My FullLife Easy-Release Fix

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BleepingBeauty
Posts: 2454
Joined: Thu Apr 02, 2009 5:30 pm
Location: Aridzona ;-)

My FullLife Easy-Release Fix

Post by BleepingBeauty » Mon Aug 17, 2009 9:44 am

Well, I'm happy to say that my amendment works well.

Here are the details on my FullLife "fix" to give the neck strap (No. 2) a quick-release so I don't have to readjust it every night.

The materials used:
one large paper clip
wire cutters
ETA: Emery board or other rasp-type file
scissors
floral tape
yarn
large-eyed sewing needle

The mask is sitting upright here, looking at the left side. The empty vertical slot (lower middle) is where the neck strap (No. 2) attaches to the mask.

Image

I cut a large paper clip at the spot where the pencil is pointing to here:

Image

That formed the piece on the right here, which is the base of my fix:

Image

ETA: If need be, smooth the cut edge of the paper clip with the emery board, and then wrap each end of the clip tightly with a small piece of floral tape before proceeding.

I used floral tape for these next two steps. Floral tape is sticky on both sides, is able to get wet, and is quite malleable. It gives whatever it covers a nice "gripper" quality. It also stretches a bit, which makes it easy to work with. (If you pull too hard and break it, you can simply overlap it, sticking it to itself, and keep going.)

I covered the paperclip with it:

Image

And then I covered the lower half of the No. 2 strap slot with it:

Image

Result so far:

Image

Then I bent the short hooked part of the clip inward slightly (so it would "hug" the top of the mask bar) and bent the long hooked part outward slightly, to result in:

Image

(The clip is actually upside-down here, considering how I'll attach it to the mask. I photographed it this way for continuity purposes.)

You can skip these next two steps. I did it to make sure the clip would fit the mask the way I wanted it to before I actually attached it, and I wanted to take pics at this stage to show you the details. I first slid the covered paper clip over the side bar. The mask is laying at a bit of an angle in this pic, facing down, with the left-side mask bar pointing up and back, to give you a good view. (The floral tape's stickiness allowed me to stick the clip to the inside of the mask bar and take this pic.) The chin support strap is in its horizontal slot to the right:

Image

Then I moved it down so that it was positioned vertically in the middle of (and behind) the open slot on the bar for Strap No. 2, with the short hook over the top of the bar. This will be the clip's position when I'm done. (Again, the floral tape allowed me to stick the clip to the mask slot and take the pic hands-free.) Sorry the pic is a little blurry, but I hope you get the idea:

Image

Next step: Cover the clip with yarn (or another similarly flexible material that you can wrap with). I laid about a half-inch of yarn along one open end of the clip, sticking it to the tape. Then I tightly wound the yarn around the clip with my fingers, covering the half-inch length that was flat against it, and then continuing along, making sure the yarn wrapped the entire clip tightly. Then I threaded the other end of the yarn with a large-eyed sewing needle and ran the needle through some of the yarn that was wrapped around the clip ("catching" a piece of yarn on several loops in a row), to secure it:

Image

For this next step, I removed the cushion from the frame so I couldn't puncture it by accident. I held the yarn-covered clip in position against the mask bar, right behind the open slot, and then used the needle and yarn to secure the clip to the mask by wrapping the lower half of the slot (where the floral tape is), going from the inside to the outside, and making sure I crossed over the underside of the clip several times in an effort to anchor it and keep it from moving. (I did that "anchoring" on each side of the slot.) When I was done wrapping, I again threaded the needle through several loops of the wrapped yarn to secure it, and then I cut it.

Here's the result (clip in place, fully wrapped and anchored vertically). I then bent the top half of the open clip inward a little bit. The empty slot you see to the right in this pic is where the chin support strap goes:

Image

I then bent the clip to the left (towards the front of the mask) a bit, to make it harder for the strap to slide up and off during the night:

Image

Here's the whole thing, assembled:

Image

I hope this fix doesn't sound (or look) too convoluted. It was easier to do than to explain , but I hope it makes sense.

Before I put the mask on last night, I made a small loop with the Velcro on the left side of the neck strap and then slipped it over the clip. I unfastened the Velcro on the right side of the strap, put the mask on, and adjusted the right side (where it attaches to the mask through its intended slot). This resulted in a large strap loop on the right side, as it now has to pick up the slack from the small loop I made on the left side. The strap stayed in place all night and worked like a charm this morning, so I deem it a success.

Effectively, all I did was extend the left-side slot for Strap No. 2 outward a little, with an opening at the top. There's no damage to the mask, and I can undo or re-do this fix whenever I want to. The right side's Velcro will stay where it is from now on, and I can simply slide the left side loop on or off the clip assembly without having to adjust anything. Mission accomplished.

