I picked up my AutoSet II today!!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
echo
Posts: 2399
Joined: Wed Jul 18, 2007 4:20 pm

Re: I picked up my AutoSet II today!!

Post by echo » Tue Nov 25, 2008 6:43 am

2girlsmom, hope you had a great night! We'll be looking forward to those reports STAT!

BTW I totally understand, I finally got my sleep study results today and I'm as giggly (giddy?) as a school girl
PR System One APAP, 10cm
Activa nasal mask + mouth taping w/ 3M micropore tape + Pap-cap + PADACHEEK + Pur-sleep
Hosehead since 31 July 2007, yippie!

User avatar
Paul56
Posts: 1016
Joined: Thu Sep 25, 2008 1:38 pm
Location: Texas

Re: I picked up my AutoSet II today!!

Post by Paul56 » Tue Nov 25, 2008 7:21 am

Bookbear wrote:
Paul56 wrote: Settling = Ramp
Settling is the terminology used with the Resmed Auto machines.
Well, not exactly. The ResMed has a ramp feature in cpap mode, which functions like other manufacturer's ramp feature... it gradually increases up from a given starting pressure up to the the single given therapy pressure over a given period of time, to allow the patient to get off to sleep before full treatment pressure is applied. Settling is used in apap mode, and it starts with the minimum given pressure and stays there for the given settling time, ignoring apnea or flow limit events until that given time has expired, then begins to respond to flow limitations or apneas with pressure increases up to the maximum given pressure. It does not gradually increase pressure at all during the settling time, unlike the ramp mode, which does.

You have a good machine, and I trust it will serve you well.

Hope this helps!
Yes, I understand that... however since the machine is being used in AUTO in this case I did not want to confuse the issue.

Okay... how did the night go?

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: AHI ~60 / Titrated @ 8 / Operating AutoSet in CPAP mode @ 12

User avatar
GuyK
Posts: 163
Joined: Mon Aug 18, 2008 7:23 pm
Location: MA

Re: I picked up my AutoSet II today!!

Post by GuyK » Tue Nov 25, 2008 7:34 am

Yes -- I'm anxious to hear too. How did it go?

I've been using an AutoSet II for about two months, and everything has been going fine.

However, something popped up recently -- I got a card reader, card and ResScan, and I downloaded my first set of data (it is one step more complicated than that -- I have a Mac, but I run XP under something called "Boot Camp" -- the good part is that I had absolutely no issues with the software or reading the card). The summary graphs are exactly in line with what I've been writing down every morning.

But -- I get a message that says "detailed data not available", and I see the summary data. I've been looking through the clinical manuals for the AutoSet II and with ResScan, but I don't see anything about enabling detailed data.

The bottom line is that at this point, I'm not sure if detailed data is even available for this machine.

Anyone able to educate me?

thanks,
Guy

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: ResScan 3.5 and card reader
Ready to fight the forces of evil, right after I finish this nap...

User avatar
2girlsmom
Posts: 235
Joined: Tue Oct 30, 2007 3:19 pm
Location: Illinois

Re: I picked up my AutoSet II today!!

Post by 2girlsmom » Tue Nov 25, 2008 8:51 am

Guy, when you d/l your data you may need to click the box to get it to d/l the detailed data. It should do up to 5 sessions.

Well last night was well, okay. I expected to feel totally different this am but I feel the same. AHI was under 5 but I really think I need to bump up the top pressure a bit. I did wake up at some point with a sore arm because I had been laying in the same position too long, so I think that is a good sign that I slept more deeply than normal. I do think I am going to increase the top # to 12, because I want more benefit than I got last night. The one good thing is I did not have the big clusters of events that I normally see. There weren't near as many together. So things are moving in the right direction.

Thanks for following my saga guys!

_________________
MaskHumidifier
Additional Comments: CPAP pressure 12.

User avatar
Paul56
Posts: 1016
Joined: Thu Sep 25, 2008 1:38 pm
Location: Texas

Re: I picked up my AutoSet II today!!

Post by Paul56 » Tue Nov 25, 2008 9:48 am

2girlsmom wrote:Guy, when you d/l your data you may need to click the box to get it to d/l the detailed data. It should do up to 5 sessions.

Well last night was well, okay. I expected to feel totally different this am but I feel the same. AHI was under 5 but I really think I need to bump up the top pressure a bit. I did wake up at some point with a sore arm because I had been laying in the same position too long, so I think that is a good sign that I slept more deeply than normal. I do think I am going to increase the top # to 12, because I want more benefit than I got last night. The one good thing is I did not have the big clusters of events that I normally see. There weren't near as many together. So things are moving in the right direction.

Thanks for following my saga guys!
That reported AHI is very good and encouraging.

Did you check on the AI & HI?

Keep in mind with the Resmed that you pretty much need to cut the HI in half to get a realistic number since the machine scores hypopneas very aggressively.

Where did the pressure spend most of the night? Did it bounce up against the max range number and stay here? If so then yes you should adjust that number upwards but don't be too quick to make changes... give yourself time to adjust and the machine to do its thing.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: AHI ~60 / Titrated @ 8 / Operating AutoSet in CPAP mode @ 12

User avatar
GuyK
Posts: 163
Joined: Mon Aug 18, 2008 7:23 pm
Location: MA

Re: I picked up my AutoSet II today!!

Post by GuyK » Tue Nov 25, 2008 12:18 pm

2girlsmom wrote:Guy, when you d/l your data you may need to click the box to get it to d/l the detailed data. It should do up to 5 sessions.
Thanks, 2girlsmom, I'll check into that tonight.

