changes such that one can tune a machine but that is not always the case.

The only sure way to make changes & know what impact they have is via the Dr or RT who hopefully will prevent a user from making changes to settings that are counter productive. Luckily we are moving closer to the machines themselves having so few parameters to change that this may become much less of an issue in the future.

The ASV machine really is best at adjusting respiration on people who have predictable patterns of periodic breathing. Those with unpredictable patterns may not benefit at all & this is what I believe NotMuffy is pointing out & raising the flag on. e.g. In Cheyennes-Stokes respiration, the periodic breathing pattern has a predictable cycle of waxing & waning & the machine's algorithm is tuned to looking for that pattern and preempting it. Erratic breathing can counter the ASV algorithm. Some forms of SDB 'illness' will cause erratic breathing & bi-level is often the better choice.

Cheers

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: An answer for low self initinated breathing on ASV??

Post by dsm » Sun Jul 03, 2011 5:47 pm

Mr Bill wrote:
<snip>

Fifth, its probably a good thing in the long run that the ASV tries to enforce a steady pattern of breaths no matter what causes breathing to be erratic. It took me months to get used to breathing 18 bpm when my natural rate is closer to 8. But I have gotten there and I keep sleeping better and better.

<snip>
Mr Bill, Re this point, I agree that some of us may well be in the habit of slow deep breathing at a very slow rate. When starting on my 1st Bilevel, had the backup rate down to 6 because of this. In fact, this habit is probably why the Resmed Vpap Adapt SV caused self so much difficulty at the start as it really wasn't going to let me run at that typical slow rate. Having used this 'breathing policeman' for years now, my breathing is typically around the 11-16 BPM rate and, it is no longer me battling the machine. It seems it trained me to change my rate.

Cheers

DSM
#2 by => my
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Problematic "initinated" Removed

Post by Paper_Nanny » Sun Jul 03, 2011 7:29 pm

NotMuffy wrote:"That said II", the "avoid" caution was for "Restrictive or neuromuscular disease".
I was reading it as "restrictive thoracic disease or restrictive neuromuscular disease."
NotMuffy wrote:"IMHO", the low baseline cannot be attributed solely to altitude.
I agree, especially since any time I have checked my oxygen saturation during the day, it has, with rare exception, been over 95%.
NotMuffy wrote:If ASV is aggressively attacking (it's using breaths as well as IPAP), then the only difference between the two is event identification.
Meaning that when the machine is running in ASV mode, it is identifying events as a non-patient triggered breath and when running in BiPAP mode, it is identifying events as apneas and hypopneas?

Note to self:content of other people's notes to self is NOT sufficient reason for freaking out

edit to correct formatting of quote

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: An answer for low self initiated breathing on ASV??

Post by Paper_Nanny » Sun Jul 03, 2011 7:42 pm

dsm wrote:
Mr Bill wrote:
<snip>

Fifth, its probably a good thing in the long run that the ASV tries to enforce a steady pattern of breaths no matter what causes breathing to be erratic. It took me months to get used to breathing 18 bpm when my natural rate is closer to 8. But I have gotten there and I keep sleeping better and better.

<snip>
Mr Bill, Re this point, I agree that some of us may well be in the habit of slow deep breathing at a very slow rate. When starting on my 1st Bilevel, had the backup rate down to 6 because of this. In fact, this habit is probably why the Resmed Vpap Adapt SV caused self so much difficulty at the start as it really wasn't going to let me run at that typical slow rate. Having used this 'breathing policeman' for years now, my breathing is typically around the 11-16 BPM rate and, it is no longer me battling the machine. It seems it trained me to change my rate.
I understand the part about it being a good thing to have a steady pattern of breathing. And I understand the part about it being not so good to battle with the machine all night over the breath rate. What I don't understand is what is the problem with having a low BPM rate. Can someone clarify?

Deborah

edit to correct spelling in subject line

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
Last edited by Paper_Nanny on Mon Jul 04, 2011 1:02 pm, edited 1 time in total.
PR System One BiPAP Auto SV Advanced

ozij
Posts: 10537
Joined: Fri Mar 18, 2005 11:52 pm

Re: No Answers Yet For Problems With ASV

Post by ozij » Sun Jul 03, 2011 8:01 pm

Slow BPM means long pauses between breaths. The problem with that is what happens to your oxygen saturation during those long pauses. The algorithm obviously assumes the if you're on the other end of the machine, you're not doing too well with that small nmber of breaths pre minutes, and it takes it upon itself to nudge your breathing along faster.

