I have a respironics bi-pap machine that suddenly started making a whistleing noise and now the blower is malfunctioning. Is there anywhere that repairs these things and has anyone had this type of problem??? I think it may have gotten wet during transit during thanksgivin,
Thanks
bipap machine broken
- BallBuster
- Posts: 51
- Joined: Fri Nov 28, 2008 9:06 pm
Re: bipap machine broken
FYI I heard that the people in the front row are gathering the eggs and tomatoes.
Re: bipap machine broken
Is it just the water that might have happened... or could it be something else?
If it's just the water... have you tried taking the card out and running it for awhile (no mask, no hose... just let it run). See if it'll clear itself up?
If it's just the water... have you tried taking the card out and running it for awhile (no mask, no hose... just let it run). See if it'll clear itself up?
Re: bipap machine broken
If it got wet you are much better off to make sure it is completely dried inside BEFORE applying power to it again.
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: bipap machine broken
Like cflame1 said sometimes they will dry out and be OK. I have not found or heard of anyone that works on them except for Respirionics. Several have tried to get parts to fix them themselves with no luck. If it is under 2 years old (from date it was shipped by Respirionics) they will fix it under warrenty. One restriction is that it must be sent back to the factory by a DME. They will not receive it from you. If you are dealing with a decent DME they will give you a loaner while it is gone. Older than 2 years it can still be sent back and they will give you an estimate to see if it cost more to repair or replace. Same shipping regulations apply. Does the screen give you an error code? Might give you some idea as to whether it is worth sending in. Good luck!!
Jerry
Jerry
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
Re: bipap machine broken
This advice will differ a bit between brands & models. If you have the old 'tank' style Bipap then I gan guarantee the following advice to be very sound
This link shows a brand new Bipap Pro II that had been put in its carry bag by the owner, while the H/H still had water in it.
http://www.internetage.ws/cpapinfo/remstar-bipap-dis-1/
So, a couple of things to watch out for.
1. If cleaning around the air filter with tissues make certain no tissue paper gets left in the air in hole
2. If water gets into the air exit port from the H/H - the water can find its way to the blower bearings and they will rust (see photos)
3. If you do get water into the air exit port - the water won't get into the electronics (that part of the machine is separate) - the best advice I can offer is to run the machine for a while with the filer removed BUT DON'T LET ANYTHING GET SUCKED IN. The aim is to dry the water out before it affects the bearings
4. If water gets spilled onto a cpap it CAN get into the electronics - that gets tricky - for myself, I would open the unit up & hand dry it (while it isn't plugged in )
Good luck
DSM
This link shows a brand new Bipap Pro II that had been put in its carry bag by the owner, while the H/H still had water in it.
http://www.internetage.ws/cpapinfo/remstar-bipap-dis-1/
So, a couple of things to watch out for.
1. If cleaning around the air filter with tissues make certain no tissue paper gets left in the air in hole
2. If water gets into the air exit port from the H/H - the water can find its way to the blower bearings and they will rust (see photos)
3. If you do get water into the air exit port - the water won't get into the electronics (that part of the machine is separate) - the best advice I can offer is to run the machine for a while with the filer removed BUT DON'T LET ANYTHING GET SUCKED IN. The aim is to dry the water out before it affects the bearings
4. If water gets spilled onto a cpap it CAN get into the electronics - that gets tricky - for myself, I would open the unit up & hand dry it (while it isn't plugged in )
Good luck
DSM