Climateline hose burst last night - only 1 month old
- 
				Laurie1041
- Posts: 162
- Joined: Sat Oct 23, 2010 10:24 pm
- Location: Kentucky
Climateline hose burst last night - only 1 month old
I have been reading that there are problems with the Climateline hose so I check it carefully for any signs of damage every night before using it.  It checked out fine until an hour later, I was awakened by a very loud noise and sudden deflation of my mask.  The hose seems to have split open very close to the outflow port.  I'll try to use some of the earlier suggestions to fix it until I can get a replacement, but it seems dicey.  Last night I had to use a standard hose from a previous loaner machine and it was so rigid and uncomfortable that I could not sleep - so maybe trying a patch is not such a bad idea.  To anyone who has sucessfully patched their Climateline hose, what seemed to work best?
Thanks! Laurie
			Thanks! Laurie
_________________
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
- 
				sleepnationtv
- Posts: 283
- Joined: Mon Jul 19, 2010 7:23 pm
Re: Climateline hose burst last night - only 1 month old
There have been a lot of problems with the climate line. Resmed has a 30 day warranty on them and our regional rep is replacing them even after the 30 day period because they know there is a problem. You may talk to your local supplier or see if you can contact your regional resmed rep.
			
			
									
									Dane Schapper, Founder cpaplibrary.com
Check out the cpaplibrary video bloghttp://www.cpaplibrary.com
						Check out the cpaplibrary video bloghttp://www.cpaplibrary.com
- 
				Laurie1041
- Posts: 162
- Joined: Sat Oct 23, 2010 10:24 pm
- Location: Kentucky
Re: Climateline hose burst last night - only 1 month old
Thank you for the info. I have emailed my DME and am waiting for a response. Lauriesleepnationtv wrote:There have been a lot of problems with the climate line. Resmed has a 30 day warranty on them and our regional rep is replacing them even after the 30 day period because they know there is a problem. You may talk to your local supplier or see if you can contact your regional resmed rep.
_________________
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
Re: Climateline hose burst last night - only 1 month old
I'd bet duct tape would get you by until you can get a REAL hose.
			
			
									
									The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...
						Unlike his passengers who died screaming as the car went over the cliff...
Re: Climateline hose burst last night - only 1 month old
This is good to know about and am wondering if it might not be prudent to get an suss hose as a back up. I have none. At least I will get some duct tape just in case.
Namaste,
MoonBear
			Namaste,
MoonBear
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
| Additional Comments: This equipment has, apart from a bit of bridge of nose irritation, worked well for me. | 
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: Climateline hose burst last night - only 1 month old
This doesn't help people who already have an S9 climateline hose, but I'm gonna put in a plug for "what I'd do" if I were thinking about buying a ResMed S9 machine.
I'd buy the S9 machine -- without buying the integrated humidifier for it -- and, of course, without buying the climateline hose. Just the machine...nothing else.
At the same time, I'd buy a Fisher & Paykel HC 150 humidifier. It comes with its own connector hose to connect the F&P humidifier to any cpap/autopap/bilevel machine. I'd use the F&P humidifier with the S9 machine.
I'd also buy either the "sleepzone heated cpap tube" (often referred to here as the "Aussie heated hose") or I'd get a Repti heat cable to keep the main air hose warm -- to prevent rainout.
LINKS to discussions about the Aussie heated hose - to prevent rainout
viewtopic.php?t=5305
Besides working fine with the S9 machine, that F&P humidifier and my heated hose would work fine with any other machine I had, or got in the future. I wouldn't ever have to buy a new "integrated" humidifier if I ever decided to change machine brands.
But most of all, I wouldn't have to worry about "climateline" hose problems when using the S9 machine.
			
			
									
									I'd buy the S9 machine -- without buying the integrated humidifier for it -- and, of course, without buying the climateline hose. Just the machine...nothing else.
At the same time, I'd buy a Fisher & Paykel HC 150 humidifier. It comes with its own connector hose to connect the F&P humidifier to any cpap/autopap/bilevel machine. I'd use the F&P humidifier with the S9 machine.
I'd also buy either the "sleepzone heated cpap tube" (often referred to here as the "Aussie heated hose") or I'd get a Repti heat cable to keep the main air hose warm -- to prevent rainout.
LINKS to discussions about the Aussie heated hose - to prevent rainout
viewtopic.php?t=5305
Besides working fine with the S9 machine, that F&P humidifier and my heated hose would work fine with any other machine I had, or got in the future. I wouldn't ever have to buy a new "integrated" humidifier if I ever decided to change machine brands.
But most of all, I wouldn't have to worry about "climateline" hose problems when using the S9 machine.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
						Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: Climateline hose burst last night - only 1 month old
Fill the humidifier with vodka and you won't notice the leak and probably won't care, lol
			
