ResScan is shutting down...
ResScan often hangs on exit. Still happens with 3.14. Has anyone figured out how to make this not happen?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
Re: ResScan is shutting down...
I find ResScan slow to close and exit but haven't had a problem w/it "hanging". Of course, I haven't updated past 3.10 yet. It just hasn't seemed worth the hassle. What's new between 3.10 and 3.14, anything worthwhile or noticeable??
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
Not sure what exactly is new, but I assume the bleeding edge includes some bug fixes.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
Re: ResScan is shutting down...
Is it Resscan, or CRWS.exe or devicescan2.exe, that is hanging? I find that after downloading data into Resscan one of those two other programs will not close on its own. There is a way to make them close but I don't know right off hand how to do it. I use to have to do it for some other program years ago and I've been thinking I ought to find that answer again. It is irritating.Breathe Jimbo wrote:ResScan often hangs on exit. Still happens with 3.14. Has anyone figured out how to make this not happen?
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
The program hangs at the "ResScan is shutting down..." window. The CRWS icon is usually, if not always, in my taskbar at the time. Whatever the cause, I need to stop ResScan in the Task Manager to close the program. ResScan is the only program listed on the program tab at the time.
According to Problem Reports and Solutions, it appears that the problem is the ResWare Shell Launcher, but that may be only what I am shutting down, rather than the cause of the problem. These are the details provided by PR&S:
Product
ResWare Shell Launcher
Problem
Stopped working
Date
4/10/2011 4:55 PM
Status
Not Reported
Problem signature
Problem Event Name: APPCRASH
Application Name: ResScan.exe
Application Version: 0.2.0.60
Application Timestamp: 00000000
Fault Module Name: oleaut32.dll
Fault Module Version: 6.0.6002.18005
Fault Module Timestamp: 49e037da
Exception Code: c0000005
Exception Offset: 0001e674
OS Version: 6.0.6002.2.2.0.768.3
Locale ID: 1033
Additional Information 1: 714b
Additional Information 2: fd1c00303581f3a3ece51bf9cafd77a4
Additional Information 3: 7bbf
Additional Information 4: 69ad086630f805a5bbe7f14dace2c961
Files that help describe the problem
Version.txt
AppCompat.txt
memory.hdmp
minidump.mdmp
Any ideas?
According to Problem Reports and Solutions, it appears that the problem is the ResWare Shell Launcher, but that may be only what I am shutting down, rather than the cause of the problem. These are the details provided by PR&S:
Product
ResWare Shell Launcher
Problem
Stopped working
Date
4/10/2011 4:55 PM
Status
Not Reported
Problem signature
Problem Event Name: APPCRASH
Application Name: ResScan.exe
Application Version: 0.2.0.60
Application Timestamp: 00000000
Fault Module Name: oleaut32.dll
Fault Module Version: 6.0.6002.18005
Fault Module Timestamp: 49e037da
Exception Code: c0000005
Exception Offset: 0001e674
OS Version: 6.0.6002.2.2.0.768.3
Locale ID: 1033
Additional Information 1: 714b
Additional Information 2: fd1c00303581f3a3ece51bf9cafd77a4
Additional Information 3: 7bbf
Additional Information 4: 69ad086630f805a5bbe7f14dace2c961
Files that help describe the problem
Version.txt
AppCompat.txt
memory.hdmp
minidump.mdmp
Any ideas?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
Re: ResScan is shutting down...
The only two things I can think of would be complete clean reinstall of Resscan and run the repair for all windows dlls. If you're using XP it's the SFC command. Can't tell you off the top of my head how to use it, but a search at Microsoft will give you all the instructions you need. SFC may work with Vista & Win 7, I don't know.Breathe Jimbo wrote:The program hangs at the "ResScan is shutting down..." window. The CRWS icon is usually, if not always, in my taskbar at the time. Whatever the cause, I need to stop ResScan in the Task Manager to close the program. ResScan is the only program listed on the program tab at the time.
According to Problem Reports and Solutions, it appears that the problem is the ResWare Shell Launcher, but that may be only what I am shutting down, rather than the cause of the problem. These are the details provided by PR&S:
[snip a bunch!]
Any ideas?