If you're used to breathing slowly, and that stern task master is nudgine you - or rushing you -along... well, you do have to get used to accepting it without a fight - and that may take time.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Sun Jul 03, 2011 9:07 pm

More "interesting" graphs... I am entertaining the idea of bagging the ASV and then myself.

Meh. I probably just need to watch Charlie the Unicorn for awhile.

https://www.youtube.com/watch?v=CsGYh8AacgY

Maybe for a very long while...

Image

Image

edit to put graphs in chronological order

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
Last edited by Paper_Nanny on Sun Jul 03, 2011 10:10 pm, edited 2 times in total.
PR System One BiPAP Auto SV Advanced

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: No Answers Yet For Problems With ASV

Post by dsm » Sun Jul 03, 2011 9:07 pm

As Ozij has implied, slow breathing is fine if the SpO2 level isn't causing desats. From my experience it was the slow breathing that would upset my wife. She would push me when she couldn't hear me breathing. I used to wonder why she did it as to me all was fine, except there was the falling asleep at work problem.

Moving to a bilevel then an ASV has got rid of the daytime sleepiness. CPAP never worked for me. Dealing with the daytime sleepiness was critical to the type of work being done & a very strong desire to remain employed.

My understanding of ASV (certainly the Vpap Adapt SV) is that it starts off the night seeing the user breathing at a slightly elevated BPM rate & tracks this (in a 3 min window). Then as the user relaxes & goes into deeper sleep the machine tracks the change. Normally breathing rate does tend to slow (but can increase during REM sleep). If the rate combined with the minute ventilation drops too far the machine will increase the Ipap pressure (that is widen the PS) and start tracking to the builtin fallback rate of 15 BPM. I believe the Bipap AutoSV will do the same. However, there was a report that if the Bipap AutoSV is set to auto BPM, the rise time can become too slow. (see http://scholarworks.boisestate.edu/cgi/ ... s%20asv%22 - last paragraph). That may be sorted out in the latest models but it did reflect my own experience when using the original Bipap AutoSV, which by and large was a wonderful machine (still have it).

As hinted at to Mr Bill, adjusting risetimes is a very tricky matter & best done with *expert* clinical guidance. Just as an aside IIRC risetime can only be adjusted on the Bipap AutoSV when in BiLevel mode. I think it is a function of the algorithm when in ASV mode.

For you Deborah, I see the issue is do you suffer from 'erratic breathing' patterns that the ASV is not best at addressing. I have no real idea at this point if that is the case - NotMuffy has pointed out what info he needs to help determine that.

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: No Answers Yet For Problems With ASV

Post by dsm » Sun Jul 03, 2011 9:14 pm

Deborah,

Re the charts for 1 & 2nd July. Which night felt better ? (was it the night of the 1st ?)

Also pls remind me what mode the machine is in (ASV or BiPap ?).

Cheers

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Sun Jul 03, 2011 10:23 pm

dsm wrote:As hinted at to Mr Bill, adjusting risetimes is a very tricky matter & best done with *expert* clinical guidance.
Thanks... I will hold off on messing with that one until such time as I feel I am being expertly guided to do so.
dsm wrote:For you Deborah, I see the issue is do you suffer from 'erratic breathing' patterns that the ASV is not best at addressing.
And what keeps thumping around in my head is, "What if I do have erratic breathing patterns that the ASV is not best at addressing? Then what? What happens after that?" Not only what happens in terms of my SDB, but what happens next in terms of my life and disease progression? I know no one on this plane of existence has the answers to that, but sometimes, I think it would make things a little easier if someone did.

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: No Answers Yet For Problems With ASV

Post by Paper_Nanny » Mon Jul 04, 2011 12:43 am

dsm wrote:Re the charts for 1 & 2nd July. Which night felt better ? (was it the night of the 1st ?)
That's a difficult question to answer. On both nights, I felt like I was barely sleeping. I also remember messing with my mask a lot during the night because I thought it was leaking. I was surprised to see when looking at the data that no, there weren't horrendous leaks. On both nights, I took the mask off after about six hours of sleep and then went back to sleep for awhile. Six hours would be a hideously short night for me.

