resmed m plus series exhale pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
dwilken
Posts: 5
Joined: Wed May 18, 2011 3:03 pm

resmed m plus series exhale pressure

Post by dwilken » Wed May 18, 2011 3:07 pm

folks. things were fine for a while. now feels like i have to push to exhale. i know i should go back "in". but that's not an option. lost job & insurance. i found the posts for getting into the clinicians mode, but i am still unsure what setting i'm looking for. once again, the inhale is fine, the exhale is hard. my pressure is 8. any help will be greatly appreciated. i'm a little over whelmed and frustrated-dave

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: resmed m plus series exhale pressure

Post by DoriC » Wed May 18, 2011 4:37 pm

Do you have cflex on your machine? That will give you some exhale relief, it has 3 settings to choose from, 1 being the least and 3 the most relief. Your machine is not data-capable so you can't see what's happening.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
Pugsy
Posts: 65014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: resmed m plus series exhale pressure

Post by Pugsy » Wed May 18, 2011 5:02 pm

http://www.apneaboard.com/CPAP%20Adjustment.htm

Check the above link as it has direct download for the M series Plus machine's clinician manual which should list your options and how to get to them. If it has CFlex that will be the exhale relief. You would just need to try each setting to see which is more comfortable.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
gasp
Posts: 2095
Joined: Fri Apr 13, 2007 4:20 pm
Location: U. S.

Re: resmed m plus series exhale pressure

Post by gasp » Wed May 18, 2011 5:50 pm

Forgive me if I missed this info somewhere, but which machine do you have?

If you have a Respironics machine with 3 buttons on top that glow a blue light, see that one is for the Flex function. Pushing the Flex button moves through 3 available settings. If you accidentally push this button during the night you may change to another Flex number.

1 = least exhalation relief
2 = medium exhalation relief
3 = maximum exhalation relief

I keep mine on 3.

_________________
MaskHumidifier
Additional Comments: Pursleep, Padacheek, Regenesis™ Pillow, Neti Pot, Travel Buddy hose hanger

dwilken
Posts: 5
Joined: Wed May 18, 2011 3:03 pm

Re: resmed m plus series exhale pressure

Post by dwilken » Tue Jul 26, 2011 7:48 am

yea, i have my c flex set on 3 but i still feel like i have to push to exhale.-any suggestions

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: resmed m plus series exhale pressure

Post by rested gal » Tue Jul 26, 2011 12:39 pm

The "push" to exhale that you feel... is that feeling happening the most as you are nearing the end of breathing out each time?
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435