using cpap and treadmill together......

Post by semreka » Sun Aug 15, 2010 10:41 am

Hi,
I am curious about using the cpap machine while exercising on treadmill.
Is there anyone who experienced this ?
I reached this idea because of lack of O2 while you are exercising?
Cpap will supply you good inhale and you will exhale it by simply opening your mouth ?
I did not try it yet , I am planning to try...
What is your opinion about this ???

)))))))))))))))))))))))

God Bless..............
Last edited by semreka on Sun Aug 15, 2010 11:45 am, edited 1 time in total.

User avatar
Emilia
Posts: 1873
Joined: Mon Jul 12, 2010 8:56 am
Location: Florida, USA

Re: using cpap and trade mill together......

Post by Emilia » Sun Aug 15, 2010 10:51 am

I think your premise is false. CPAP is used to open the airway when at rest...because the throat muscles relax or one's tongue falls back and closes off the airway while you are asleep. If you are awake, upright, and exercising, your airway is open! A cpap machine does NOT provide additional oxygen, it just provides pressurized room air to open a closed airway. So.... I would NOT be trying this whilst walking on a treadmill!
Yes, that blue eyed beauty is my cat! He is a seal point, bi-color Ragdoll. I adopted him in '08 from folks who could no longer care for him. He is a joy and makes me smile each and every day.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: using cpap and trade mill together......

Post by chunkyfrog » Sun Aug 15, 2010 10:54 am

No; did you really mean 'treadmill'?
Still no, in my opinion.
Try breathe-rite strips.
Or home-made ones (tape and a strip of plastic like off a cottage cheese lid)
Good luck!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
roster
Posts: 8162
Joined: Wed Jan 18, 2006 8:02 pm
Location: Chapel Hill, NC

Re: using cpap and trade mill together......

Post by roster » Sun Aug 15, 2010 11:27 am

I think you should try it and report back. Turn the pressure to the highest setting.
Rooster
I have a vision that we will figure out an easy way to ensure that children develop wide, deep, healthy and attractive jaws and then obstructive sleep apnea becomes an obscure bit of history.https://www.youtube.com/watch?v=0ycw4uaX ... re=related

User avatar
semreka
Posts: 61
Joined: Fri Jul 30, 2010 4:49 am

Re: using cpap and trade mill together......

Post by semreka » Sun Aug 15, 2010 11:40 am

As an engineer who has taken courses about fluid mechanics in university education , I can say the pressure generated by cpap machine is directly related with the additional debit in the air flow through your lungs. The machine is using the air in your room .It is respirable (approx. 80 % of nitrogen and 20 % oxygen ) . I takes the air in your room and insufflate it into your nose. The whiffled air through your weasand generates a air barrier in your throat and prevents the collapse in soft tissue of your throat. It is simple mechanism. And the air is air. To generate a presure there should be forced flow. that means additional O2 molecules.
It may be useful for your cells to produce more ATP . But I am not sure ....
If any one has some idea about the subject please feel free to contribute.
GOD BLESS....

User avatar
semreka
Posts: 61
Joined: Fri Jul 30, 2010 4:49 am

Re: using cpap and trade mill together......

Post by semreka » Sun Aug 15, 2010 11:42 am

semreka wrote:As an engineer who has taken courses about fluid mechanics in university education , I can say the pressure generated by cpap machine is directly related with the additional debit in the air flow through your lungs. The machine is using the air in your room .It is respirable (approx. 80 % of nitrogen and 20 % oxygen ) . It takes the air in your room and insufflate it into your nose. The whiffled air through your weasand generates a air barrier in your throat and prevents the collapse in soft tissue of your throat. It is simple mechanism. And the air is air. To generate a presure there should be forced flow. that means additional O2 molecules.
It may be useful for your cells to produce more ATP . But I am not sure ....
If any one has some idea about the subject please feel free to contribute.
GOD BLESS....

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: using cpap and treadmill together......

Post by Julie » Sun Aug 15, 2010 11:49 am

And... if you were to open your mouth to exhale, all the air would be gone anyhow!

User avatar
semreka
Posts: 61
Joined: Fri Jul 30, 2010 4:49 am

Re: using cpap and treadmill together......

Post by semreka » Sun Aug 15, 2010 11:56 am

and,
when exhaling finishes , you can close your mouth and inhale through your nose...
I do not know ....
This is not a important topic...
I just want to have brain storming on this subject for fun....

I thank everyone who worries about me ...
Fell free, I will never put my self in a dangerous condition....

