ResMed Autoset LCD Reading

Post by atab » Thu Jul 23, 2009 7:07 am

I got switched from Elite II to Autoset for auto-titration. My earlier sleep lab titrated pressure was 6 and that wasn't working. I used to write down leak, AHI, AI after reading off LCD of elite and I do the same with autoset. On autoset "pressure" is also readable. Last night it read 11.8. Autoset is set to 6-12. So what exactly is the pressure reading here? Highest pressure during the night, average during apneas, or average through-out that night?

-atab

_________________
MachineMask
Additional Comments: AutoSet setting 9-18

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: ResMed Autoset LCD Reading

Post by jdm2857 » Thu Jul 23, 2009 8:28 am

Your machine is displaying the 95 centile pressure. That is the pressure that the machine was at or below for 95% of the night. It was above that pressure for 5% of the night.

When you switch from day to week or month, etc. the machine desplays the median of the displayed values for the days in that period. The median is the value for which half of the days the data was below that value and half of the days the data was above that value. It is an average calculation that is less affected by extreme data (say a very bad day) than the mode (regular average).

If you have the AutoSet permanently, you should consider getting the ResScan software and a card reader.
jeff

User avatar
atab
Posts: 103
Joined: Sun May 24, 2009 9:50 am

Re: ResMed Autoset LCD Reading

Post by atab » Thu Jul 23, 2009 9:13 am

Thank you jdm.

So I did not understand something about autoset. My pressure is set to 6-12. I thought the pressure stays at 6, and when an apnea happens it goes up as necessary and then comes down after the apnea incident. If that is the case 95th percentile would be very close to 6. (Last night I had AHI 7.4 and AI 2.1, that is a very small fraction of night .) The reading of 11.8 says that I was near 12 all night. So is it the case that when the pressure goes up, it stays there? That explains why I had a hell of a problem with mask leaking. I think pressure of 12 is very different form 6.

Yes, I will buy software if I stay on autoset. At the moment it is a temporary machine for re-titration.

atab.

_________________
MachineMask
Additional Comments: AutoSet setting 9-18

User avatar
Pugsy
Posts: 65345
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Autoset LCD Reading

Post by Pugsy » Thu Jul 23, 2009 9:31 am

atab wrote:The reading of 11.8 says that I was near 12 all night. So is it the case that when the pressure goes up, it stays there?
Yep, it stays in that range because it needs to be there or it would fall back to 6 more often. The software will show you what it is responding to and it will become clearer to you.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: ResMed Autoset LCD Reading

Post by jdm2857 » Thu Jul 23, 2009 9:37 am

The software responds to snores, flow limitations, and apneas (only at pressures less than 10). It increases pressure in response to events, and slowly decreases the pressure when there are no events. The fact that you pressure stayed up (at a pressure over 10) means that you must have been having either snores or flow limitations that kept the pressure up.

Unfortunately, ResMed software does not show snores or flow limitations, so it is hard to relate pressure changes to events.
jeff

User avatar
atab
Posts: 103
Joined: Sun May 24, 2009 9:50 am

Re: ResMed Autoset LCD Reading

Post by atab » Thu Jul 23, 2009 9:53 am

That clears up my understanding. Thanks all.

Last night I was waking up frequently with mask leaks. I now attribute this to higher pressure. I was on 6 with Elite. I use quattro FF. I guess the mask has to be looser for higher pressure. But how do I go to sleep at a pressure of 6 (starting pressure of APAP) ready for a pressure of 12?

atab

_________________
MachineMask
Additional Comments: AutoSet setting 9-18

jnk
Posts: 5781
Joined: Mon Jun 30, 2008 3:03 pm

Re: ResMed Autoset LCD Reading

Post by jnk » Thu Jul 23, 2009 10:25 am

atab wrote: . . . My pressure is set to 6-12. I thought the pressure stays at 6, and when an apnea happens it goes up as necessary and then comes down after the apnea incident. . . .
An auto attempts to be at the lowest, yet still effective, pressure at any given moment for preventing events, as best it can tell. That's the idea, anyway, as I understand it.
atab wrote: . . . The reading of 11.8 says that I was near 12 all night. . . .
I don't think you can know that without software. The reading tells you only a number that you spent 95% of the night at OR BELOW. You MAY have spent much of the night WELL below. Then again, not. But it does seem your pressure did top out at the limit, though, at some point, so the upper-limit pressure may need to go higher.
atab wrote:. . . So is it the case that when the pressure goes up, it stays there? . . .
My understanding is that it comes down gradually if it senses nothing to make it stay there. It does so gradually so as not to disturb your sleep needlessly or cause events.