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: ResScan is shutting down...
Breathe Jimbo,Breathe Jimbo wrote:The program hangs at the "ResScan is shutting down..." window. The CRWS icon is usually, if not always, in my taskbar at the time. Whatever the cause, I need to stop ResScan in the Task Manager to close the program. ResScan is the only program listed on the program tab at the time.
According to Problem Reports and Solutions, it appears that the problem is the ResWare Shell Launcher, but that may be only what I am shutting down, rather than the cause of the problem. These are the details provided by PR&S:
Product
ResWare Shell Launcher
Problem
Stopped working
Date
4/10/2011 4:55 PM
Status
Not Reported
Problem signature
Problem Event Name: APPCRASH
Application Name: ResScan.exe
Application Version: 0.2.0.60
Application Timestamp: 00000000
Fault Module Name: oleaut32.dll
Fault Module Version: 6.0.6002.18005
Fault Module Timestamp: 49e037da
Exception Code: c0000005
Exception Offset: 0001e674
OS Version: 6.0.6002.2.2.0.768.3
Locale ID: 1033
Additional Information 1: 714b
Additional Information 2: fd1c00303581f3a3ece51bf9cafd77a4
Additional Information 3: 7bbf
Additional Information 4: 69ad086630f805a5bbe7f14dace2c961
Files that help describe the problem
Version.txt
AppCompat.txt
memory.hdmp
minidump.mdmp
Any ideas?
As I suggested in our PM discussion, some of us have found 3.12 more stable than 3.13. Perhaps the same may hold true for 3.14. I'm curious if anyone else has had this experience with 3.14. I have not.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
I've had this problem with the last few versions of ResScan. On a hunch, I tried using a different card reader. ResScan has not yet hung and is closing faster, without leaving the CRWS icon in the taskbar. However, I have not yet added new data since switching card readers. I should have a better sense in a few days whether the problem is due to a flakey card reader.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
Re: ResScan is shutting down...
Very well could be. I've got a card reader that causes no problems except when I have Lotus 1-2-3 open and it wants to automatically save a new unsaved file. It comes up with some oddball msg and I just click on continue and all is OK. Only happens if the card reader is plugged in even with no card in the reader.Breathe Jimbo wrote:I've had this problem with the last few versions of ResScan. On a hunch, I tried using a different card reader. ResScan has not yet hung and is closing faster, without leaving the CRWS icon in the taskbar. However, I have not yet added new data since switching card readers. I should have a better sense in a few days whether the problem is due to a flakey card reader.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
No hangs since I started using a different card reader. Therefore, the problem seems to have been caused by the flakey card reader that I stopped using. The one that I got from our host is working perfectly. More expensive, but reliable.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: ResScan is shutting down...
Card readers do wear out, too. I'm glad the new reader was the fix.Breathe Jimbo wrote:No hangs since I started using a different card reader. Therefore, the problem seems to have been caused by the flakey card reader that I stopped using. The one that I got from our host is working perfectly. More expensive, but reliable.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
As I try to figure out the report printing problem, I now get hangs all the time. The CRWS icon remains in my toolbar even after I use the Task Manager to end the ResScan program. After shutting the program, I still see processes for crws.exe and devicescan2.exe.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: ResScan is shutting down...
Reboot.Breathe Jimbo wrote:As I try to figure out the report printing problem, I now get hangs all the time. The CRWS icon remains in my toolbar even after I use the Task Manager to end the ResScan program. After shutting the program, I still see processes for crws.exe and devicescan2.exe.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- Breathe Jimbo
- Posts: 954
- Joined: Sun Sep 12, 2010 3:02 pm
- Location: Pasadena, CA
Re: ResScan is shutting down...
Stopping crws.exe and devicescan2.exe in the Task Manager also works.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap |
-
WaterSkimmer73
Re: ResScan is shutting down...
Hi Folks,
I also found that replacing the card reader instantly fixed the problem of ResScan (v3.14) hanging/going very slowly when shutting down. It even works fine on an external USB hub.
WaterSkimmer
I also found that replacing the card reader instantly fixed the problem of ResScan (v3.14) hanging/going very slowly when shutting down. It even works fine on an external USB hub.
WaterSkimmer