)))))))))))))

User avatar
roster
Posts: 8162
Joined: Wed Jan 18, 2006 8:02 pm
Location: Chapel Hill, NC

Re: using cpap and trade mill together......

Post by roster » Sun Aug 15, 2010 12:09 pm

semreka wrote:To generate a presure there should be forced flow. that means additional O2 molecules.

When CPAP is working properly, there is no additional air flow as compared to normal breathing. CPAP creates a pressure to splint the airway open. The rate of airflow is determine by the movement of the diaphragm.

Now with CPAP, you are getting 'slightly' pressurized air into the lungs. A high CPAP pressure is 20 cm H2O. One atmosphere is 1033 cm H2O. Add the CPAP pressure and you are at 1053 cm. This is only a 2% increase in pressure, so you are bringing very few additional molecules into the lungs.

When you try it, post it on YouTube and give us the link.
Rooster
I have a vision that we will figure out an easy way to ensure that children develop wide, deep, healthy and attractive jaws and then obstructive sleep apnea becomes an obscure bit of history.https://www.youtube.com/watch?v=0ycw4uaX ... re=related

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: using cpap and treadmill together......

Post by BlackSpinner » Sun Aug 15, 2010 12:24 pm

The lack of O2 when running (assuming properly functioning lungs) is more due to the breathing method and lack of practice. Many people start breathing very shallowly and very fast (panting). Try to teach yourself to breathe deeply using your diaphragm and all of your lungs and hold the air in your lungs before exhaling - yoga style breathing. It makes a huge difference. I usually use my heart beat to count with . 4 in, 4 hold, 4 out 4 hold - this means very long slow breathing during meditation and quick deep breathes when running. The trick is to start breathing like that before you start, if you wait until you NEED air you won't have the control to do it.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: using cpap and treadmill together......

Post by Goofproof » Sun Aug 15, 2010 1:11 pm

Not a good idea, XPAP machines are designed as sleep use devices. When we sleep our need for volume of air goes down, just the backwards from exercising. A XPAP doesn't supply the flow rates needed for exercise. Don't believe me, have sex with your mask on, you will want a bigger hose. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
semreka
Posts: 61
Joined: Fri Jul 30, 2010 4:49 am

Re: using cpap and treadmill together......

Post by semreka » Sun Aug 15, 2010 1:32 pm

Goofproof wrote:Not a good idea, XPAP machines are designed as sleep use devices. When we sleep our need for volume of air goes down, just the backwards from exercising. A XPAP doesn't supply the flow rates needed for exercise. Don't believe me, have sex with your mask on, you will want a bigger hose. Jim
Perfect answer......................
You made me laugh.........)))))))))))))))))))))))
I am happy with my hose when I am making sex.....(even it is not the hose of the cpap)
But we should take the idea of the partners also...
It is hard to hear from a lady , the word "enough"..........
)))))))))))))))))))))
Thanks for laughs.......
GOD BLESS..

User avatar
bearded_two
Posts: 459
Joined: Mon Aug 10, 2009 8:01 pm

Re: using cpap and treadmill together......

Post by bearded_two » Sun Aug 15, 2010 5:07 pm

It would be a better idea to adapt a 12 volt automotive alternator with appropriate electronics to your treadmill and use it to charge a backup battery for your CPAP.

User avatar
cpapernewbie
Posts: 685
Joined: Fri Oct 13, 2006 5:53 pm

Re: using cpap and treadmill together......

Post by cpapernewbie » Mon Aug 16, 2010 12:17 am

this give me a new idea:
sleep walking still using my S9 and on a treadmill

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: CPAP history: dumb tank, auto, PR M, PR System 1, PR BIPAP, PR System 1 model 60, Resmed S9, Resmed S10, Dreamstation
Resmed airsense 10

User avatar
mreewh
Posts: 40
Joined: Wed Nov 04, 2009 10:58 pm

Re: using cpap and treadmill together......

Post by mreewh » Mon Aug 16, 2010 12:49 am

Goofproof wrote:have sex with your mask on, you will want a bigger hose.


But back to the original topic ... I think this is a legit question. I've noticed that now that I'm using my hybrid mask, I get much improved airflow through my nose at night. It also seems better most of the day, but by the time I'm ready for bed again, I notice the positive effect once more when ramped back up to full pressure.

I've always had problems getting enough air during exercise, from as early as I can remember from childhood.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: CMS-50E Pulse Oximeter
Marie
Severe OSA, diagnosed August 2009