User avatar
atab
Posts: 103
Joined: Sun May 24, 2009 9:50 am

Re: ResMed Autoset LCD Reading

Post by atab » Thu Jul 23, 2009 11:45 am

I don't think you can know that without software. The reading tells you only a number that you spent 95% of the night at OR BELOW. You MAY have spent much of the night WELL below. Then again, not. But it does seem your pressure did top out at the limit, though, at some point, so the upper-limit pressure may need to go higher.
Good point. I am a little rusty on my statistics. Yes, 95th percentile can be well off of the mean (or even median). Thanks for pointing out.

atab.

_________________
MachineMask
Additional Comments: AutoSet setting 9-18

User avatar
ricochetv1
Posts: 181
Joined: Sat Jul 04, 2009 4:37 pm

Re: ResMed Autoset LCD Reading

Post by ricochetv1 » Thu Jul 23, 2009 3:19 pm

It slowly ramps down to the low pressure over a 20 min. period.

joeydan
Posts: 9
Joined: Thu Aug 07, 2008 1:08 pm

Re: ResMed Autoset LCD Reading

Post by joeydan » Thu Jul 23, 2009 5:38 pm

I sometimes wake up and check my Resmed AutoSet 11 with my mask on and machine running. To my surprise the LCD shows a high pressure even though I'm awake and there is no flow limitation.I've watched for as long as 1 minute with no change in the pressure.Has anyone else noted this and should it take 20 minutes to drop to settling pressure?

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: ResMed Autoset LCD Reading

Post by jdm2857 » Thu Jul 23, 2009 7:16 pm

I shouldn't take 20 minutes for the pressure to begin dropping, but it may not happen in a minute either. Because of conscious control, breathing when awake is less regular than when asleep, and that can confuse the algorithm. For example, holding your breath for 10 seconds can be scored an apnea, and shallow breaths can be scored hypopneas.
jeff

User avatar
ricochetv1
Posts: 181
Joined: Sat Jul 04, 2009 4:37 pm

Re: ResMed Autoset LCD Reading

Post by ricochetv1 » Fri Jul 24, 2009 8:46 am

And that is the exact reason I set a long settling time on mine.

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: ResMed Autoset LCD Reading

Post by Slinky » Fri Jul 24, 2009 9:58 am

And a high leak rate can disrupt the accuracy of your data. Rule # 1, get your leak under control. It took a long time for forum members to get that to sink into my thick skull. The addition of a chin strap PROVED just how smart our forum members are!!! My sleep doctor and DME supplier's RT kept telling me these xPAPs can compensate quite nicely for leaks. And maybe they can - but those high leaks DO make the data unreliable.

Score: Forum Members 1, Sleep Profession 0

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
ricochetv1
Posts: 181
Joined: Sat Jul 04, 2009 4:37 pm

Re: ResMed Autoset LCD Reading

Post by ricochetv1 » Fri Jul 24, 2009 10:22 am

Is a .18 L/s rate good?

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: ResMed Autoset LCD Reading

Post by Slinky » Fri Jul 24, 2009 12:58 pm

Since you are using a non-Resmed mask (or is the LIberty a Resmed mask???) - I don't know.

0.4 L/s or 24 L/M are the suggested limit of Leak. But since the Resmeds subtract the allowed vent rate for their own masks IF the correct Mask Selection has been set I can't say because I don't know if the Liberty is a Resmed mask or not nor do I know the Mask Selection being used and if it is the correct one for the Liberty if it is a Resmed mask.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.