BTW, although the mechanics of therapy have been going fine for me, I don't always feel a dramatic effect either. There are times I still feel groggy on waking and mid-afternoon, although I think that things have been better overall. I know I'm getting REM every now and then, and while I used to frequently wake up in the middle of the night and not be able to get back to sleep, it seems like that hasn't happened since I started therapy.

I recall many people on this forum saying they didn't feel like they were comfortable until about the third month. I'm looking forward to that.

The last time I visited my PCP, which was about week 3, I had a dramatically reduced blood pressure. Lowest number I've seen in years. So I bought a B.P. machine so that I can monitor it. Overall, I think the numbers are fairly good. I hope to make a case to my PCP that I can come off my B.P. medication (while monitoring, of course) or at least reduce the dosage.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: ResScan 3.5 and card reader
Ready to fight the forces of evil, right after I finish this nap...

User avatar
2girlsmom
Posts: 235
Joined: Tue Oct 30, 2007 3:19 pm
Location: Illinois

Re: I picked up my AutoSet II today!!

Post by 2girlsmom » Tue Nov 25, 2008 4:05 pm

Median pressure was 10.8, 95%ile was 11 and max pressure was 11. I went ahead and put max at 12.

AHI was 3.7 with an AI of 0.4 and HI of 3.3. So it is good, but I went to bed late and was woken up at 6:30 by my 5yo. I still got some clustering of events even @ 11. I did wake up a few times as well. It just wasn't the magical night I had envisioned. I do think we're headed in the right direction so that is encouraging.

_________________
MaskHumidifier
Additional Comments: CPAP pressure 12.

twocents
Posts: 34
Joined: Sun Oct 26, 2008 5:15 pm
Location: Long Island
Contact:

Re: I picked up my AutoSet II today!!

Post by twocents » Tue Nov 25, 2008 4:31 pm

enjoy... way to go..
two cents ¢¢

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: I picked up my AutoSet II today!!

Post by Slinky » Tue Nov 25, 2008 7:54 pm

Thanks for sharing, 2GirlsMom!! Maybe not a miracle occurred but things are looking and sounding good. Just hang in there. Patience is a virtue.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Last edited by Slinky on Tue Nov 25, 2008 8:15 pm, edited 1 time in total.
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: I picked up my AutoSet II today!!

Post by dsm » Tue Nov 25, 2008 8:05 pm

2girlsmom wrote:Guy, when you d/l your data you may need to click the box to get it to d/l the detailed data. It should do up to 5 sessions.

Well last night was well, okay. I expected to feel totally different this am but I feel the same. AHI was under 5 but I really think I need to bump up the top pressure a bit. I did wake up at some point with a sore arm because I had been laying in the same position too long, so I think that is a good sign that I slept more deeply than normal. I do think I am going to increase the top # to 12, because I want more benefit than I got last night. The one good thing is I did not have the big clusters of events that I normally see. There weren't near as many together. So things are moving in the right direction.

Thanks for following my saga guys!
That danged adrenaline just messed up your sleep - i know you will grow into it

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
2girlsmom
Posts: 235
Joined: Tue Oct 30, 2007 3:19 pm
Location: Illinois

Re: I picked up my AutoSet II today!!

Post by 2girlsmom » Tue Nov 25, 2008 9:20 pm

I'll post in the morning about my night tonight. My husband is off so I may even be able to sleep in if I need to. I am looking forward to seeing how increasing the max to 12 will go for me.

_________________
MaskHumidifier
Additional Comments: CPAP pressure 12.

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: I picked up my AutoSet II today!!

Post by Slinky » Tue Nov 25, 2008 10:37 pm

Hey, DSM may well have hit the nail on the head! I know my first two titrations I was so excited about finally getting som RELIEF and some SLEEP I couldn't sleep for beans. And worse than you - I only slept 48 minutes out of 6 hours bedtime so they brought me back in for another titration and I only slept 98 minutes out of 6 hours bedtime! And its NOT like I hadn't had some experience sleeping w/a mask and pressure and wires and all.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
2girlsmom
Posts: 235
Joined: Tue Oct 30, 2007 3:19 pm
Location: Illinois

Re: I picked up my AutoSet II today!!

Post by 2girlsmom » Tue Nov 25, 2008 11:00 pm

I think adrenaline did keep me from sleeping well. And, I drank plenty of caffeine. Today I had only one soda and it was very early so I am confident I will konk out when I head to bed shortly.

_________________
MaskHumidifier
Additional Comments: CPAP pressure 12.

User avatar
kopoloff
Posts: 398
Joined: Mon Nov 03, 2008 5:20 am
Location: Melbourne Australia

Re: I picked up my AutoSet II today!!

Post by kopoloff » Wed Nov 26, 2008 12:30 am

As i understand it, the S8 Autoset only stores 4 sessions of detailed data, so you have to download at least every 4 days. Once you've downloaded it and saved it, of course you have it forever.

K

User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Re: I picked up my AutoSet II today!!

Post by Snoredog » Wed Nov 26, 2008 1:19 am

2girlsmom wrote:I'll post in the morning about my night tonight. My husband is off so I may even be able to sleep in if I need to. I am looking forward to seeing how increasing the max to 12 will go for me.
Go ahead and put the Maximum up at 20 cm, no use letting that getting in the way to skew your results. If pressure during the night bumps into that 12 cm pressure it will throw off the 95% data and the rest of the data won't be that accurate either.

Since you are already over 10 cm pressure, it is only going up to 11 and higher due to Flow Limitations and Snore. FL's will show up under HI reason they are always higher than compared to other machines.

What you should go by is Leak, and AI. Get AI down to about 1 by increasing the Minimum pressure up and you'll have it, once you get AI=1 then start fine tuning in .2 to .5 pressure increases until HI drops below 5 and you will have it. Leak must me within limits for any of the data is to be considered accurate.
someday science will catch up to what I'm saying...