			
									
									
						- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Climateline hose burst last night - only 1 month old
My first Climateline spread in the same place without splitting--on the third night.
After the DME replaced it, I wrapped both ends with vetwrap over the heavy rubber end
and onto the hose for a couple of inches to to give it more of a transition from stiff to soft.
It's been working fine for a couple of months. Knock on wood.
			After the DME replaced it, I wrapped both ends with vetwrap over the heavy rubber end
and onto the hose for a couple of inches to to give it more of a transition from stiff to soft.
It's been working fine for a couple of months. Knock on wood.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Airsense 10 Autoset for Her | 
- 
				ghstrainer
- Posts: 1
- Joined: Sat Oct 02, 2010 10:02 pm
- Location: New England
Re: Climateline hose burst last night - only 1 month old
I had mine split last night also after just 5 nights of use!!!!  It split right at the junction at the tube/collar connection (mask end).  The clear material that comprises the tube seems thinner yet harder than either my slimline or my standard hose. I was thinking, is anyone aware of a bad batch of these, or if recently produced versions have been improved.  If so, is there a manufacture date or SN that we can check?  I was just starting to enjoy the consistency and comfort of the climateline system.
			_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
| Additional Comments: Began tx Oct. 2010; Pressure @ 10-15; EPR @ 2; Rescan v3.12 | 
					Last edited by ghstrainer on Sat Dec 11, 2010 10:34 pm, edited 1 time in total.
									
			
									
						- 
				Laurie1041
- Posts: 162
- Joined: Sat Oct 23, 2010 10:24 pm
- Location: Kentucky
Re: Climateline hose burst last night - only 1 month old
"Honey, sorry about your Grey Goose, I thought it was distilled water! Best sleep I ever had."Rick B wrote:Fill the humidifier with vodka and you won't notice the leak and probably won't care, lol
_________________
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
- 
				Laurie1041
- Posts: 162
- Joined: Sat Oct 23, 2010 10:24 pm
- Location: Kentucky
Re: Climateline hose burst last night - only 1 month old
I just purchased some F4 self-sealing silicone tape and will give that a try. I purchased another Climateline from our sponsor which I hope gets here quickly. I will have a back-up as soon as I get the replacement hose from my DME. I was afraid to try duct tape in the event it leaked and end up with duct tape that I could not remove without destroying the hose.LinkC wrote:I'd bet duct tape would get you by until you can get a REAL hose.
_________________
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
- 
				Laurie1041
- Posts: 162
- Joined: Sat Oct 23, 2010 10:24 pm
- Location: Kentucky
Re: Climateline hose burst last night - only 1 month old
Excellent idea. Obviously Resmed knows there is an issue with the Climateline hose. Whether or not they are doing something about it is unknown to me at this time. My concern is whether something more insidious than a "blow-out" could occur because of manufacturing defects. After all, the hose is plugged into an electrical circuit. I will be writing to Resmed concerning my experience with their defective Climateline and letting them know that I am reporting the defect to the FDA.sleepnationtv wrote:There have been a lot of problems with the climate line. Resmed has a 30 day warranty on them and our regional rep is replacing them even after the 30 day period because they know there is a problem. You may talk to your local supplier or see if you can contact your regional resmed rep.
_________________
| Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear | 
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control | 
Re: Climateline hose burst last night - only 1 month old
My slimline lasted 2 weeks.  It split where the hose connects to the rubber cuff.  So the first thing I did when I received my climateline was to wrap both ends where the hose meets the rubber cuff with F4 tape.  Its worked fine now for a couple of months.
I don't understand why resmed doesn't fix this problem.
			