I hope this post is worthwhile to someone else here, especially if it prompts another that results in a better mousetrap. If anyone has a question about what I did (or how I did it), fire away.
Last edited by BleepingBeauty on Mon Aug 17, 2009 10:46 am, edited 1 time in total.
Veni, vidi, Velcro. I came, I saw, I stuck around.

Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy. ;))

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.

User avatar
WindFlyer
Posts: 262
Joined: Wed Nov 17, 2004 12:35 pm
Location: Michigan, USA

Re: My FullLife Easy-Release Fix

Post by WindFlyer » Mon Aug 17, 2009 9:47 am

AH, the tradition of us CPAP'ers being inventive and doing what it takes continues! Good show, BleepingBeauty!

_________________
Mask
Flow generator: Mask: Fisher/Paykel HC-431 Full Face Mask Humid:
Pressure: 21/17cm H2O
40,000 hours on CPAP
100% Compliant for 15 yrs.+

Mask breathers unite!

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: My FullLife Easy-Release Fix

Post by jdm2857 » Mon Aug 17, 2009 10:05 am

You added steel to your mask. Big mistake. Now Wile E. can use the big Acme electromagnet on you while you're sleeping.

Clever and practical solution.

The only tip I would add would be to hit the cut end of the paper clip with a file (even an emery board would do) to remove any sharp burrs.
jeff

ozij
Posts: 10474
Joined: Fri Mar 18, 2005 11:52 pm

Re: My FullLife Easy-Release Fix

Post by ozij » Mon Aug 17, 2009 10:13 am

Wonderful!!!

Are you left handed? Or is it easier to reach across your face and remove the left strap with your right hand?

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
BleepingBeauty
Posts: 2454
Joined: Thu Apr 02, 2009 5:30 pm
Location: Aridzona ;-)

Re: My FullLife Easy-Release Fix

Post by BleepingBeauty » Mon Aug 17, 2009 10:36 am

WindFlyer wrote:AH, the tradition of us CPAP'ers being inventive and doing what it takes continues! Good show, BleepingBeauty!
Thanks!
jdm2857 wrote:You added steel to your mask. Big mistake. Now Wile E. can use the big Acme electromagnet on you while you're sleeping.
Rut-roh. (I know, I know, that's another cartoon. )
Clever and practical solution.

The only tip I would add would be to hit the cut end of the paper clip with a file (even an emery board would do) to remove any sharp burrs.
Good point, jdm, and I did think about mentioning that in my OP. I found that I didn't have a sharp edge on the paper clip after I cut it, for some reason, so I didn't have to file anything. Nonetheless, I did make sure to cover both ends pretty well with the sticky floral tape.
ozij wrote:Wonderful!!!

Are you left handed? Or is it easier to reach across your face and remove the left strap with your right hand?

O.
Thanks, O. No, I'm right-handed, and I'm not really sure why I chose the left side for the amendment... But this fix could be done on either side (or both, for that matter).

I think I'll apply the same fix to the over-the-ears strap, too, so that I never have to stretch the headgear again.
Veni, vidi, Velcro. I came, I saw, I stuck around.

Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy. ;))

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.

User avatar
PST
Posts: 986
Joined: Sat May 30, 2009 9:56 pm

Re: My FullLife Easy-Release Fix

Post by PST » Mon Aug 17, 2009 10:48 am

BleepingBeauty, you will have to change your motto to "I came, I sew, I stuck around."

User avatar
mdintx
Posts: 296
Joined: Sun Jun 14, 2009 4:24 am
Location: DFW-TX

Re: My FullLife Easy-Release Fix

Post by mdintx » Mon Aug 17, 2009 11:48 am

I was thinking that getting a couple of the headgear buckles for Swift:

https://www.cpap.com/productpage/resmed ... ft-II.html

and using extra straps from unused headgear might also work.

User avatar
2ndGenCPAPgal
Posts: 79
Joined: Tue Jun 23, 2009 3:56 pm
Location: Rockville, MD

Re: My FullLife Easy-Release Fix

Post by 2ndGenCPAPgal » Mon Aug 17, 2009 12:03 pm

I found this while looking at replacement clips. I think those clips are available at craft/fabric stores so I might try to make something similar.

https://www.cpap.com/productpage/Respir ... -Clip.html

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: 13cmH2O

User avatar
socalmonkey
Posts: 260
Joined: Wed Jul 01, 2009 12:11 pm
Location: Southern California

Re: My FullLife Easy-Release Fix

Post by socalmonkey » Mon Aug 17, 2009 12:59 pm

Very innovative! Are you sure you won't end up with impression of a paper clip on your face?

But you do have too much time on your hands to come up with that solution!

