DOC WOULD NOT PRESCRIBE AUTO C-FLEX

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Tue Aug 02, 2005 9:57 pm

Not only conflict of interest, but possibly a violation of the Sherman anti-trust laws (it doesn't have to be a big company). Like any other prescription you get from a doctor, you should be able to take that prescription to any pharmacist (or in this case a DME provider) of your choice to have it filled. I did some searching on this subject the other day (but didn't have a whole lot of time). What I DID find on one item that came up was a reference to the AMA Code of Ethics. I just don't see how they can MAKE you buy the equipment from THEM.

It ain't ethical (and maybe not legal either)!

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: DOC WOULD NOT PRESCRIBE AUTO C-FLEX

Post by dsm » Tue Aug 02, 2005 10:48 pm

rpalmer wrote:
Am I just being paranoid?

Rol
Nup, It is a conflict of interest as I see it.

In my case my family doctor sent me to a respiratory specialist (doctor). He sent me to a Sleep Clinic. The Sleep clinic did the 2 nights & recommended a cms of 15. They are the ones who than advised what CPAP gear was on the market & who to go visit to see it & buy it from. They did (as has already been discussed at length) warn against AUTOs *but* left it to us to decide for ourselves.

My family doctor has *no* idea what xPAP gear I have. My respiratory specialist has no idea what xPAP gear I bought. The sleep clinic know because I told them & they are being very supportive about it.

That is how it should work.

But the issue here is that we don't have the quirky laws you seem have in the US re xPAP nor a draconian health regulatory body to threaten anyone.
But, the biggest difference here seems to be the involvement of health insurance companies. That seems to be the spoiling factor there.

As an aside ...
The new ResMed S8 I have just been looking at has written on the bottom ...

"CAUTION: USA ONLY: Federal law restricts this device to sale by or on oder of a physician"



Cheers

DSM

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): resmed, CPAP

xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

fstanmyre
Posts: 118
Joined: Tue May 10, 2005 9:06 am
Location: Dallas

for DSM, runaway pressures

Post by fstanmyre » Wed Aug 03, 2005 8:42 am

I have both the Remstar Auto and my newly purchased Goodknight 420E. Personally, I've never experienced a pressure "runaway" with either. Thinking about it, I'm not sure I even know what that means. Does it mean that the pressure keeps going up to the upper apap limit for no apparent reason? Anybody have a comment in this regard?

I do have a leakage problem with my FF mask which I use whenever I have nasal congestion, ususally due to pollution or allergies here in big D (normally use a Breeze). This case I have a built in safety valve since the FF mask blows it's seal at any pressure over about 12. I can't imagine any of the FF masks keeping a seal at higher pressures. Maybe some folks are confusing large mask leaks with "runaways" but this is a mask problem, not an apap problem.

Last night just for the heck of it I set the apap minimum pressure at 8 which is about my 90% pressure and the max at 15. With the Breeze the machine ran very nicely most of the time right at 8 with an occasional excursion to 11 obviously in response to obstructive apnea. AHI for almost 7 hours was 2.4.

Reading some of the papers on the subject, there seem to be so many variables in how some people respond to cpap or apap. For me apap is just great resulting in lower pressures most of the night and AHI's under the magic 5 number. I haven't called the lady in Australia yet because I'm still digesting some of the papers on the subject, which she should be doing btw.

Meanwhile however, it seems to be plain old common sense that patients should be provided with a good apap machine unless the doctor has some good reason why it should not be provided, which I have not heard yet from anybody other than the somewhat nebulous "runaway" argument. Apap obviously works very well for many of us in this forum but if it doesn't work for some, possibly because of mask problems associated with higher pressures, then you can always turn the apap off and go straight cpap. What the heck is so complicated about this approach?

Fred Stanmyre

User avatar
neversleeps
Posts: 1141
Joined: Wed Apr 20, 2005 7:06 pm
Location: Minnesota

Re: DOC WOULD NOT PRESCRIBE AUTO C-FLEX

Post by neversleeps » Wed Aug 03, 2005 8:58 am

dsm wrote:"As an aside ...
The new ResMed S8 I have just been looking at has written on the bottom ...

CAUTION: USA ONLY: Federal law restricts this device to sale by or on oder [sic] of a physician"
Good old USA.... The last bastion of ludicrous regulations. What one must surmise was originally devised to protect us, has, in this instance, resulted in preventing us from making a move without lining someone else's pockets. Certainly I am at much greater risk of dying from swallowing a bottle of pills I can easily legally obtain WITHOUT order of a physician than I am from buying a glorified leaf blower.

GRRRRR!