			
									
									
						I don't understand why resmed doesn't fix this problem.
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: Climateline hose burst last night - only 1 month old
Several background threads about problems with the ResMed S9 climateline hose:
May 2010 - a six page thread:
viewtopic.php?p=481136#p481136
Resmed™ S9 Climateline hose durability - topic started by Uncle_Bob
On page 4 of that thread:
_________________________________________________
July 2010:
viewtopic.php?p=504492#p504492
S9 Climate Line has a hole! - topic started by Big Daddy RRT,RPSGT
_________________________________________________
August 2010:
viewtopic.php?p=505610#p505610
ClimateLine Hose how to tell "Original" vs "Improved" Model - topic started by ldj325
A guest wrote:
September 2010
viewtopic.php?p=520459#p520459
Resmed S9 H5is and ClimateLine Tubing - topic started by Johnny Goodman (owner of cpap.com:)
September 2010:
viewtopic.php?p=519554#p519554
Climateline hose and S9 Heated Humidifier - topic started by Sanadol
________________________________________________
			
			
									
									May 2010 - a six page thread:
viewtopic.php?p=481136#p481136
Resmed™ S9 Climateline hose durability - topic started by Uncle_Bob
On page 4 of that thread:
WillSucceed wrote:I spoke today with the ResMed rep here in Toronto, telling him about the concerns expressed with hoses leaking and, limited supply of hoses (I've been waiting over a month for a Climate-line).
He is aware of a durability problem with the S9 hoses but thinks it is not a big a problem as it is being made out to be by users. He has assurred me that ResMed is on top of this and that hoses should be available by late next week.
_________________________________________________
July 2010:
viewtopic.php?p=504492#p504492
S9 Climate Line has a hole! - topic started by Big Daddy RRT,RPSGT
steamfan wrote:I got my S9 and the assorted hoses at the Mayo Clinic Sleep Store, and they were at great pains to point out that the Slimline and Climateline both suffered from a problem where they developed leaks near the connector if you pulled on the hose while disconnecting it. They said that ResMed is working on a redesign.
_________________________________________________
August 2010:
viewtopic.php?p=505610#p505610
ClimateLine Hose how to tell "Original" vs "Improved" Model - topic started by ldj325
A guest wrote:
_________________________________________________guest wrote:The S9 Senior Product Manager of Resmed in San Diego was made aware of this ClimateLine topic and responded, "I'm not aware of any changes. They may have made some molding changes with the new equipment, but I haven't heard of any actual changes."
September 2010
viewtopic.php?p=520459#p520459
Resmed S9 H5is and ClimateLine Tubing - topic started by Johnny Goodman (owner of cpap.com:)
________________________________________________johnnygoodman wrote:Howdy All,
We received communication back from Resmed regarding the S9 ClimateLine and Humidifier. Here are the core points of their bulletin:
1. These operating conditions present no known safety issue. The S9 displaying a fault message under these conditions is not intended and is a software bug.
2. This is a software issue that causes an incorrect message to be displayed. A software upgrade is being performed on units returned for warranty repair that will prevent the "Heated Tube fault, replace tube" error from preventing system operation.
3. To cause this issue one must cause the temperature in the ClimateTube to exceed 90 deg F / 32 deg C. This temperature can only be achieved if the tube is operated under one's sheets for an extended period, the room is over 77 degrees F, or the tube is in direct sunlight while in use.
4. The error goes away each time the temperature drops below the 90 Deg threshold and reappears each time the threshhold is exceeded. If your S9 shuts down during sleep, try unplugging the unit, then powering up again. The CPAP will normally come back on until the 90 Deg threshold is again triggered.
5. Substitute a Slimline tube if the problem persists as a temporary fix.
Here is advice issued in the bulletin:
1. Make sure your humidifier tub is full at the start of therapy. A dry plate humidifier can increase the chance of exceeding temperature bounds.
2. Turn down the humidifier/climateline setting.
If you bought your unit from CPAP.com and have concerns or want a software upgrade, call us at 1-800-356-5221.
Johnny
September 2010:
viewtopic.php?p=519554#p519554
Climateline hose and S9 Heated Humidifier - topic started by Sanadol
Sanadol wrote:My Climateline hose suddenly stopped working after 3 days of use. I had to send it back to cpap.com.
I got a call from them yesterday saying that I need to send my my humidifier too.
The guy i spoke with told me that Resmed got back to them that the "heated tube failure. replace tube." error is caused by the humidifier and replacing the hose alone will only lead to more climateline tube failures.
Has anyone heard of any type of humidifier recall from Resmed or 'am just the only one with the hose "frying" humidfier
________________________________________________
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
						Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: Climateline hose burst last night - only 1 month old
I've been using the ClimateLine hose for six months now (unheated most of that time).  Because I was aware that there was a tendency for these hoses to split near the rubber cuff, I wrapped both ends with self-fusing, heat-resistant silicone Rescue Tape before I started using it.  I purchased the tape on amazon.com.  No problems so far.
			
			
									
									
						 
                 
                         
                         
                         
                         
                         
                         
                         
			 
	