S9, can't get AHI breakdown & data missing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

S9, can't get AHI breakdown & data missing

Post by Uncle_Bob » Mon Mar 15, 2010 11:53 am

Lately I've been trying to access the AHI breakdown on the machines display which I've never managed to do.

I did read a post about accessing another menu to do this, not the clinical menu but an advance info menu (seriously how many menus )

To access the advance info menu i hold down the info and setup button for three seconds and all that happens is the screen flickers, i just get AHI and mask fit etc.

I've tried being in the clinical menu/mode before doing this also. Sleep Quality is also set to "On".

Since this did not work i popped out the SD card and marched off to my computer to download my data for the night.

And now i find that all i can get for the detailed data is just the flow, i can't select any other data items

What did i do

Image

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: S9, can't get AHI breakdown & data missing

Post by jmelby » Mon Mar 15, 2010 12:07 pm

Uncle_Bob wrote:Lately I've been trying to access the AHI breakdown on the machines display which I've never managed to do.

I did read a post about accessing another menu to do this, not the clinical menu but an advance info menu (seriously how many menus )

To access the advance info menu i hold down the info and setup button for three seconds and all that happens is the screen flickers, i just get AHI and mask fit etc.
From there you need to wheel down to the next screen down on the left and click on it to show it, then click on it again to go into the right-hand pane...then you can scroll down to the AHI breakout... I'll post a picture later after I get home. The change from the standard info page is that you get that extra screen to look at--it is not displayed by default. You also get a more detailed screen that shows all of the machine settings--the patient screen just shows options that you can change as a patient.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: S9, can't get AHI breakdown & data missing

Post by jmelby » Mon Mar 15, 2010 12:09 pm

Uncle_Bob wrote:And now i find that all i can get for the detailed data is just the flow, i can't select any other data items

What did i do
Now that is strange... I have noticed some quirks with the 3.10 software myself--where my summary graphs don't show up, though I can get to the detailed ones. Restarting ResScan seems to fix that.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: S9, can't get AHI breakdown & data missing

Post by Nord » Mon Mar 15, 2010 1:01 pm

Hi Uncle Bob...

Ya the DL from the SD card changes unless it is set. Next time... When you import your data on your card... in the box "Select Report" check the "All Available Data" and beside that in the same box check the box "Select Data Type to Download"...

A new dialog box comes up "Select Data Type to Download":
1. check the box "All Summary data and..."
2. Drop Down box... select "All"
3. "Check" include High Data Rate Sessions (etc)
and most importantly
4. "Check" make these default options...
and click "OK"

You should be good to go...
Nord

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: S9, can't get AHI breakdown & data missing

Post by Uncle_Bob » Mon Mar 15, 2010 1:19 pm

Nord wrote:Hi Uncle Bob...

Ya the DL from the SD card changes unless it is set. Next time... When you import your data on your card... in the box "Select Report" check the "All Available Data" and beside that in the same box check the box "Select Data Type to Download"...

A new dialog box comes up "Select Data Type to Download":
1. check the box "All Summary data and..."
2. Drop Down box... select "All"
3. "Check" include High Data Rate Sessions (etc)
and most importantly
4. "Check" make these default options...
and click "OK"

You should be good to go...
Nord

I've got those settings already so all i have to do is select download and not specify all data each time. So i'm wondering what else it could be ...

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: S9, can't get AHI breakdown & data missing

Post by Nord » Mon Mar 15, 2010 1:34 pm

I'm doing this from memory... heh heh, so please forgive if not exactly... but as I remember:

In 3.10 go to Tools menu and select Options and Graphs: Make sure that in the Detailed options that all the options are available in 3.10 and Apply...

If so then may be the card that is not getting the info as the blower; If not then it may be a deeper software issue in 3.10.

Good luck...
Nord

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: S9, can't get AHI breakdown & data missing

Post by DreamDiver » Mon Mar 15, 2010 3:06 pm

Looks like there are still some random issues with the software...

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: S9, can't get AHI breakdown & data missing

Post by Uncle_Bob » Mon Mar 15, 2010 8:49 pm

Nasty software, error i get on uninstall and re-install. Uninstall and reboot caused my hp PC recovery screen to show up i had to unplug the power to reboot. Still all i get is flow.

Image

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: S9, can't get AHI breakdown & data missing

Post by DreamDiver » Tue Mar 16, 2010 5:48 am

Uncle_Bob wrote:Nasty software, error i get on uninstall and re-install. Uninstall and reboot caused my hp PC recovery screen to show up i had to unplug the power to reboot. Still all i get is flow.
Have you considered attempting to re-download the software? It may be yours was unintentionally corrupted during download. When I see dll errors like that, it makes me think of install file corruption.

I would use md5 check 2.1 in windows to make sure the file has a valid md5 sum.
If you compare it with this one, they should be identical:
D5C72D83D3A6F9FF120A49C14D0EAA73

If that checks out, here are some other avenues:
Has this always been the case, or did everything except flow just disappear one day?

I'm on WinXP. Are you using Vista or 7? I'm guessing Vista.
Would it help to the install in WinXP emulation?