User avatar
rpalmer
Posts: 80
Joined: Wed Jan 26, 2005 11:30 am
Location: Maryland, but heart & soul are in the Swamp!
Contact:

Re: DOC WOULD NOT PRESCRIBE AUTO C-FLEX

Post by rpalmer » Wed Aug 03, 2005 10:33 am

neversleeps wrote:Good old USA.... The last bastion of ludicrous regulations. What one must surmise was originally devised to protect us, has, in this instance, resulted in preventing us from making a move without lining someone else's pockets. Certainly I am at much greater risk of dying from swallowing a bottle of pills I can easily legally obtain WITHOUT order of a physician than I am from buying a glorified leaf blower.

GRRRRR!
Unfortunately, neversleeps, I think that's called The Law of Unintended Consequences. As our Big Nanny in Washington seeks to control more & more of our lives, this is the inevitable result.
Or as Thomas Sowell wrote: "The government forces those who sell pharmaceutical drugs to list the possible side effects, even if only a few people will suffer those side effects. Unfortunately, the government itself never tells us about the bad side effects of the things it prescribes."

I think DSM lives in Canada, and they certainly don't take a back seat to anyone when it comes to government regulations, so it's interesting that he has so much more freedom when it comes to Dr/Sleep Clinic/DME/xPAP issues.

DSM's right. That is EXACTLY how it should work!
“The best cure for insomnia is to get a lot of sleep.â€

ozij
Posts: 10463
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Aug 03, 2005 12:07 pm

fstanmyre wrote:Does it mean that the pressure keeps going up to the upper apap limit for no apparent reason?
Yes. And if you experience it you know there's nothing nebulous about it...

O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Post by rested gal » Wed Aug 03, 2005 12:45 pm

you can always turn the apap off and go straight cpap. What the heck is so complicated about this approach?
Fred, I agree absolutely. A versatile autopap machine that can always be easily set to simple CPAP if "straight pressure" is what's needed makes sense to me.
Personally, I've never experienced a pressure "runaway" with either. Thinking about it, I'm not sure I even know what that means. Does it mean that the pressure keeps going up to the upper apap limit for no apparent reason?


Yes, that's what it means. Pressure running up to the top and pretty much staying there for no apparent reason, or more precisely, for no "good" reason. Runaways that are not being caused or continued by leaks of any kind.

The 420E auto that I dearly loved did that (pressure climbing and staying at whatever upper pressure I had it set for, even 19!!) until -SWS looked at my data and suggested I turn off IFL1 in the advanced settings. From that point on, the 420E behaved perfectly for me -- staying down around 9 or 10 most of the time.

Later, I bought a REMstar Auto from a lady who no longer needed it. She was having success with a dental device and was selling all her year-old "cpap" equipment. I had been following her posts for some time about her trials and tribulations not getting effective treatment, trying many different masks, and many different pressure settings.

On the REMstar auto she was experiencing constant pressure runaways...pressure climbing to whatever upper pressure she set. Same way the 420E had been doing me. Happily her REMstar never did a runaway with me.

I've read people reporting that the ResMed Spirit auto has done "runaway" pressure for them if the humidifier water chamber was filled above the "fill-line". I think I've also read that the Spirit's aggressive reaction to snores can also do it for a few people.

So, yes, pressure runaway on any autopap can and does happen for some people. That's why I think it's a good idea to keep a leash on the upper pressure setting... setting it no more than a few points above a person's titrated pressure, especially when using a new brand of autopap that you haven't tried before.

Helps also to have the software to take a look and see that the margin you've set at the top indeed does not have to be used. I want the extra cm's up there to be there for "just in case" I need more than usual on some night....not be a pressure the machine generally uses for me.

I'll hasten to say that it is unusual for "runaway" pressure to happen. Most autopaps work fine for most people, just as the two you're using are doing for you, Fred. Just because runaway pressure scenarios can happen to a few people (and did for about a month for me until I made a setting change) does not make me shy away in the least from using any autopap and recommending that others use them.

As with news, our ears tend to perk up and notice "bad news", paying little attention to good news. In the same way, I'd say people should NOT worry about the possibility of "pressure runaway" with autopaps. Most people report getting wonderful treatment from their autopaps. I sure do.

A few might experience pressure runaway, but it's usually easily remedied if you know what to do.

Or.... just switch the machine to straight cpap.

All said and done, I do think the reluctance and/or refusal of some medical professionals to prescribe autopaps (for whatever their reason) is a disservice to the many who might get much more comfortable treatment from an autopap rather than the straight cpap they're on.

