cpap pressure not as strong? Help plz

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Geminidream
Posts: 118
Joined: Sun Dec 13, 2009 8:25 am
Location: Ohio

cpap pressure not as strong? Help plz

Post by Geminidream » Fri Apr 02, 2010 4:00 pm

Hello all,
Just in the last few days, my cpcp machine has felt like the air pressure is not as strong coming out of the tube as it felt previously and my AHI/AI rate has climbed up as well. I have changed the filter and hose and have not changed anything else. I have a copy of the clinical manual, but the troubleshooting page was usless for this issue. Anyone else ever have this problem?

Geminidream

User avatar
bearded_two
Posts: 459
Joined: Mon Aug 10, 2009 8:01 pm

Re: cpap pressure not as strong? Help plz

Post by bearded_two » Fri Apr 02, 2010 4:50 pm

You can make or build a manometer to check your pressure or you can take your blower to your DME to check the pressure.

Country4ever
Posts: 1373
Joined: Wed Oct 31, 2007 6:22 pm

Re: cpap pressure not as strong? Help plz

Post by Country4ever » Fri Apr 02, 2010 5:05 pm

Hi Geminidream,
I had the same problem with my Resmed 8 auto vantage. I checked the pressure myself and also had the DME check it.....although she wasn't real helpful as she said the pressure was somewhere between 7 and 10 when it was set to about 7.6. I've heard that the machines are seldom wrong.
Can you test the short tube for leaks? Is your leak rate going up too, or just your AHI?
My AHI and leak rates both got so bad, and I started feeling bad, so I switched from a Swift nasal pillow to the Hybrid. Just my second night, so I'm not sure that will help yet.
When I asked the same question you're asking, someone suggested getting a loaner from the DME and sleeping a few nights with that one, and seeing if your numbers are consistent on a different machine. I thought that was a good suggestion. If my full mask doesn't help my numbers, I'll do that next.
I swear that my pressure is less in the morning........although lots of people say we just get used to it by then. But I've noticed that if I let the air fill my mouth when I go to bed, my cheeks puff out big. But when I do the same thing in the morning, they hardly puff out at all.
Is there anything different in your life? A few more pounds, a cold, stress? Many of us seem to go through cycles and can't explain why all of a sudden our numbers get worse. I'm ready for a good cycle! Good luck to you!

_________________
MachineMask
Additional Comments: Resmed HumidAir 11 for humidifier

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: cpap pressure not as strong? Help plz

Post by Goofproof » Fri Apr 02, 2010 10:16 pm

Country4ever wrote:Hi Geminidream,
I had the same problem with my Resmed 8 auto vantage. I checked the pressure myself and also had the DME check it.....although she wasn't real helpful as she said the pressure was somewhere between 7 and 10 when it was set to about 7.6. I've heard that the machines are seldom wrong.
Can you test the short tube for leaks? Is your leak rate going up too, or just your AHI?
My AHI and leak rates both got so bad, and I started feeling bad, so I switched from a Swift nasal pillow to the Hybrid. Just my second night, so I'm not sure that will help yet.
When I asked the same question you're asking, someone suggested getting a loaner from the DME and sleeping a few nights with that one, and seeing if your numbers are consistent on a different machine. I thought that was a good suggestion. If my full mask doesn't help my numbers, I'll do that next.
I swear that my pressure is less in the morning........although lots of people say we just get used to it by then. But I've noticed that if I let the air fill my mouth when I go to bed, my cheeks puff out big. But when I do the same thing in the morning, they hardly puff out at all.
Is there anything different in your life? A few more pounds, a cold, stress? Many of us seem to go through cycles and can't explain why all of a sudden our numbers get worse. I'm ready for a good cycle! Good luck to you!
If your machine has a 3 cm difference between the pressure you set it for (ON CPAP), it is junk, and your DME is a Quack, the spec is 1 CM. JIM

As we use XPAP we get used to the pressure, that's why when we first turn it on it seems to be blowing harder than in the morning when we take it off.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Komodo
Posts: 464
Joined: Tue Jun 09, 2009 1:29 pm
Location: Zephyrhills FL

Re: cpap pressure not as strong? Help plz

Post by Komodo » Sat Apr 03, 2010 4:54 am

Geminidream wrote:Hello all,
Just in the last few days, my cpcp machine has felt like the air pressure is not as strong coming out of the tube as it felt previously and my AHI/AI rate has climbed up as well. I have changed the filter and hose and have not changed anything else. I have a copy of the clinical manual, but the troubleshooting page was usless for this issue. Anyone else ever have this problem?

Geminidream
I had the same problem, with the same machine. Pressure started out fine, then by the middle of the night it was barely putting out anything. I called my DME and had them exchange it. No problems with the new one.

I will ask you the same question everyone here asked me when I posted about my machine: "Are you sure it's the machine, and not that you're just used to the pressure and don't feel it, even though it IS working?"