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
Nord
Posts: 565
Joined: Sun Mar 07, 2010 4:30 pm
Location: GTA Canada

Re: S9, can't get AHI breakdown & data missing

Post by Nord » Tue Mar 16, 2010 6:20 am

DreamDiver wrote:
Uncle_Bob wrote:Nasty software, error i get on uninstall and re-install. Uninstall and reboot caused my hp PC recovery screen to show up i had to unplug the power to reboot. Still all i get is flow.
Have you considered attempting to re-download the software? It may be yours was unintentionally corrupted during download. When I see dll errors like that, it makes me think of install file corruption.

I would use md5 check 2.1 in windows to make sure the file has a valid md5 sum.
If you compare it with this one, they should be identical:
D5C72D83D3A6F9FF120A49C14D0EAA73

If that checks out, here are some other avenues:
Has this always been the case, or did everything except flow just disappear one day?

I'm on WinXP. Are you using Vista or 7? I'm guessing Vista.
Would it help to the install in WinXP emulation?
Hi DD

I think since Uncle Bob DL the main link and I UL from his... and it was working properly originally... worth a try but...
I'm also on XP Pro thru a virtual and my DL works fine. It doesn't say that the software is 7 compatible... just Vista, 2000, XP etc. You would think that the software would have alerted to an incompatible OS but who knows...

Bob... First... are you on 7 or what is your OS??? Second... did you do a complete re-install after deleting all traces of the original ResScan??? Is your PC working properly in all other ways since this developed. Perhaps there is some corruption to the OS that needs to be addressed...

Nord

User avatar
dave21
Posts: 720
Joined: Tue Mar 09, 2010 1:05 am
Location: United Kingdom

Re: S9, can't get AHI breakdown & data missing

Post by dave21 » Tue Mar 16, 2010 7:53 am

Failed to register means either you're not an administrator (try downloading the EXE installer to your hard disk and right clicking and say RUN AS ADMINISTRATOR) or login with Administrator rights to your PC.

There's also a few more links available on Google at http://www.google.co.uk/search?hl=en&cl ... =&aql=&oq= that might help a little.

It's possible that that DLL might already be copied into system area and be registered or there could be a problem with your registry.

_________________
Machine
Additional Comments: Running AirSense 10 AutoSet CPAP, previously S9 AutoSet and S8 AutoSet Spirit
Image

ozij
Posts: 10515
Joined: Fri Mar 18, 2005 11:52 pm

Re: S9, can't get AHI breakdown & data missing

Post by ozij » Tue Mar 16, 2010 10:21 am

http://support.microsoft.com/kb/249873

Also:
Error 1904.Module [filepath] Failed to Register. HRESULT -2147220473.

Summary
Installation problem can be caused by invalid atl.dll registration.

Applies To: Copernic Enterprise Search 3+

Symptom
During installation, these messages can be encountered (paths to files may vary):

· Error 1904.Module C:\Program Files\Copernic Enterprise Search 3\Bin\CESAdmin300NT.dll failed to register. HRESULT -2147220473. Contact your support personnel.

· Error 1904.Module C:\Program Files\Copernic Enterprise Search 3\Bin\CESMonitor300NT.dll failed to register. HRESULT -2147220473. Contact your support personnel.

· Error 1904.Module C:\Program Files\Copernic Enterprise Search 3\Bin\CESSearch300NT.dll failed to register. HRESULT -2147220473. Contact your support personnel.

· Error 1904.Module C:\Program Files\Copernic Enterprise Search 3\Bin\CIDXConverterLib310NT.dll failed to register. HRESULT -2147220473. Contact your support personnel.

Cause
It may be caused by a system dll (atl.dll) not being correctly registered on the host system.

Solution
1. Locate atl.dll, generally found in “C:\WINNT\system32” or “C:\WINDOWS\system32”.

2. Open the command-line prompt

a. Start Menu/Run

b. Type “cmd”

c. Press ENTER

3. Using the atl.dll filename and path, call regsvr32, i.e. at the command-line prompt, type:

a. regsvr32 “C:\WINNT\system32\atl.dll”

b. Press ENTER

4. You should see a regsvr32 window, confirming success: Close it.

5. Restart Coveo Enterprise Search installation.

Just change the dll details for yours.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: S9, can't get AHI breakdown & data missing

Post by Uncle_Bob » Tue Mar 16, 2010 12:38 pm

Thanks guys I will try all your suggestions tonight. On my window XP machine.

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: S9, can't get AHI breakdown & data missing

Post by Uncle_Bob » Wed Mar 17, 2010 1:11 pm

So last night i tried a clean install on my laptop running XP and using an SD card reader. The result was not good. I only got 4 items in the detailed graphs which is an improvement over just flow but i didn't get the events chart and some others still.

The night that showed just the flow on my desktop ResScan now shows as no detailed data available on my laptop.

Gremlins from Australia

User avatar
dave21
Posts: 720
Joined: Tue Mar 09, 2010 1:05 am
Location: United Kingdom

Re: S9, can't get AHI breakdown & data missing

Post by dave21 » Wed Mar 17, 2010 2:30 pm

Image

_________________
Machine
Additional Comments: Running AirSense 10 AutoSet CPAP, previously S9 AutoSet and S8 AutoSet Spirit
Image