I'm not a doctor, but I don't see how it could hurt to at least try other types of machine, be it cpap, bi-level, or autopap, to see what suits the person best. Any machine, no matter how advanced, can be set to straight cpap at any time, just as Fred said.

fnorette
Posts: 78
Joined: Fri Jul 01, 2005 6:01 pm

Runaways

Post by fnorette » Wed Aug 03, 2005 1:19 pm

I have used both a ResMed CPAP and the ResMed AutoPAP Spirit, and have experienced runaways with both. Interestingly, I AM a heavy snorer ... Rested Gal probably has made a good observation.


fstanmyre
Posts: 118
Joined: Tue May 10, 2005 9:06 am
Location: Dallas

Post by fstanmyre » Wed Aug 03, 2005 1:42 pm

Hmmmm!, a runaway with cpap?

Fred Stanmyre

User avatar
WAFlowers
Posts: 1172
Joined: Wed Jul 27, 2005 10:59 am
Location: Clearwater FL
Contact:

Post by WAFlowers » Wed Aug 03, 2005 1:47 pm

I'm pleased, happy, thrilled (take your pick or use your own adjective) to report that I finally have my prescription in hand. Not only that, but the doctor actually prescribed an auto unit. In fact, he specifically prescribed the 420E and also specified a heated humidifier.

YIPEE! I've already faxed it to cpap.com and am about to followup with a phone call to get the order in.

My sympathies to anyone who has a doc who isn't as cooperative as mine is.

The CPAPer formerly known as WAFlowers

fstanmyre
Posts: 118
Joined: Tue May 10, 2005 9:06 am
Location: Dallas

osij

Post by fstanmyre » Wed Aug 03, 2005 1:58 pm

osij says she does experience runaways for no apparent reason. I suggest there is a reason and it needs to be deciphered. RG had the problem and found the reason. Anyhow, solution simple, if you can't find the reason, switch to straight cpap as long as you are getting AHI's below 5. If you are not getting the low AHI's then a good sleep clinic is the answer.

Fred Stanmyre

User avatar
WillSucceed
Posts: 1031
Joined: Sun Nov 07, 2004 7:52 am
Location: Toronto, Ontario

Post by WillSucceed » Wed Aug 03, 2005 2:10 pm

Good old USA.... The last bastion of ludicrous regulations.
Last bastion of ludicrous regulations, you say? Too bad you folk can't figure out gun control. It would seem to make more sense to encourage folk to take responsibility for, and control of, their health care and limit their access to those things that really hurt them, like guns.
I think DSM lives in Canada, and they certainly don't take a back seat to anyone when it comes to government regulations, so it's interesting that he has so much more freedom when it comes to Dr/Sleep Clinic/DME/xPAP issues
You are right about this... we have regulations up the ying-yang and some of them are good ones.
Buy a new hat, drink a good wine, treat yourself, and someone you love, to a new bauble, live while you are alive... you never know when the mid-town bus is going to have your name written across its front bumper!

fstanmyre
Posts: 118
Joined: Tue May 10, 2005 9:06 am
Location: Dallas

fnorette

Post by fstanmyre » Wed Aug 03, 2005 2:19 pm

Derek's studies as I recall did show a significant Remstar response to snoring. If this is shown to be a problem during the sleep study it might have been better for the sleep lab or doctor to prescribe the Goodknight which does allow for adjustments to the machines response to apnea, acoustic vibrations(snoring), and flow limitation.. I've not fiddled with these settings yet, just using the defaults, but no doubt will in due course.

Anyhow point is once again, did fnorette's sleep lab prescribe the right machine for her particular condition?
Fred Stanmyre

User avatar
rpalmer
Posts: 80
Joined: Wed Jan 26, 2005 11:30 am
Location: Maryland, but heart & soul are in the Swamp!
Contact:

CONGRATULATIONS, BILL!

Post by rpalmer » Wed Aug 03, 2005 3:17 pm

WAFlowers wrote:I'm pleased, happy, thrilled (take your pick or use your own adjective) to report that I finally have my prescription in hand. Not only that, but the doctor actually prescribed an auto unit. In fact, he specifically prescribed the 420E and also specified a heated humidifier.

YIPEE! I've already faxed it to cpap.com and am about to followup with a phone call to get the order in.
Congratulations, Bill! Looks like the wait was well worth it.

We'll look forward to reading your posts in a few days reporting how much better you feel.

Rol

“The best cure for insomnia is to get a lot of sleep.â€

fstanmyre
Posts: 118
Joined: Tue May 10, 2005 9:06 am
Location: Dallas

For WillSucceed

Post by fstanmyre » Wed Aug 03, 2005 3:18 pm

DSM lives in Australia.
Fred Stanmyre