Sorry, this what I found. It was for the 3 button Plus but it looks like the lite.
To go to the next setting, press and release the pressure on/off button.
To change settings, use the ramp or humidifier buttons.
First setting shown is the ELEVATION SETTING (altitude above sea level.)
Lo = Less than 2,500 feet.
M = 2,500-5,000 feet.
Hi = 5,000 - 7,500 feet.
PRESSURE SETTING:
The CPAP pressure setting will appear. To change the setting, press and release the ramp or humidifier button.
Press and release the pressure on/off button each time you want to go to the next setting.
FINE ADJUSTMENT SETTING:
The CPAP fine adjustment setting will appear. This setting allows you to calibrate the device so that the pressure can be verified with a manometer. If you DO NOT want to calibrate the device, press and release the pressure on/off button to go to the next setting.
My Note: do not change this setting unless you have a manometer.
RAMP TIME SETTING:
Range 0 - 45 minutes. The patient also has access to this in the Patient Setup menu.
RAMP STARTING PRESSURE SETTING:
Range 4 - the Pressure Setting.
Did say something about the blower would run all through this process.