User avatar
Geminidream
Posts: 118
Joined: Sun Dec 13, 2009 8:25 am
Location: Ohio

Re: cpap pressure not as strong? Help plz

Post by Geminidream » Sat Apr 03, 2010 5:22 am

bearded_two wrote:You can make or build a manometer to check your pressure or you can take your blower to your DME to check the pressure.
I have no idea what a manometer is.....could you explain? I may just wind up taking my machine to the DME next week, however.

User avatar
Geminidream
Posts: 118
Joined: Sun Dec 13, 2009 8:25 am
Location: Ohio

Re: cpap pressure not as strong? Help plz

Post by Geminidream » Sat Apr 03, 2010 5:25 am

Komodo wrote:
Geminidream wrote:Hello all,
Just in the last few days, my cpcp machine has felt like the air pressure is not as strong coming out of the tube as it felt previously and my AHI/AI rate has climbed up as well. I have changed the filter and hose and have not changed anything else. I have a copy of the clinical manual, but the troubleshooting page was usless for this issue. Anyone else ever have this problem?

Geminidream
I had the same problem, with the same machine. Pressure started out fine, then by the middle of the night it was barely putting out anything. I called my DME and had them exchange it. No problems with the new one.

I will ask you the same question everyone here asked me when I posted about my machine: "Are you sure it's the machine, and not that you're just used to the pressure and don't feel it, even though it IS working?"
No, its starts out feeling like less air is coming out when I first turn it on.....and then I feel like I am not getting enough air right off the bat as I try to go to sleep. This morning's numbers were pretty decent, but I still feel like I had to struggle with it all night. I have been on cpap therapy faithfully since the week before Christmas last year and nothing has changed until now.

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Re: cpap pressure not as strong? Help plz

Post by ozij » Sat Apr 03, 2010 5:54 am

Any chance you inserted the filter with the blue towards inside instead of outside?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Geminidream
Posts: 118
Joined: Sun Dec 13, 2009 8:25 am
Location: Ohio

Re: cpap pressure not as strong? Help plz

Post by Geminidream » Sun Apr 04, 2010 4:45 am

ozij wrote:Any chance you inserted the filter with the blue towards inside instead of outside?
No...filter is in the right position. Now, the last two nights my numbers have been like they were and I feel okay, but it still does not feel like its putting out as much pressure as it had been. I may be making a trip to the DME Tues to check it out.

bigk
Posts: 608
Joined: Tue Jan 22, 2008 5:35 pm
Location: Central Coast near Sydney Australia

Re: cpap pressure not as strong? Help plz

Post by bigk » Sun Apr 04, 2010 5:06 am

What ramp time do you have set? What is the start ramp pressure? What is the treatment pressure? Do you have EPR on? What value is that?

_________________
MaskHumidifier
Additional Comments: S9 Autoset, H5i, Resmed Swift FX, Resscan 3.10
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose

User avatar
Spirit
Posts: 433
Joined: Fri Aug 07, 2009 9:27 pm
Location: Central Tx

Re: cpap pressure not as strong? Help plz

Post by Spirit » Sun Apr 04, 2010 10:18 am

Hi,
Have you cleaned your hose recently? I had similar experience, after rinsing out my hose, air pressure decreased alot !!! felt like I was suffocating. Hose had water droplet build up. Do you have humidifier set to high ? Possible water build up in hose, take off hose, clean, completely dry, then see if pressure returns back to normal.

User avatar
Geminidream
Posts: 118
Joined: Sun Dec 13, 2009 8:25 am
Location: Ohio

Re: cpap pressure not as strong? Help plz

Post by Geminidream » Mon Apr 05, 2010 11:58 pm

Thanks for the replies!

I have no water in the hose....don't even have the humidifier on right now because its warm here now. I don't have the ramp on at all, pressure is set at 6 and the epr was not turned on by the DME, apparently.....soooo I am off to see the DME today and have it checked. She said on the phone I may just be getting used to the pressure, but its not just different when the mask is on my face, I feel a difference coming out of the hose without the mask attached to it as well. We shall see.

Will let ya know what the result is!

Geminidream

ZzzzzGoodnight

Re: cpap pressure not as strong? Help plz

Post by ZzzzzGoodnight » Sun Oct 02, 2016 12:55 am

Check the damb machine to see if it may have sucked in some obstacle. People? What's the matter with you? He wants some suggestions not a novel.

baja05
Posts: 10
Joined: Sun Sep 25, 2016 5:43 pm

Re: cpap pressure not as strong? Help plz

Post by baja05 » Sun Oct 02, 2016 3:09 am

The motor can be getting worn out. It may build enough pressure but the flow may be diminished. A calibration may be the only way to tell. What machine is it and how many hours?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: cpap pressure not as strong? Help plz

Post by Julie » Sun Oct 02, 2016 3:47 am

You're responding to a 6 mo. old note and the poster's long gone...