I think I generally felt more restless the night of the 2 July. And I felt worse the morning of 3 July than the morning of 2 July. I don't know if that is soley because of the lack of restful sleep that night or if it was a cumulative effect. I remember dreaming more on 1 July.
dsm wrote:Also pls remind me what mode the machine is in (ASV or BiPap ?).
It is set in BiPAP Mode 12/8.

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

dsm-guested

Re: No Answers Yet For Problems With ASV

Post by dsm-guested » Mon Jul 04, 2011 2:02 am

Paper_Nanny wrote:
dsm wrote:Re the charts for 1 & 2nd July. Which night felt better ? (was it the night of the 1st ?)
That's a difficult question to answer. On both nights, I felt like I was barely sleeping. I also remember messing with my mask a lot during the night because I thought it was leaking. I was surprised to see when looking at the data that no, there weren't horrendous leaks. On both nights, I took the mask off after about six hours of sleep and then went back to sleep for awhile. Six hours would be a hideously short night for me.

I think I generally felt more restless the night of the 2 July. And I felt worse the morning of 3 July than the morning of 2 July. I don't know if that is soley because of the lack of restful sleep that night or if it was a cumulative effect. I remember dreaming more on 1 July.
dsm wrote:Also pls remind me what mode the machine is in (ASV or BiPap ?).
It is set in BiPAP Mode 12/8.

Deborah
Deborah

How much do you think anxiety is intruding into your xpap activities/experiments and sleep ? ( it can play a significant spoiler role)
Are you aware of higher than normal levels of anxiety at the moment ?

Cheers

DSM

User avatar
NotMuffy
Posts: 962
Joined: Wed Sep 16, 2009 6:56 am
Location: Dunno. GPS is dead.

Much Better!

Post by NotMuffy » Mon Jul 04, 2011 4:09 am

Paper_Nanny wrote:
NotMuffy wrote:If ASV is aggressively attacking (it's using breaths as well as IPAP), then the only difference between the two is event identification.
Meaning that when the machine is running in ASV mode, it is identifying events as a non-patient triggered breath and when running in BiPAP mode, it is identifying events as apneas and hypopneas?
I might offer that when the machine is running in ASV mode, it is not necessarily identifying events. The multi-factorial rate determining portion of the algorithm establishes your RR Target at ~12, so if you haven't had a breath in 5 seconds (5 seconds short of an "event"), ready or not, here one comes. If it successfully delivers a breath, no event. If it sorta delivers a breath, then it may register as a hypopnea. It if can't, apnea. Consequently, it may be that while you could have no events registered, if RR is going crazy, then sleep may be highly disrupted. The shortfall of only monitoring respiratory parameters. I mean, you might not even be asleep during some of those periods.
Paper_Nanny wrote:Note to self:content of other people's notes to self is NOT sufficient reason for freaking out
What, we need the PFT results for the next discussion point.
"Don't Blame Me...You Took the Red Pill..."

User avatar
NotMuffy
Posts: 962
Joined: Wed Sep 16, 2009 6:56 am
Location: Dunno. GPS is dead.

I'll Get It Right Yet...

Post by NotMuffy » Mon Jul 04, 2011 5:46 am

OK, for a sneak preview, Magdy Younes offered an explanation for chain central events in "The Loop":

Image

Keeping in mind that breathing is a dynamic process, at any moment in time you travel along one of the 3 vectors during breath cycle. Forget about apneas and hypopneas, only consider RR change (like a respiratory event is when your RR is less than 6).

Further, you can alter the position of each vector. For instance, start breathing on aggressive mechanical ventilation (represented by PAV (Proportional Assist Ventilation)) and you left-shift the Controller Gain vector.

CompSAS patients probably have a left shift of the Feedback Gain (have "increased chemoresponsiveness"); that is the root of all CompSAS evil, and a reduced FRC will additionally shift the Plant Gain

Image

So if you also have a reduced FRC, you have the Perfect Storm, and why you can take a normal AHI and blow it up into the 30s once you try to "fix" it.

See how easy this is?

Let's do World Hunger next.
"Don't Blame Me...You Took the Red Pill..."