User avatar
socalmonkey
Posts: 260
Joined: Wed Jul 01, 2009 12:11 pm
Location: Southern California

Re: My FullLife Easy-Release Fix

Post by socalmonkey » Mon Aug 17, 2009 1:02 pm

2ndGenCPAPgal wrote:I found this while looking at replacement clips. I think those clips are available at craft/fabric stores so I might try to make something similar.

https://www.cpap.com/productpage/Respir ... -Clip.html
I like that! Now that is a simple easy fix!

User avatar
rogelah
Posts: 125
Joined: Mon Apr 20, 2009 5:30 am
Location: Weston, Florida

Re: My FullLife Easy-Release Fix

Post by rogelah » Mon Aug 17, 2009 3:12 pm

Okay Sleeping Beauty, your next assignment, should you choose to accept it, is to design a fix for "healthcare".

This message will self-destruct in 60 seconds...60...59...58...57...
ResMed VPAP III ST
IPAP:23.0, EPAP:12.0, RR:15, RT:250ms, IPAPmax:2.0s, MAX I:E:1:1, IPAPmin:1.0s

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: My FullLife Easy-Release Fix

Post by DoriC » Mon Aug 17, 2009 3:26 pm

BB, I really admire your creativity and patience. I can't even sew a straight hem. But I'm asking the same question that I'm sure has been asked here a million times, "Why can't the designers of these masks and machines recognize the flaws in their products in the first place"? Somehow looking at all your photos makes me angry if you know what I mean. Thanks for posting. Dori

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

fuzzy96
Posts: 448
Joined: Sun Sep 02, 2007 5:34 pm
Contact:

Re: My FullLife Easy-Release Fix

Post by fuzzy96 » Mon Aug 17, 2009 4:16 pm

good job bb!!!!

so when ya gunna weld new rocker panels on my taurus for me??????????????????????????

the one picture looked like two inch worms doin the hoochie-coochie Image

this aint no wierd rorshak test is it?

User avatar
BleepingBeauty
Posts: 2454
Joined: Thu Apr 02, 2009 5:30 pm
Location: Aridzona ;-)

Re: My FullLife Easy-Release Fix

Post by BleepingBeauty » Mon Aug 17, 2009 6:15 pm

PST wrote:BleepingBeauty, you will have to change your motto to "I came, I sew, I stuck around."
Good one!
2ndGenCPAPgal wrote:I found this while looking at replacement clips. I think those clips are available at craft/fabric stores so I might try to make something similar.

https://www.cpap.com/productpage/Respir ... -Clip.html
Definitely an easier fix than what I constructed. I didn't even know these things existed. Thanks.
socalmonkey wrote:Very innovative! Are you sure you won't end up with impression of a paper clip on your face?
Thanks. And no, there's no paper clip impression on my cheek. The mask's side bar angles away from the face. So there!
But you do have too much time on your hands to come up with that solution!
Hey! I never claimed to have a life.
rogelah wrote:Okay Sleeping Beauty, your next assignment, should you choose to accept it, is to design a fix for "healthcare".

This message will self-destruct in 60 seconds...60...59...58...57...
Ha! I choose NOT to accept such an assignment. (I'm sleep-deprived and can't be trusted with such a task.)
DoriC wrote:BB, I really admire your creativity and patience. I can't even sew a straight hem. But I'm asking the same question that I'm sure has been asked here a million times, "Why can't the designers of these masks and machines recognize the flaws in their products in the first place"? Somehow looking at all your photos makes me angry if you know what I mean. Thanks for posting. Dori
Thanks, Dori. I'm not a seamstress, either, but this involved no real sewing. Your question is a really good one, and it's sad that there's no answer. I figured out early on that the people who design CPAP equipment likely don't use it, themselves. (If they did, they might realize deficiencies or flaws and fix them before the product ever went to market.) We're on our own, so I just did what I could to come up with a fix for myself. Granted, a lot more work went into my fix than purchasing those handy little strap clips, but I didn't know about them. *shrug*
fuzzy96 wrote:good job bb!!!!
Thanks, fuzz!
so when ya gunna weld new rocker panels on my taurus for me??????????????????????????
Hmm. I haven't tried welding. Yet. I'll let you know.
the one picture looked like two inch worms doin the hoochie-coochie
You've got a dirty mind, fuzz. No wonder I like you.
this aint no wierd rorshak test is it?
Whether it was or it wasn't, I won't be telling your Freud-follower shrink what you saw in that pic. Your twisted mind is safe with me (and the 40,000 other members of the forum, I'm sure).
Veni, vidi, Velcro. I came, I saw, I stuck around.

Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy. ;))

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.

User avatar
timbalionguy
Posts: 888
Joined: Mon Apr 27, 2009 8:31 pm
Location: Reno, NV

Re: My FullLife Easy-Release Fix

Post by timbalionguy » Mon Aug 17, 2009 8:56 pm

Thanks for introducing me to a new 'engineering material'-- floral tape!
Lions can and do snore....